In our case, we are in the US and have stopped selling our monitor in the EU because we are not 60601-1 3rd edition compliant, but would like to continue to support customers there by repairing their units, replacing accessories, etc.
We are considering not renewing with our Authorized Rep. We cannot supply the info they request anyway since the unit will not pass the updated standard, and we are maybe two years away from introducing anything in the EU.
Without an AR can we still support our EU customers? It seems we would have a problem shipping a repaired unit back into the EU with the AR's name on the back but without them acting as our AR.
Hi,
The situation you describe is complex. "Continue to support customers" is not a clear-enough terminology to make decisions by. Repair of units already on the market and "replacing accessories" (= supply of new accessory units, if I got it right) are quite different from a regulatory perspective. Requirements applying to "etc." depend on the details of that "etc.".
As MIREGMGR noted, if you don't have a registered place of business in the EC (in simple words, an EC branch or office) and you don't have an EC rep, you are not eligible for CE marking (MDD article 14, s. 2). So if you go forward and discontinue the representation, the question should be "what are you still allowed to do when you're not eligible for the CE mark any more?"
Supply of new accessory units would (in my understanding) be regulated just as the supply of new device units. Therefore, not being eligible to CE mark those new accessory units, you won't be allowed to supply them any more.
With repair of units already on the market, the situation is different. The MDD's article 17, s. 1 states that "Devices... must bear the CE marking of conformity
when they are placed on the market." (emphasis added).
As detailed in the European Commission's interpretative document on
PLACING ON THE MARKET OF MEDICAL DEVICES (
https://ec.europa.eu/health/medical-devices/files/guide-stds-directives/placing_on_the_market_en.pdf), various Directives and Regulation define that "placing on the market" means "
first making available..." and so on (emphasis added); so, placing a device
back on the market (e.g. after a repair) is excluded from the scope. This is further clarified by The European Commission's
Guide to the implementation of directives based on the New Approach and the Global Approach, on p. 16:
Products which have been repaired (for example following a defect), without changing the original performance, purpose or type, are not to be considered as new products according to New Approach directives. Thus, such products need not undergo conformity assessment... This applies even if the product has been temporarily exported to a third county for the repair operations... Thus, maintenance operations are basically excluded.
and on p. 17:
Consumer products outside the field of application of New Approach directives and other Community legislation (for example... repaired products) come under the Directive on general product safety, where they are supplied in the course of commercial activity.
With regards to shipping repaired units back to the EC market, bearing outdated EC rep details, I think that even if there's no regulatory issue, you should sort it with that EC rep first. If things get complicated, maybe consider modifying those details on the units coming in for repair, to state something like "Initially placed on the EC market under representation of..."; however, a written opinion from your NB would probably be the safest measure.
Cheers,