jmalm
I am having difficulty trying to meet part of the 60601-2-27 (2nd edition) subclause 50.102.5 Multichannel crosstalk.
The subclause states:
"When an input signal limited in amplitude and rate as per 50.102.2 is applied to any one LEAD WIRE of the multi-channel EQUIPMENT, with all other LEAD WIRES connected to the N (RL) LEAD WIRE, the unwanted output in the unused LEAD WIRES shall not be greater than 5 % of the applied input signal.
Close switches S, S2 and S4. Set the EQUIPMENT SENSITIVITY to 10 mm/mV. Connect LEAD WIRES F(LL), V1, and if provided, the Frank E to the R (RA) LEAD. All other LEAD WIRES are connected to the N (RL) LEAD WIRE. [...]. From the signal generator, apply 2,5 mV peak-to-valley 30 Hz triangular wave between the F (LL), V1, and if provided, the Frank E and the N (RL) LEAD WIRE. Set the EQUIPMENT SENSITIVITY to 10 mm/mV using any display device. The output in the other channels shall be ≤ 5 % peak-to-valley of the full scale output."
The problem I'm having is that some of my inputs V1-V6 are referenced to a Wilson Centre Terminal voltage derived from an average of RA,LL,and LA signals. If I apply a voltage as stipulated in this clause to a subset of RA,LL, or LA, I will have a significant voltage that shows up on leads V1-6, which is the way it's supposed to work in the field.
However the way the clause is written, I will never be able to pass this test. Am I misinterpreting this clause or is the way my circuit is constructed improper to satisfy this clause? Is this going to be changed in the upcoming 3rd-edition release?
Thanks!
Jesse
Thanks!
Jesse