Sean Kelley said:
When we conduct Measurement Sustem Analysis the important step is to conduct % Gauge R&R study.After conducting it what are the acceptable limits? Is there any min limit for it? Can we have some thing like 0.08% variation due to % G R&R and say 99.92% variation due to parts we have used.
That's great! That's what you want to see. Actual Part variation should be the major contributor to total observed variation. Contribution of GRR error should be minimum. I trust you have taken samples of the 'same' part. If this is not the case, then observed variation is unduly inflated as compared to GR&R.
Usually, R&R error of 10% of Study variation is acceptable. The lower the better. If R&R is between 10 - 30%, the measurement system needs improvement and can be 'conditionally' accepted. Over 30% error is not acceptable.
You also have the choice of comparing the GRR with process variation and tolerance, in addition to the Study variation (TV).
I suggest you get hold of the AIAG MSA Manual (if you haven't already)..