The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

View Full Version : Part 11 - Changes to a system validated to Part 11 - What to revalidate?


Pataha
7th September 2005, 03:53 PM
Hi all,

If one has validated a system to part 11, if you make a change to the system, can you verify just that element of the change or should you re validate the whole system?

Thanks for your help.

Al Rosen
7th September 2005, 04:07 PM
Hi all,

If one has validated a system to part 11, if you make a change to the system, can you verify just that element of the change or should you re validate the whole system?

Thanks for your help.From the FDA Guidance Document, General Principles of Software Validation.4.7. SOFTWARE VALIDATION AFTER A CHANGE
Due to the complexity of software, a seemingly small local change may have a significant global system
impact. When any change (even a small change) is made to the software, the validation status of the
software needs to be re-established. Whenever software is changed, a validation analysis should
be conducted not just for validation of the individual change, but also to determine the extent
and impact of that change on the entire software system. Based on this analysis, the software
developer should then conduct an appropriate level of software regression testing to show that
unchanged but vulnerable portions of the system have not been adversely affected. Design controls and
appropriate regression testing provide the confidence that the software is validated after a software
change.