The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


Related Topic Tags
micrometers, working standards (metrology), digital measurement equipment, calibration (general topics)
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 15th July 2011, 07:33 AM
Dongzkie Dongzkie is offline
Involved in Discussions

 
Registration Date: Dec 2010
Location: Cebu Philippines
 
Posts: 39
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 13
Karma: 10
Dongzkie has less than 100 Karma points so far.
Question What is the correct Working Standard for Digital Micrometer?

What is the correct working standard for digital micrometer?
I am confuse as far as i have known gage block is the right working standard for micrometer but we are using pin gage. I've asked one of my colleague he told me that we are measuring diameter of our product. Is this really right?
  #2  
Old 15th July 2011, 08:07 AM
Jim Wynne's Avatar
Jim Wynne Jim Wynne is offline
Cross Forum Moderator

 
Registration Date: Jan 2005
Location: Southeast Wisconsin
 
Posts: 13,332
Thanks Given to Others: 1,779
Thanked 5,354 Times in 3,431 Posts
Blog Entries: 2
Karma Power: 1445
Karma: 39214
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Re: micrometer working standard

Quote:
In Reply to Parent Post by Dongzkie View Post

What is the correct working standard for digital micrometer?
I am confuse as far as i have known gage block is the right working standard for micrometer but we are using pin gage. I've asked one of my colleague he told me that we are measuring diameter of our product. Is this really right?
It depends on what you mean by "working standard." In general if you're calibrating a standard micrometer, blocks are preferable because they engage the entire contact surface. If you're using a some form of standard on the shop floor to verify a mic prior to use, and the pins simulate typical workpieces, then pins are probably the way to go.
__________________
We aim to please. You aim too, please.
Thanks to Jim Wynne for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 15th July 2011, 03:40 PM
step30044 step30044 is offline
Involved in Discussions

 
Registration Date: Nov 2007
 
Posts: 26
Thanks Given to Others: 0
Thanked 3 Times in 3 Posts
Karma Power: 25
Karma: 25
step30044 has less than 100 Karma points so far.
Re: What is the correct Working Standard for Digital Micrometer?

You will need to account for the accuracy of the pins and micrometer as well as the condition of the pins
  #4  
Old 15th July 2011, 03:42 PM
Jim Wynne's Avatar
Jim Wynne Jim Wynne is offline
Cross Forum Moderator

 
Registration Date: Jan 2005
Location: Southeast Wisconsin
 
Posts: 13,332
Thanks Given to Others: 1,779
Thanked 5,354 Times in 3,431 Posts
Blog Entries: 2
Karma Power: 1445
Karma: 39214
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Re: What is the correct Working Standard for Digital Micrometer?

Quote:
In Reply to Parent Post by step30044 View Post

You will need to account for the accuracy of the pins and micrometer as well as the condition of the pins
That's true, but it would also be true of blocks.
__________________
We aim to please. You aim too, please.
  #5  
Old 15th July 2011, 09:19 PM
Dongzkie Dongzkie is offline
Involved in Discussions

 
Registration Date: Dec 2010
Location: Cebu Philippines
 
Posts: 39
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 13
Karma: 10
Dongzkie has less than 100 Karma points so far.
Re: What is the correct Working Standard for Digital Micrometer?

thank you very much for the advice.
As for now we are using Class Z gage + which has the accuracy of 0.0025mm use in Mitutoyo digital micrometer has accuracy of 0.001mm and i don't think this is acceptable.
  #6  
Old 16th July 2011, 08:44 AM
Jim Wynne's Avatar
Jim Wynne Jim Wynne is offline
Cross Forum Moderator

 
Registration Date: Jan 2005
Location: Southeast Wisconsin
 
Posts: 13,332
Thanks Given to Others: 1,779
Thanked 5,354 Times in 3,431 Posts
Blog Entries: 2
Karma Power: 1445
Karma: 39214
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Re: What is the correct Working Standard for Digital Micrometer?

