I am in the process of framing the procedure for calibrating the on-line pH meter.
Can any one help on how to calibrate the on line pH meter.
Regards / Raghu
amanbhai
20th March 2009, 05:03 AM
What we do is to dip this rod in the CRM and records its findings. this we call verification/ calibration and records its results. The same work instruction we have written.
SilverHawk
20th March 2009, 07:16 AM
I thought it woul dbe standardization of the pH solution to validate the pH Meter. I would prepare 3 standardised solutions of pH = 3; pH = 7 and pH = 10 to be used for calibrating the pH meter and make necessary adjustments when the probe is immersed in each pH solutions.
It is verification / validation or calibration of pH Meter?
v9991
20th March 2009, 08:42 AM
am sure that you will find many references if you google the subject,
the best source for defining the procedure and acceptance criteria for pH meter is its user manual [manufacturer's recommendations]
Generally, we use point & slope method; outline of procedure is as follows:
1.use any-two-standard-solution and adjust the instrument response to match standard values.
2. measure the response for third standard, and evaluate the closeness of the value to the standard-reading.
3. calculate the slope of the measured response; and compare it against the manufacturer's recommended acceptance-criteria.
since you are suggesting that its an online probe, you will also need to evaluate the frequency of calibration & conditions as well.
and if you are in the purview of any regulatory agencies, then better refer to relevant guidelines/pharmacopoeia as well. [ which may have specific criteria for acceptance-criteria, standard solutions etc.,]
hope this helps.
thanks
valiveti.