The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > ISO 13485 and ISO 14969 - Medical Devices - Quality Management Systems
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality

Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 4th January 2005, 04:54 PM
Dan LaBelle
Unregistered Guest

 
 
Posts: n/a
Thank You! Implementing ISO 13485:2003 - 8.5.1 - Advisory Notices - Recall Procedure

Hello,
I was wondering if I could get some ideas from anyone who has already or is soon to be implementing ISO 13485:2003 in regards to Advisory Notices. I may not be looking into it far enough, but in 8.5.1 states the requirement for a procedure(s) for the issue and implementation of advisory notices. Following the definition of Advisory Notice, would it be wise to add an additional process in the Recall Procedure, where it would document the notification to customer(s) as an Advisory Notice (use of, modification of, return of, or destruction of) in regards to the medical device? I was thinking of also adding a "yes/no", in the Engineering Change approval process to ensure if an Advisory Notice is needed.

Let me know your thoughts.

Thanks,

Dan

Sponsored Links
  #2  
Old 5th January 2005, 10:30 AM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Super Moderator

 
Registration Date: Jun 2002
Location: Lawn Guyland
Age: 62
 
Posts: 3,547
Thanks Given to Others: 71
Thanked 717 Times in 442 Posts
Karma Power: 413
Karma: 7090
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
I added the issuance of advisory notices to our recall procedure. I did not, however, revise our EC approval process. What % of Engineering Changes require an Advisory Notice?
__________________
Al
Thanks to Al Rosen for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 5th January 2005, 03:12 PM
Dan LaBelle
Unregistered Guest

 
 
Posts: n/a
Quote:
In Reply to Parent Post by Al Rosen

I added the issuance of advisory notices to our recall procedure. I did not, however, revise our EC approval process. What % of Engineering Changes require an Advisory Notice?
Thanks Al, you are right, it would be a very low %, and usually that would result from a Customer Complaint which is linked to the recall process.
  #4  
Old 9th October 2005, 01:17 AM
Chan J Chan J is offline
Inactive Registered Visitor

 
Registration Date: Oct 2005
Location: Singapore
 
Posts: 9
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 31
Karma: 10
Chan J has less than 100 Karma points so far.
Hi, just would like to know, that do we need to send an advisory notice to our suppliers as well should a product recall happen? Since we control our supplier, should they have a vigilence system? Who actuallly are the stakeholders should we issue an advisory notice? Sorry for the many questions.Thanks
  #5  
Old 9th October 2005, 09:34 AM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Super Moderator

 
Registration Date: Jun 2002
Location: Lawn Guyland
Age: 62
 
Posts: 3,547
Thanks Given to Others: 71
Thanked 717 Times in 442 Posts
Karma Power: 413
Karma: 7090
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
Quote:
In Reply to Parent Post by Chan J

Hi, just would like to know, that do we need to send an advisory notice to our suppliers as well should a product recall happen?
It would seem to me that your supplier should be involved long before an advisory notice is sent.

Quote:
In Reply to Parent Post by Chan J

Since we control our supplier, should they have a vigilence system?
I think not.

Guidance Document MEDDEV 2.12-1 describes the Vigilence System.

Quote:
In Reply to Parent Post by Chan J

Who actuallly are the stakeholders should we issue an advisory notice?
This is also addressed in MEDDEV 2.12-1.
Attached Files: 1. Scan for viruses before using, 2. Please report any 'bad' files by Reporting this post, 3. Use at your Own Risk.
File Type: pdf meddev 2_12-1 vigilence-sys.pdf (115.5 KB, 557 views)
__________________
Al
Thanks to Al Rosen for your informative Post and/or Attachment!
  #6  
Old 9th October 2005, 04:05 PM
Roland Cooke Roland Cooke is offline
Appreciated Member

 
Registration Date: Sep 2005
Location: Houston (ex-pat Brit)
 
Posts: 602
Thanks Given to Others: 45
Thanked 292 Times in 197 Posts
Karma Power: 94
Karma: 1366
Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.
For companies that are unlikely to experience much need for advisory notices, performing a dummy recall exercise is never a bad thing.

