class ii medical device, embedded software, eu validation, fda (food and drug administration), installation operational and performance qualification, installation qualification (iq), mdd (medical device directive), operational qualification (oq), performance qualification (pq), requirements and specifications, iec 62304 - medical device software life cycle processes  |
 |
|

28th March 2008, 11:48 AM
|
|
Getting Involved (6 to 9 Posts)
Registration Date: Jan 2008
Location: Switzerland
|
|
Posts: 8
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 22 Karma: 10 
|
|
EU Validation Requirement vs. FDA Validation Requirement - EN 62304
Dear all,
Our software development SOP (for a class 2 medical device with an embedded software) is based on EN 62304. I am now trying to submit a 510(k) and realised that the terminology for software testing requirement has changed from Unit Testing and verification/System testing to IQ/OQ/PQ (Installation/Operational/Performance Qualification).
Do we have to change our terminology documentation format to fulfill FDA's requirement?
The same goes to the final device functionality test protocol. Whilst this is fine for MDD do I need to create an IQ/OQ/PQ protocol for the US market?
Your advice will be greatly appreciated
H
Last edited by Marc; 28th March 2008 at 12:01 PM.
Reason: Added (Installation/Operational/Performance Qualification) for search engines.
|

29th March 2008, 10:55 AM
|
 |
Forum Administrator
Registration Date: Dec 2005
Location: Massachusetts, USA
Age: 49
|
|
Posts: 5,123
Thanks Given to Others: 1,471
Thanked 2,214 Times in 1,427 Posts
Karma Power: 400
|
|
|
Re: EU Validation Requirement vs. FDA Validation Requirement - EN 62304
Did you have a look at IEC 62304:2006 ? This is recognized by FDA.
__________________
Well done is better than well said - Benjamin Franklin
|
|
Thanks to Ajit Basrur for your informative Post and/or Attachment!
|
|

29th March 2008, 09:10 PM
|
|
Super Moderator
Registration Date: Sep 2005
Location: Johore/Malaysia
|
|
Posts: 5,959
Thanks Given to Others: 1,409
Thanked 2,752 Times in 1,912 Posts
Karma Power: 662
|
|
|
Re: EU Validation Requirement vs. FDA Validation Requirement - EN 62304
Quote:
In Reply to Parent Post by Hrobot
Dear all,
Our software development SOP (for a class 2 medical device with an embedded software) is based on EN 62304. I am now trying to submit a 510(k) and realised that the terminology for software testing requirement has changed from Unit Testing and verification/System testing to IQ/OQ/PQ (Installation/Operational/Performance Qualification). ........
|
Somebody correct me if I am wrong. Unit testing applies to software - unit testing is a procedure used to validate that individual units of source code are working properly whereas IQ/OQ/PQ refers to the need to validate the complete equipment or machine at the relevant stages suggested by the term itself.
|
|
Thanks to harry for your informative Post and/or Attachment!
|
|

31st March 2008, 05:17 AM
|
|
Getting Involved (6 to 9 Posts)
Registration Date: Jan 2008
Location: Switzerland
|
|
Posts: 8
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 22 Karma: 10 
|
|
|
Re: EU Validation Requirement vs. FDA Validation Requirement - EN 62304
I think I made a mistake there to reference unit testing where I should have just referred to the final system testing.
I have checked the standards which are recognised by FDA and it includes 62304 Software life cycle, 60601 medical electronic equipment and 14971 risk management which are all the standards that we adhere to.
DOes this mean that I could use the Abrreviated 510(k) with a declaration of conformity to these standards?
regards,
H
|

31st March 2008, 06:23 PM
|
 |
Robots R Us
Registration Date: Dec 2005
Location: Santa Barbara, California, USA
|
|
Posts: 246
Thanks Given to Others: 58
Thanked 58 Times in 26 Posts
Karma Power: 56
|
|
|
Re: EU Validation Requirement vs. FDA Validation Requirement - EN 62304
I just went back through and read both the FDA 510(k) new Paradigm and the requirements for using an Abbreviated 510(k). Sounds like your case is a good example of why they created this abbreviated path! Good Luck!
|

31st March 2008, 06:32 PM
|
 |
Robots R Us
Registration Date: Dec 2005
Location: Santa Barbara, California, USA
|
|
Posts: 246
Thanks Given to Others: 58
Thanked 58 Times in 26 Posts
Karma Power: 56
|
|
|
Re: EU Validation Requirement vs. FDA Validation Requirement - EN 62304
I am also very interested in the interpretation of the IQ/OQ/PQ comment in HROBOTs first posting. Is the implication that if these particular terms are not mentioned in IEC 62304 that they are not required by FDA? I'm far from a software expert, hence my naivete/curiosity.
|

5th April 2008, 02:43 PM
|
|
Appreciated Information Resource
Registration Date: Jun 2005
Location: Plano, TX
Age: 53
|
|
Posts: 560
Thanks Given to Others: 369
Thanked 334 Times in 240 Posts
Karma Power: 97
|
|
|
Re: EU Validation Requirement vs. FDA Validation Requirement - EN 62304
Quote:
In Reply to Parent Post by Watchwait
I am also very interested in the interpretation of the IQ/OQ/PQ comment in HROBOTs first posting. Is the implication that if these particular terms are not mentioned in IEC 62304 that they are not required by FDA? I'm far from a software expert, hence my naivete/curiosity.
|
There is nothing in 21 CFR 820 (regulation by which FDA requires compliance for device development and manufacturing) that requires a particular test paradigm for product verification and validation.
In my experience, IQ/OQ/PQ are typically (only) used in manufacturing equipment validation, not product V&V.
|
|
Thanks to yodon for your informative Post and/or Attachment!
|
|

6th April 2008, 09:09 AM
|
|
Involved - Posts
Registration Date: Nov 2003
Location: Israel
Age: 64
|
|
Posts: 61
Thanks Given to Others: 0
Thanked 20 Times in 10 Posts
Karma Power: 45 Karma: 90 
|
|
|
Re: EU Validation Requirement vs. FDA Validation Requirement - EN 62304
I recently submitted a 510(k) for a medical device with embedded SW. I based my SW documentation and terminology on the outlines described in the FDA Guidance document entitled:" Guidance for the Content of Premarket Submissions for Software Contained in Medical Devices, dated May 11, 2005".
From my experience with the FDA, when they publish a Guidance Document that is applicable to your situation, then you should carry out the recommendations contained in the Guidance Document as much as possible/applicable.
Note that if you submit Declarations of Conformity in lieu of actual test reports, then they should comply with the FDA requirements listed in the attached table.
Also, the FDA has several Guidance Documents on the Use of Standards in Substantial Equivalence Determinations.
Regards,
Asherlee
|
|
Thank You to Asherlee for your informative Post and/or Attachment!
|
|
Lower Navigation Bar
|
|
|
Do you find this discussion thread helpful and informational?
|
Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
|
|
|
| Thread Tools |
Search this Thread |
|
|
|
| Display Modes |
Rate Thread Content |
Linear Mode
|
|
Forum Posting Settings
|
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
HTML code is Off
|
|
|
|
|