jayquality
hello covers...
working on a med dev class i with measuring function certification
the mdd states:
the intervention by the notified body is limited to:
in the case of devices with a measuring function, only the aspects of
manufacture concerned with the conformity of the products with the
metrological requirements
lrqa has a nice document stating:
The aspects relating to the metrology of Class I Measuring Devices that LRQA would assess include, but are not limited to:
• those elements of the risk analysis and essential requirements checklist relating to measuring,
• manufacturers records and procedures relating to measuring aspects and calibration,
• the traceability of calibration,
• units of measure,
• post market surveillance and vigilance systems,
• draft labels, packing and instructions for use bearing the CE marking,
• procedure to prepare and review the Declaration of Conformity,
• procedure to identify what is a significant change and what should be advised to the notified body.
My questions:
What experience do you have with the 'limited' scope of the assesment?
Note the 'not limited to..' clause
Anyone got non conformities on issues not related to measurement functions?
Does the NB look at the whole technical file in detail or just diagonal and picking out the metrological items?
Do they actually look at say 60601-1 or 60601-1-2 (EMC) or 10993-1 reports, the 2 latter being not relative to a measuring function?
I guess this also applies to the QMS as required by annex V but no certificate to 13485 would be possible with the limited scope, right?
iow to get a 13485 certificate you need to implement an entire qms (providing no exclusions in the scope)?
On a slightly different topic, assume you have the technical capabilities to perform 60601-1 testing (leak current, test finger etc) or even equipment to do EMC measurements (i have a test chamber at my disposal but not accredited for medical stuff), can you generate your own report and use that?
Or does the NB only accept these compliance reports from accredited labs?
I can't find in the 60601-1-2 that you cannot do your own emc measurement...
thanks
Jay
the mdd states:
• units of measure,
My questions:
thanks
Jay