The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality

Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 6th August 2007, 05:49 AM
sarasheed sarasheed is offline
Involved in Discussions

 
Registration Date: May 2005
Location: India, State:Andhra pradesh
 
Posts: 46
Thanks Given to Others: 3
Thanked 2 Times in 2 Posts
Karma Power: 37
Karma: 24
sarasheed has less than 100 Karma points so far.
Question Validity of 7 Year Old Calibration Certificate

Our Laboratory is certified to ISO 17025:2005.We have procured a Master slip Gauge Box from Mutituyo on 15-10-2006 with calibration certificate traceable to Japan National Standard and put use as it was in good condition suitable for useing as a Master Standard.Date of certificte is 21-10-1999 and there is no indication about the validity of the certificate.The Auditor has raised a Major Non-confirmance since it has been calibrated 7 years back which is very long period.We have set a periodicity of calibration as 2 years in our system.

Is the Auditor's decision in raising the NC is correct?.We presumed that we can use it upto 15-10-2008 which is 2 years from the starting date of use .

sarasheed
  #2  
Old 6th August 2007, 08:35 AM
Jennifer Kirley's Avatar
Jennifer Kirley Jennifer Kirley is offline
Forum Moderator

 
Registration Date: Jan 2004
Location: Maine, USA
 
Posts: 5,225
Thanks Given to Others: 3,017
Thanked 2,843 Times in 1,634 Posts
Karma Power: 615
Karma: 27213
Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.
Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.
Re: Validity of Test certificte

Even though 7 years is a long period, I am confused about why you were iussued a major NC. Can you tell us exactly what the NC says?
__________________
"If you only have a hammer, you tend to see every problem as a nail." Abraham Maslow
Sponsored Links

  #3  
Old 6th August 2007, 08:36 AM
Jerry Eldred's Avatar
Jerry Eldred Jerry Eldred is offline
Forum Moderator

 
Registration Date: Dec 1999
Location: Texas
 
Posts: 512
Thanks Given to Others: 7
Thanked 228 Times in 139 Posts
Karma Power: 114
Karma: 4257
Jerry Eldred is appreciated, and has over 1700 Karma points.
Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.
Re: Validity of Test certificte

Since the certificate is seven years old and you have a two year interval, you would have to justify that it is valid to begin your calibration interval on the first date you used it. And you would also have to justify what has happened to the unit for the years since 1999.

Since it was calibrated in 1999, you have to prove that nothing has changed in such a long time (since it is multiples of your entire calibration interval).

I unfortunately would have to agree that if it were my standard, I would definitely send it out and have it re-certified rather than take the risk of the amount of time passed since it was calibrated something had possibly changed.

The only exception would be if you just bought it new from the manufacturer still in it's original box, and they gave you written justification that it is still valid to use after all those years since the original calibration date.

Sorry to be the bearer of a difficult position.
Thank You to Jerry Eldred for your informative Post and/or Attachment!
  #4  
Old 6th August 2007, 10:10 AM
BradM's Avatar
BradM BradM is offline
Forum Administrator

 
Registration Date: Sep 2006
Location: Arlington,Texas
 
Posts: 5,367
Thanks Given to Others: 1,309
Thanked 1,835 Times in 1,238 Posts
Karma Power: 400
Karma: 16759
BradM is appreciated, and has over 1700 Karma points.
BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.
Re: Validity of Test certificte

Quote:
In Reply to Parent Post by Jerry Eldred View Post

Since it was calibrated in 1999, you have to prove that nothing has changed in such a long time (since it is multiples of your entire calibration interval).

I unfortunately would have to agree that if it were my standard, I would definitely send it out and have it re-certified rather than take the risk of the amount of time passed since it was calibrated something had possibly changed.

The only exception would be if you just bought it new from the manufacturer still in it's original box, and they gave you written justification that it is still valid to use after all those years since the original calibration date.

Sorry to be the bearer of a difficult position.
I agree entirely with Jerry's reply. You have no idea where/how this standard has been stored, the use of it before you got it, etc. Was this brand new? How had it been stored? Etc.

Did you perform any type of cross-verification with other standards in your lab prior to use? Did you perform even a simple 1 to 1 check of some sorts?

You need to have this thing re-certified. The cost of recertification can be insignificant to the cost of making poor measurements with a potentially O.O.T. standard. Agreed?

