View Full Version : Calibration Interval (Frequency) Adjustment using Hilliard type 2 algorithm


A C Diggs
29th September 2004, 01:58 AM
We have a new manager for our pharmaceutical calibration lab, and he has asked me for help deciphering our SOP on cal interval adjustments. Currently, the SOP states that we use the Hilliard type 2 calibration interval assignment algorithm. I have run some senerios through the formula and cannot come up with a situation that would qualify to extend the interval (not even zero OOTs worked). Perhaps, I'm not applying the formula correctly. :o
I work in a fair size shop with 3000 instruments to cal (approx 400 per month). We do not have a validated program to crunch these numbers for interval adjustment according to the Hilliard method. :frust:
We have kicked around the idea of rewriting the SOP and using a simple look-up table based on the previous 4 calibrations, but then we ran into the question of how do we test this table for its ability to meet our relaibility target? :confused:

Marc
19th December 2004, 12:31 PM
Can any stats experts help out on this one?

I know it's dated, but I'm trying to catch up on some zero reply threads.

Hershal
20th December 2004, 10:36 AM
I confess to ignorance of the specific algorithm (until I read up on it). Graeme might be the best source for a good answer here.

Hershal

Marc
21st December 2004, 05:14 AM
I tried a search for "Hilliard type 2 algorithm" (with and without the quotes) on just about every search engine and haven't found anything. I did find references to Cahn-Hilliard type equations but they are beyond me.

rmccallum
21st December 2004, 10:21 AM
If you are willing to spend a few $$ there is software out there that will optimize your cal intervals for you. I would have to assume it is statistically significant. Correct me if I am wrong.

Marc
21st December 2004, 01:04 PM
I've never seen any company go so far as to use something like a Hilliard type 2 algorithm (whatever that is). It's typically factors such as use frequency and environment, and results of previous calibrations.

Let's see - If we're going to analyze data to set calibration intervals, I'd look at 'as received' vs. 'as returned' if adjusted.

Anyone do data analysis and have a formula, or do most of you folks just look at the calibration results and 'gut feel' it?

This thread may be of interest: http://Elsmar.com/Forums/showthread.php?t=997

rmccallum
21st December 2004, 02:10 PM
Currently we don't statistically analize to establish calibration frequency. We use the "gut feeling" method. Curious if this is acceptable for a ISO 17025 accredited automotive lab. The standard does call out uncertainty analysis. I just assumed this would be applied to calibration optimization. I did find the calculations on a google search to optimize calibration using fitted modelling and reliability but I don't think I have the time or willpower to program an excel spread sheet so I found a company that supplies the software to perform this function.

I am just starting out in pursuit of the accreditation. I have a pretty good foundation but a long ways to go. You will most likely see me hanging around over the next while. One question that may sound stupid but can I reference ISO 17025 requirements such as internal auditing or corrective action to our ISO 9000: 2001 system or do I need to rewrite all these?

Hershal
21st December 2004, 04:45 PM
If we get away from the algorithm, then I recommend bouncing over to NCSLI RP (Recommended Practices) 1.

http://www.ncsli.org

It is well respected, established, and sound.

Some of the calibration software in the newer versions include cycle adjustment. Check with Cybermetrics and Blue Mountain.

Hershal

Marc
21st December 2004, 04:57 PM
:topic: I was all over the site but I kept seeing 'NCLSI'. Nowhere did I see what the acronym stands for (admittedly I''m 3/4 blind).

What is NCLSI? National Calibration Laboratories Standards Institute or something such?

I will admit I am really put off by going to a site where the acronym is everywhere but nowhere is it defined.

Hershal
22nd December 2004, 10:24 AM
Oops, I should have included that. My apologies.

National Conference of Standards Laboratories - International. It used to be NCSL, then in Toronto in 2000 the International was added.

:topic: I have some great party stories from that conference.........

The RPs and other documents including ANSI/NCSL Z540-1-1994, ANSI/NCSL Z540-2-1997 (the U.S. GUM), and ANSI/ISO 17025:2000 are available through the publications page.

Hope this helps.

Hershal

TownDawg
3rd January 2005, 11:07 AM
http://www.isgmax.com/Freeware/A3%20Method.pdf

Interesting... never really realized how scientifically one can develop calibration intervals. I also have always used the "gut method."

:)

Also... this freeware might be of interest.. http://www.isgmax.com/Freeware/A3Test.exe

Graeme
7th January 2005, 03:19 PM
I confess to ignorance of the specific algorithm (until I read up on it). Graeme might be the best source for a good answer here.

Hershal

:( I've been away from the Forums for a while ... too busy on other things. But now I'm back. :magic:

Hilliard agorithm -- that's a new one to me, but then I've only been in this business for 23+ years. When free to do so, I've always used methods from NCSL RP-1 or the tools from Dr. Howard Castrup. (The first half of my experience was working for the Navy, where they have a thick book that tells you what calibration interval to use!)

Hershal -- where did you find information on it?

(Now if it was a halliard -- I know what that is! A piece of running rigging on a sailboat.)