The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


Related Topic Tags
digital measurement equipment, calibration (general topics)
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 30th December 2004, 02:49 PM
JSmith JSmith is offline
Email Activation Not Completed

 
Registration Date: Mar 2002
Location: Clawson, MI
Age: 39
 
Posts: 20
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 0
Karma: 107
JSmith is appreciated, and has over 100 Karma points.JSmith is appreciated, and has over 100 Karma points.
Question Digital Caliper Calibration vs. Verification - Full Scale vs. Used Range

Hello,

I have a question regarding the calibration / verification of Digital Calipers.

If my company uses calipers to measure parts up to 1 1/2 " only, would it be acceptable for me to verify the 6" calipers from 1 - 2 " with my gage blocks over that range and not send them out for calibration to an outside company?

Thanks

Jen

Last edited by Wes Bucey; 30th December 2004 at 02:52 PM. Reason: fix typo in headline for better use by search spiders

Sponsored Links
  #2  
Old 30th December 2004, 03:06 PM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

 
Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 57
 
Posts: 2,100
Thanks Given to Others: 1
Thanked 625 Times in 444 Posts
Karma Power: 265
Karma: 8599
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Only thing to consider here is: What is the likelihood that the calipers will ever be used outside the 2" range?

If none (or almost none), then calibrating to only the needed range is advisable. Caliper calibration is cheap so ou may not save any mone, but ask.

Of course, I always recommend using an calibration lab accredited to SO/IEC 17025.

Hershal
Sponsored Links

  #3  
Old 30th December 2004, 03:18 PM
Kevin H's Avatar
Kevin H Kevin H is offline
Lab Manager

 
Registration Date: Jun 2004
Location: Western Pennsylvania, USA
Age: 60
 
Posts: 489
Thanks Given to Others: 19
Thanked 50 Times in 40 Posts
Karma Power: 87
Karma: 1203
Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.
Hi Jen, ,my experience is a little dated, as I left when the Mechanical Testing lab I ran was transitioning from ISO Guide 25 to ISO Guide 17025 about 4 years ago. That said, in our case the supervisor (me) was responsible for quarterly calibration of the digital electronic calipers we used. The calipers were used to measure elongation from tensile testing of flat rolled stainless steel. The primary measurement area was between 2" and 3", 2" being the initial measurement before beginning tensile testing and 3" being reached or exceeded only on rare occasions. I used gage blocks to check over the ranges we'd typically use, beginning with verifying the zero location, then checking at 1/2", 1", 2", 3" and 4". If they were off, I had some adjustments I could do to bring them into calibration, or could send them out for work & recalibration by an external source. On a daily basis, they were verified by each turn by checking the zero value and then verifying against a 2" gauge block. At that time the program satisfied our Guide 25 registrar - A2LA.
Of course, we also did round robin testing of tensile specimens that was reported to both us and our registrar. That testing included comparing all results from specimens prepared and tested by the lab including elongation.

Hope the information helps in making your decision.
  #4  
Old 3rd January 2005, 11:20 AM
JSmith JSmith is offline
Email Activation Not Completed

 
Registration Date: Mar 2002
Location: Clawson, MI
Age: 39
 
Posts: 20
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 0
Karma: 107
JSmith is appreciated, and has over 100 Karma points.JSmith is appreciated, and has over 100 Karma points.
Thank you for your replies!

Jennifer
  #5  
Old 6th January 2005, 12:46 PM
gaugefixer gaugefixer is offline
E-Mails Invalid or Rejected by Recipient System

 
Registration Date: Jul 2004
Location: Ontario, Canada
 
Posts: 30
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 0
Karma: 48
gaugefixer has less than 100 Karma points so far.
Hi Jen,

Yes, perfectly acceptable. I am wondering if you do any calibrations in house or do you send them all out? Seems like a waste to send out a 6" digital caliper for calibration if you have your own gauge blocks and a few basic tools.
If you have gauge blocks you can do your own calibrations; you would just need to record the results and the procedure you used.

Roger



Quote:
In Reply to Parent Post by JSmith

Hello,

I have a question regarding the calibration / verification of Digital Calipers.

If my company uses calipers to measure parts up to 1 1/2 " only, would it be acceptable for me to verify the 6" calipers from 1 - 2 " with my gage blocks over that range and not send them out for calibration to an outside company?

Thanks

Jen
  #6  
Old 7th January 2005, 08:46 AM
JSmith JSmith is offline
Email Activation Not Completed

 
Registration Date: Mar 2002
Location: Clawson, MI
Age: 39
 
Posts: 20
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 0
Karma: 107
JSmith is appreciated, and has over 100 Karma points.JSmith is appreciated, and has over 100 Karma points.
Hi Roger,

I have gage blocks to calibrate the od side. Will I need to calibrate (verify) the ID side as well? If so do I have to use a set of ring gages or could I use certified micrometers to calibrate the ID?

Thanks for your help,

Jennifer
  #7  
Old 7th January 2005, 09:04 AM
Kevin H's Avatar
Kevin H Kevin H is offline
Lab Manager

 
Registration Date: Jun 2004
Location: Western Pennsylvania, USA
Age: 60
 
Posts: 489
Thanks Given to Others: 19
Thanked 50 Times in 40 Posts
Karma Power: 87
Karma: 1203
Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.
Jen, If you're using both ID & OD sides of the calipers you need to calibrate both. You can end up with a minor, visually undetectable flaw that will invalidate the measurement. I've used micrometers to verify the ID jaws of a digital caliper by first using a gage block to set the micrometer, locking the micrometer, then calibrating the ID jaws of the caliper, then double checking the locked micrometer with the gauge block. This was acceptable to our A2LA auditors for an ISO Guide 25 mechanical testing lab at a steel mill.
  #8  
Old 7th January 2005, 12:13 PM
gaugefixer gaugefixer is offline
E-Mails Invalid or Rejected by Recipient System

 
Registration Date: Jul 2004
Location: Ontario, Canada
 
Posts: 30
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 0
Karma: 48
gaugefixer has less than 100 Karma points so far.
Hi again,

I agree totally with what Kevin said. You can also use master rings if you have them.

Roger
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Dial Caliper or Digital Caliper - Which is the best? jhoniegudel ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 27 17th May 2011 03:52 PM
Gage not calibrated to full use/range of the gage - Surveillance Audit - Calibration 5litre ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 7 15th August 2009 07:40 PM
Best Caliper for Shop Use? Digital vs. Dial Caliper vs. Vernier-type Lundberg ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 53 29th September 2008 02:39 PM
Parts Counting Scale - Verification or Calibration? andre-soares ISO/TS 16949 - International Automotive Quality Systems Standard 1 19th September 2005 07:10 AM
Full Scale Tolerance - How to apply full-scale accuracy specification to a gage? e006823 ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 4 2nd August 2004 04:34 PM



The time now is 01:57 AM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


   


Marc Timothy Smith - Elsmar.com
8466 LeSourdsville-West Chester Road, West Chester, Ohio 45069-1929
513 341-6272