Quote:
In Reply to Parent Post by Dongzkie View Post

thank you very much for the advice.
As for now we are using Class Z gage + which has the accuracy of 0.0025mm use in Mitutoyo digital micrometer has accuracy of 0.001mm and i don't think this is acceptable.
What are the tolerances of things you're measuring?
__________________
We aim to please. You aim too, please.
Thanks to Jim Wynne for your informative Post and/or Attachment!
  #7  
Old 16th July 2011, 10:10 AM
JRKH's Avatar
JRKH JRKH is offline
Involved - Posts

 
Registration Date: Apr 2001
Location: Cincinnati, Ohio
Age: 58
 
Posts: 936
Thanks Given to Others: 484
Thanked 261 Times in 170 Posts
Karma Power: 150
Karma: 3667
JRKH is appreciated, and has over 1700 Karma points.
JRKH is appreciated, and has over 1700 Karma points.JRKH is appreciated, and has over 1700 Karma points.JRKH is appreciated, and has over 1700 Karma points.JRKH is appreciated, and has over 1700 Karma points.JRKH is appreciated, and has over 1700 Karma points.JRKH is appreciated, and has over 1700 Karma points.
Re: What is the correct Working Standard for Digital Micrometer?

I must say that, I am not particularly convinced that using pins is a good idea. The justification that one is measuring "round parts" doesn't convince me.

The issue of checking round parts may effect the wear pattern (flatness) of the anvil and to a lesser extent the spindle but if this is significant then it is something best evaluated in a cal lab setting and not on the shop floor. The wear on the Anvil will likely be in a single direction due to fact that it is fixed, and the particulars of measuring round parts. The wear on the spindle will be different and less pronounced depending on the number of different diameters are measured due to the fact that the spindle rotates and thus contacts the surface differently with different dia's.
The readout, digital or analog - should have no bearing on the standard used.

That is just the gut reaction of an old machinist who ran turret and engine lathes for many years and measured thousands and thousands of round parts, before even coming to the inspection/quality field.

Peace
James
__________________
Low tech is better than no tech.

The only constant is those who declare, "Things around here will never change!!"
  #8  
Old 16th July 2011, 10:34 AM
Jim Wynne's Avatar
Jim Wynne Jim Wynne is offline
Cross Forum Moderator

 
Registration Date: Jan 2005
Location: Southeast Wisconsin
 
Posts: 13,332
Thanks Given to Others: 1,779
Thanked 5,354 Times in 3,431 Posts
Blog Entries: 2
Karma Power: 1445
Karma: 39214
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Re: What is the correct Working Standard for Digital Micrometer?

Quote:
In Reply to Parent Post by JRKH View Post

I must say that, I am not particularly convinced that using pins is a good idea. The justification that one is measuring "round parts" doesn't convince me.

The issue of checking round parts may effect the wear pattern (flatness) of the anvil and to a lesser extent the spindle but if this is significant then it is something best evaluated in a cal lab setting and not on the shop floor. The wear on the Anvil will likely be in a single direction due to fact that it is fixed, and the particulars of measuring round parts. The wear on the spindle will be different and less pronounced depending on the number of different diameters are measured due to the fact that the spindle rotates and thus contacts the surface differently with different dia's.
The readout, digital or analog - should have no bearing on the standard used.

That is just the gut reaction of an old machinist who ran turret and engine lathes for many years and measured thousands and thousands of round parts, before even coming to the inspection/quality field.

Peace
James
I'm going under the assumption that the OP is using calibrated micrometers and calibrated pins (or other "working" standards), and using the standards on the floor as just a quick verification that things are as they should be. When doing a full calibration, blocks of the appropriate class should be used. In the circumstances described by the OP, where diameters are being measured, I don't see a problem with using pins of the appropriate size to do the verification.
__________________
We aim to please. You aim too, please.
Thanks to Jim Wynne for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Results of a GR&R study done on a 0-1" Digital Micrometer dbulak Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5 7th April 2010 01:27 PM
In-House Calibration - Digital Micrometer found Out of Calibration tknchancs ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 10 3rd November 2009 01:09 AM
GRR (Gage R&R) procedure for Mitutoyo Digital Micrometer wanted dbulak Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1 3rd June 2009 02:14 AM
Need of working standard for vernier and micrometer calipers rizy951 ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 2 28th May 2009 01:42 AM
What is the useful life for a good 0-1" digital micrometer? ScottK Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7 25th April 2007 10:45 AM



The time now is 04:14 AM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


   


Marc Timothy Smith - Elsmar.com
8466 LeSourdsville-West Chester Road, Olde West Chester, Ohio 45069-1929
513 341-6272