From: "Doucet, Frank (INMS)" <Frank.Doucet@nrc.ca>
To: "'ISO 25 discussion group'" <iso25@quality.org>
Subject: ISO 10012 and ISO/IEC FDIS 17025 (still a draft but not for long - no more changes allowed)
Date: Thu, 8 Apr 1999 20:08:13 -0400
To: All concerned with the interactions of these two documents
These are some of my views that I would like to share and discuss with
anyone interested:
My views are that these two documents (ISO 10012 and ISO/IEC 17025) are supposed to address different situations and are both needed in the market. Because both of them deal with measuring equipment, it is reasonable to expect that they will overlap and cover similar subjects. ISO 10012 is a document supposed to be concerned with quality assurance requirements for measuring equipment at all levels of use from calibration labs to manufacturing activities, including incoming inspection, etc. On the other hand, 17025 is a document supposed to be concerned with the requirements for the competence of testing and calibration laboratories to assure that test and calibration results are technically valid. I will cite a some examples to emphasize these distinctions.
I am a buyer of a product and the contract that I have with the producer might include a stipulation that the measuring equipment used in all phases of manufacturing complies with ISO 10012. This requirement could be to assure that equipment are labeled properly (calibration status including cal date, due date, etc. ), and that adequate calibration intervals are established to give desired reliability levels, to name a few, with the main purpose of demonstrating the producers capability to provide reliable measurement results. In addition to this, the contract could stipulate ISO 9001 registration and accredited calibrations (17025) for all reference standards, etc. Someone might argue that 17025 will do the same thing in this case as 10012 and because of this, 10012 is redundant. Unfortunately requiring only 17025 will not assure that these requirements are met. For example, 17025 also requires labeling of equipment to indicate calibration status, but this requirement is only for equipment held by the lab and does not apply to equipment sent in for calibration by a client (although most labs will label client's equipment as a routine practice). In addition, 17025 stipulates that calibration reports and labels shall not contain any recommendation on the calibration interval, except where this has been agreed with the client.
These are only a few of the many possible examples and I'm sure that other readers will come up with more.
And finally, for those laboratories considering using 10012 instead of 17025, I would highly recommend comparing the two documents and I'm quite confident that after an objective comparison, 17025 will be the choice of the vast majority. This is because 10012 does not address a large percentage of the important aspects of a laboratory operation, including its competence.
-----------------
Frank J. Doucet, Technical Advisor, Electrical
National Research Council Canada
Institute for National Measurement Standards
Calibration Laboratory Assessment Service (CLAS)
M36-RM22
Ottawa, Ontario, Canada K1A 0R6
Tel.(613)993-0159 Fax.(613)952-1394
<mailto:frank.doucet@nrc.ca> frank.doucet@nrc.ca
www.cisti.nrc.ca/inms/