The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 1 votes, 3.00 average. Display Modes
  #1  
Old 14th March 2005, 01:56 PM
ozziegood's Avatar
ozziegood ozziegood is offline
Registered Visitor

Registration Date: Jan 2005
Location: Holland, MI
 
Posts: 24
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 20
Karma: 10
ozziegood has less than 100 Karma points so far.
Question Scale Calibration Frequency - Low Volume - Current Cycle is 6 Months

I am wondering if ther is some sort of calculation to the frequency of calibrating scales. We do not use them in high volume and are on a 6 month calibration. I am not sure that this is feasible. Anyone have any thoughts?!
Reply With Quote

Sponsored Links
  #2  
Old 14th March 2005, 02:02 PM
Teri Teri is offline
Courtesy Access

Registration Date: Mar 2003
Location: Ohio
Age: 39
 
Posts: 196
Thanks Given to Others: 3
Thanked 17 Times in 13 Posts
Karma Power: 34
Karma: 197
Teri is appreciated, and has over 100 Karma points.Teri is appreciated, and has over 100 Karma points.
Default

Ours is also on a 6 month shedule... never been a problem. We do them right before our year end inventory and again during a July shutdown.
__________________
Teri
Mistakes are a fact of life. It is the response to errors that count!
Reply With Quote
Sponsored Links

  #3  
Old 14th March 2005, 02:14 PM
Jim Wynne's Avatar
Jim Wynne Jim Wynne is offline
Courtesy Access

Registration Date: Jan 2005
Location: Southeast Wisconsin
Age: 57
 
Posts: 9,216
Thanks Given to Others: 755
Thanked 2,298 Times in 1,550 Posts
Karma Power: 612
Karma: 20405
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Default

Quote:
Originally Posted by ozziegood

I am wondering if ther is some sort of calculation to the frequency of calibrating scales. We do not use them in high volume and are on a 6 month calibration. I am not sure that this is feasible. Anyone have any thoughts?!
Calibration intervals should be based on experience. If, when devices are calibrated, it's found that they need significant adjustment, intervals should be shortened. Conversely, if no significant adjustments are needed, you might be justified in extending the intervals. It's not a good idea to set intervals arbitrarily.
__________________
Some men are born mediocre, some men achieve mediocrity, and some men have mediocrity thrust upon them.-- Joseph Heller
Reply With Quote
  #4  
Old 14th March 2005, 02:16 PM
Kevin H's Avatar
Kevin H Kevin H is offline
Quality Metallurgist

Registration Date: Jun 2004
Location: Western Pennsylvania, USA
Age: 57
 
Posts: 466
Thanks Given to Others: 19
Thanked 33 Times in 27 Posts
Karma Power: 54
Karma: 1123
Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.
Default

We're at 6 months for shipping scales, etc. One reference to check is NIST Handbook 44, Specifications, Tolerances, and Other Technical Requirements for Weighing and Measuring Devices. It is viewable online at: http://ts.nist.gov/ts/htdocs/230/235/h442003.htm

The other question to ask is if you decrease your calibration, and you discover you were out of calibration, investigated and found possible non-conforming product shipped to your customer(s), and were then required to notify your customer(s), what effect would that have on your business on a long term basis?
Reply With Quote
  #5  
Old 14th March 2005, 08:03 PM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 53
 
Posts: 1,621
Thanks Given to Others: 1
Thanked 310 Times in 224 Posts
Karma Power: 186
Karma: 6526
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Default

Another source to consider is RP-1 (Recommended Practice) from NCLSI (National Conference of Standards Laboratories - International).

http://www.ncsli.org

It is the RP that provides guidance for establishing cal intervals.

Hershal
Reply With Quote
  #6  
Old 15th March 2005, 02:23 PM
Randy Stewart's Avatar
Randy Stewart Randy Stewart is offline
VINI, VIDI, DORMIVI

Registration Date: Feb 2002
Location: Walled Lake, Michigan
 
Posts: 1,153
Thanks Given to Others: 20
Thanked 79 Times in 51 Posts
Karma Power: 84
Karma: 1481
Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.
Default

ozziegood,
When I worked as a cal tech we always used a 10% rule - that is increase or decrease cal intervals by 10%. That way you don't find yourself giving the scales an extra 6 months, pushing the cal out to a year, only to find they failed at 7 months.
__________________
"What this country needs is more unemployed politicians." Edward Langley
Reply With Quote
  #7  
Old 18th March 2005, 02:02 AM
practitioner practitioner is offline
Inactive Registered Visitor

Registration Date: Mar 2005
Location: California
 
Posts: 2
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 20
Karma: 10
practitioner has less than 100 Karma points so far.
Default Calibration Intervals

Some good advice has already been stated. Base your frequency on experience. More important is to maintain a verification scheme that allows you to quickly and frequently check your scales and balances. That way you don't get surprises and you maintain confidence in your measurements in between calibrations.
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Implementation of TS 16949 for Low Volume Production Ramatein ISO/TS 16949 - International Automotive Quality Systems Standard 2 30th September 2009 07:40 AM
SPC Techniques - Low volume environment - What can be applied? cbearden SPC Monitoring and Statistical Analysis Techniques 6 15th February 2009 08:51 PM
Supplier Rating System - Weighting low volume suppliers vs. high volume suppliers fcardillo Supplier Quality Assurance and other Supplier Issues 2 2nd May 2008 08:05 AM
Low Volume Manufacturing - What is Low Volume Manufacturing? Ady12 Definitions, Acronyms, Abbreviations and Interpretations 9 5th September 2006 07:40 PM
Short Run - Low Volume Cpk Calculation tohbin Capability, Accuracy and Stability - Process, Machine, etc. 10 20th February 2002 05:28 PM



The time now is 04:09 AM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts