I have a customer asking if there is any documentation to state that the CMM must be calibrated every year or is 18 month intervals okay. Does anyone have anything in writing from 17025 that says this or is it just an industry accepted standard.
Thanks
Dan
AndyN
23rd October 2009, 12:23 PM
I have a customer asking if there is any documentation to state that the CMM must be calibrated every year or is 18 month intervals okay. Does anyone have anything in writing from 17025 that says this or is it just an industry accepted standard.
Thanks
Dan
Dan:
There's nothing in ISO/IEC 17025 about such a thing. The selection of a calibration frequency is dependent on a number of issues, one being the stability of the equipment. If a cmm is located in a 'harsh' environment it may not even last 12 months before it needs a check! What's needed is data to help make the decision. Even the 'as new' calibration can give you a clue.
The best advice I've found is to start a schedule of intermediate checks, using an artifact - the ANSI ball bar test is one way - to see what kinds of readings you get. The manufacturer may suggest a yearly calibration, but without the data even that may be too frequent and the cmm may go much longer without a full blown calibration.
John Broomfield
23rd October 2009, 06:00 PM
I have a customer asking if there is any documentation to state that the CMM must be calibrated every year or is 18 month intervals okay. Does anyone have anything in writing from 17025 that says this or is it just an industry accepted standard.
Thanks
Dan
Dan,
I agree with Andy. Your calibration process for each device should show how often you need to calibrate. Do not stick to manufacturer's recommended frequency because they do not know the circumstances of use. For example, is the device used once a month or once an hour? Another example, is the device used in a dusty environment or clean environment?
You may find that you are wasting money on too frequent calibrations or too infrequent verifications.
John
True Position
23rd October 2009, 08:46 PM
Personally, I do not consider adjusting calibration intervals until I have a few years of records showing that it was in spec when it was calibrated. Only then I start considering moving the time back. I have yet to have an auditor have an issue with that.
If you want to check the machine between calibration cycles, I use gage blocks to check all 3 axis and a couple ring gages. Southern Gage also makes an artifact for CMM checking that some people have recommended:
http://www.gsgage.com/content/our_products/surveillance_masters.asp
cbearden
24th October 2009, 07:47 PM
There is nothing wrong with having an 18 month Calibration Interval....as long as the CMM is giving you accurate results and you can correltate those results using another method, it's your choice on the Calibration Interval.