It means that should a problem arise at some point in the future, your company is not testing its system, out for the first time in the midst of what might well be a crisis situation.

Traceability and supplier/distributor communication lines can often be significantly strengthened as a result of this exercise, and if nothing else it helps spread a sense of confidence and competence in the system.
  #7  
Old 19th October 2005, 06:22 AM
Miragesy's Avatar
Miragesy Miragesy is offline
Inactive Registered Visitor

 
Registration Date: Sep 2005
Location: turkey
 
Posts: 2
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 32
Karma: 10
Miragesy has less than 100 Karma points so far.
Hi,
I'm new in Elsmar Cove. I find the site very outstanding.
I also have trouble with advisory notice procedure. Should there be a seperate procedure for advisory notices or are recall, customer complaint, post production surveillance, etc. procedures enough?
Where can I find a sample for advisory notice procedure if these are not enough?
Thank you in advance.
Serap
  #8  
Old 19th October 2005, 06:40 AM
Roland Cooke Roland Cooke is offline
Appreciated Member

 
Registration Date: Sep 2005
Location: Houston (ex-pat Brit)
 
Posts: 602
Thanks Given to Others: 45
Thanked 292 Times in 197 Posts
Karma Power: 94
Karma: 1366
Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.Roland Cooke is appreciated, and has over 1300 Karma points.
Quote:
In Reply to Parent Post by Miragesy

Hi,
I'm new in Elsmar Cove. I find the site very outstanding.
I also have trouble with advisory notice procedure. Should there be a seperate procedure for advisory notices or are recall, customer complaint, post production surveillance, etc. procedures enough?
Where can I find a sample for advisory notice procedure if these are not enough?
Thank you in advance.
Serap
(In European terminology) a recall is one type of advisory notice, specifically demanding the product be returned/destroyed.

AFAIK, in the US, a recall is any type of regulatory/complaint communication.



In your QMS, you should have a cascading system:


1. Risk Management - problem prevention

2. Post Market Surveillance (not just complaints!!!!!!!)

3. Complaints Handling (sorting price complaints from safety complaints)

4. Vigilance Complaints Handling (confirmation, determination of seriousness, communication with regulators)

5. Advisory Notices (issued by manufacturer to users, advising them of product issues. Often the product does NOT need to be returned/destroyed. Usually would be copied to regulators.)

6. Recall Process (a specific advisory notice, which requires the product to be returned/destroyed. Mandatory to involve regulators when a recall occurs. An established procedure to ensure that the recall will be effective, however running dummy recalls is recommended to validate the system especially wrt communication and traceability.).

7. Risk Management - incorporating lessons learned.


There is no requirement to have more than one procedure covering these different processes. I would however strongly recommend it!

Last edited by Roland Cooke; 19th October 2005 at 06:52 AM. Reason: tweaked #4
Thank You to Roland Cooke for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > ISO 13485 and ISO 14969 - Medical Devices - Quality Management Systems

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Established procedure for Advisory Notices - ISO 13485 Chiefs491 ISO 13485 and ISO 14969 - Medical Devices - Quality Management Systems 7 25th June 2008 10:31 AM
Corrections and Removals and Advisory Notices: 21 CFR 806 (Procedure needed) CAHARDIN Documentation Control Systems, Procedures, Forms and Templates 1 9th October 2007 09:07 AM
Advisory Notices - Who is responsible for dealing with Advisory Notices? Marc ISO 13485 and ISO 14969 - Medical Devices - Quality Management Systems 7 11th September 2007 04:41 PM
ISO 13485 - Advisory Notices - Regulations - Quality Manual QA Tech ISO 10013 - Quality Management System (QMS) Manuals 13 28th September 2006 05:00 AM