Now, just my opinion, but I don't agree with the major NC on this one. But I do think it should be sent in.
__________________
Now, 75 years later in an abundant society where people have laptops, cell phones, iPods and minds like empty rooms, I still plod along with books. ”
—Harper Lee
  #5  
Old 6th August 2007, 10:33 AM
lawriematkin lawriematkin is offline
Getting Involved (6 to 9 Posts)

 
Registration Date: Sep 2006
Location: Switzerland
 
Posts: 8
Thanks Given to Others: 1
Thanked 2 Times in 1 Post
Karma Power: 28
Karma: 20
lawriematkin has less than 100 Karma points so far.
Re: Validity of Test certificate

As you procured in 2006. Proof that the measurement equipment still conformed to requirements is required. As this is used as a master calibrator the auditor may have taken the N/C as systemic as all other gauges calibrated to the master on the evidence be out of calibration to the master.
Purchase, without a currect calibration certificate (2006) means you have no evidence that supports conformity
Thank You to lawriematkin for your informative Post and/or Attachment!
  #6  
Old 6th August 2007, 10:44 AM
BradM's Avatar
BradM BradM is offline
Forum Administrator

 
Registration Date: Sep 2006
Location: Arlington,Texas
 
Posts: 5,367
Thanks Given to Others: 1,309
Thanked 1,835 Times in 1,238 Posts
Karma Power: 400
Karma: 16759
BradM is appreciated, and has over 1700 Karma points.
BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.
Re: Validity of Test certificate

Quote:
In Reply to Parent Post by lawriematkin View Post

As you procured in 2006. Proof that the measurement equipment still conformed to requirements is required. As this is used as a master calibrator the auditor may have taken the N/C as systemic as all other gauges calibrated to the master on the evidence be out of calibration to the master.
Purchase, without a currect calibration certificate (2006) means you have no evidence that supports conformity
Hello! Welcome to the Cove!! Thanks for the great first post!

You make a valid point concerning the NC. I would go with a major if there was no certificate or evidence of calibration whatsoever. To me (and I am not a 17025 expert) this would be a minor, for citing not having a current, valid certification. That is an opinion, and I would defer to those much more proficient in 17025 than I.

I hope to see you contribute and add to our knowledge base here!
__________________
Now, 75 years later in an abundant society where people have laptops, cell phones, iPods and minds like empty rooms, I still plod along with books. ”
—Harper Lee
  #7  
Old 6th August 2007, 12:25 PM
Ken K's Avatar
Ken K Ken K is offline
E-Mails Invalid or Rejected by Recipient System

 
Registration Date: Jun 2000
Location: USA
 
Posts: 406
Thanks Given to Others: 0
Thanked 17 Times in 14 Posts
Karma Power: 0
Karma: 530
Ken K is appreciated, and has over 500 Karma points.Ken K is appreciated, and has over 500 Karma points.Ken K is appreciated, and has over 500 Karma points.Ken K is appreciated, and has over 500 Karma points.Ken K is appreciated, and has over 500 Karma points.Ken K is appreciated, and has over 500 Karma points.
Re: Validity of Test certificate

I'm rather surprised Mutituyo sold you a standard with a 7 year old certificate.

I'm not surprised you received a major for using it. You indicate you have used it for a Master Standard, yet you have not indicated you have performed any type of verification on it. Chances are it is within tolerance for your use, but you have to proof except for an out of date calibration certificate.

Personally, I would send it out for calibration and wait for the results. If it is out of tolerance, you have nearly a year of data which will be questionable that I suspect the auditor will take a good look at.

Good luck.
  #8  
Old 6th August 2007, 03:48 PM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

 
Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 57
 
Posts: 2,114
Thanks Given to Others: 1
Thanked 637 Times in 452 Posts
Karma Power: 266
Karma: 8659
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Re: Validity of 7 Year Old Calibration Certificate

A corrective action is valid in this case, because the blocks had not been calibrated in 7 years, the original cert showed no evidence of validity, and your own system says two years. I would also have written a non-conformance.

The reality is, you cannot state that you know the value +/- the uncertainty value, of any block in the set. Temperature changes, skin oils, and many other factors will change the values.

As for the major/minor, each AB makes its own decision as to that grade system.....some use it, some do not.....I can't offer any real opinion on that.

Send it to an accredited lab and have it cal'd.....then staying on cycle is much easier.

Hope this helps.
Thanks to Hershal for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Validity Duration of USP / FDA Certificate? cisko Other ISO and International Standards and European Regulations 5 30th September 2011 03:57 AM
AS9100 Rev. B Certificate Validity Joy AS9100, Nadcap and related Aerospace Standards and Requirements 6 23rd March 2010 11:40 PM
IATF portal for TS 16949 Certificate Validity Check DilekTKY ISO/TS 16949 - International Automotive Quality Systems Standard 2 19th January 2010 01:58 AM
ISO 9001:2000 Certificate - How to have a 3 year validity instead of 1? cmfugen ISO 9000, ISO 9001, and ISO 9004 - Questions and Discussions 15 8th September 2008 05:19 AM
Ever come across an ISO 9001 certificate with more than 3 years validity period? cheahga ISO 9000, ISO 9001, and ISO 9004 - Questions and Discussions 35 19th September 2007 04:15 AM



The time now is 09:28 AM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


   


Marc Timothy Smith - Elsmar.com
8466 LeSourdsville-West Chester Road, Olde West Chester, Ohio 45069-1929
513 341-6272