Calibration of Tape Measures used on our production floor - What are 'Best Practices'
We recently received a minor nonconformance during a surveillance audit. The finding concerned the calibration (or lack thereof) of tape measures used on our production floor. These tapes are used to confirm lengths after a cutting operation. I am looking for a logical, simple method for calibrating tape measures. Does anyone have any "best practices" to share? I appreciate the assistance.
Dom-
|

27th July 1999, 07:30 AM
|
|
|
I think that you may need to provide more information on the subject.
Are the tapes used to "determine acceptance of the product" for movement to the next function area or for shipment?
If not, label them as "reference tools" (assuming that you have some other method of measuring/testing the product for conformance to specification.
If they are, I believe that you may be in a pinch. Tape measures are subject to so much regular abuse, malformation, and other circumstances that would impact their accuracy (to some tolerances) --> you might find yourself hard pressed to keep them calibrated (or spend a ton of time rechecking them).
Apologies for only being semi-helpful. Throw some more details out there so that those more knowledgeable than I on the subject can give you some better advice.
ALM
|

27th July 1999, 09:44 PM
|
|
|
dscumaci: I must agree with ALM, more info is needed to give a helpful reply.
What type of verification is used at your facility to determine calibration of tape measures as specified under your calibration procedures?
------------------
GSH
Quality Mgr.
QS9000 Coordinator
|

10th August 1999, 08:07 AM
|
|
|
|
Gentlemen,
Thank you for your response and I apologize for the vagueness of the question. To clarify the matter, the tape measures are used to make accept/reject decision after steel bar has been cut. Tolerances for length of bar and plate range from 1/4" and 1/16". However, my question is what my be a logical means of calibrating the tape measures? Does anyone have experience implementing a process for calibrating tape measures? What are some options we may have?
Thank you for your help;
Dom
|

10th August 1999, 08:40 AM
|
|
An Original Cover!
Registration Date: Dec 1998
Location: Greenville, SC
|
|
Posts: 471
Thanks Given to Others: 0
Thanked 4 Times in 4 Posts
Karma Power: 54 Karma: 45 
|
|
|
You can't calibrate a tape measure (because you can't adjust it to bring it back into conformance), but you can verify aspects of it. You can verify that it is still readable and you can compare it to a standard (if you suspect warpage, for instance). One option is to verify it at incoming inspection and then establish reasonable intervals for verification. Also, come up with rules of action, such as what to do if the tape measure has been dropped or been exposed to unusual wear conditions.
|

10th August 1999, 12:01 PM
|
 |
One of THE Original Covers!
Registration Date: Nov 1998
Location: Wallingford, CT USA
Age: 43
|
|
Posts: 1,158
Thanks Given to Others: 22
Thanked 63 Times in 43 Posts
Karma Power: 94
|
|
|
Remember to keep in mind that you are required to calibrate equipment through their usable range. Purchase the right Tape measure for the job. To add to Roger's post, consider the things that break or become loose (i.e. the end-hook on the tape measure). Is there any play?
|

10th August 1999, 04:26 PM
|
|
|
|
Do you need to consider the "readibility" factor, where the measurement device must read one more decimal than the stated tolerance? 1/16 is .0625. Your device should be readible to .006? (1/125?) This may be splitting hairs, but could this be an audit target, and more importantly an issuue for you? I guess it depends upon the criticality of the measurement.
|

11th August 1999, 05:04 PM
|
|
Courtesy Access
Registration Date: Dec 1998
Location: South Central Massachusetts
|
|
Posts: 788
Thanks Given to Others: 0
Thanked 15 Times in 12 Posts
Karma Power: 63 Karma: 151  
|
|
|
keep a steel rule for verification when tape gets worn/abused ...discard (or tag) and replace with a new one when wear reaches an unacceptable level...the steel rule comes in calibrated/verified or what ever........and if not heavily used has an infinitye cal cycle....or tap on one of 15 years.....You set the rules, what makes sence for you
|
Lower Navigation Bar
|
|
|
|
Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
|
|
|
| Thread Tools |
Search this Thread |
|
|
|
| Display Modes |
Rate Thread Content |
Linear Mode
|
|
Posting Settings
|
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
HTML code is Off
|
|
|
|
|