Recently we were audited for our ANVISA registration. One of the topics was 100% testing and Certificate of Conformance for the crash cart we use. The cart is CE marked and tested to 60601-1 3rd edition, but 100% testing is not done on each cart. I would like to push this requirment back onto the cart manufacture but I am not sure which standard I should be applying. Is it the clauses from 60601-1:2005?
This is the NC which was written up:
8.1 and 10.1-10.5 Lucid products are 5 V/12 Vdc rated USM powered from the mains through a cart supplied with an isolation transformer. The cart is supplied to Lucid by lTD Model KD6402.9. Functional Testing is conducted on the system as per procedure document 02603 Rev E, attached. Safety production line tests for earth continuity, electric strength, and leakage current tests conducted on 100% production at lTD in Germany. Lucid to obtain Safety Test Procedure from lTD including the functional procedure for the test equipment. Lucid also must obtain a CoG from lTD declaring that the safety tests have been conducted on 100% production.
This is the NC which was written up:
8.1 and 10.1-10.5 Lucid products are 5 V/12 Vdc rated USM powered from the mains through a cart supplied with an isolation transformer. The cart is supplied to Lucid by lTD Model KD6402.9. Functional Testing is conducted on the system as per procedure document 02603 Rev E, attached. Safety production line tests for earth continuity, electric strength, and leakage current tests conducted on 100% production at lTD in Germany. Lucid to obtain Safety Test Procedure from lTD including the functional procedure for the test equipment. Lucid also must obtain a CoG from lTD declaring that the safety tests have been conducted on 100% production.