Is it possible to have a process that is too good? Cpk is 14.63
Is it possible to have a process that is too good. The reason I ask, is that I have conducted a stability study on a gauge. The Xbar chart shows points out of control. The mean is 36.007. The UCL is 36.014 and the LCL is 36.001 The rbar is .011 and the specification limits are 36.000±.2 I also get a CPk figure of 14.63!! Which I know is invalid if the process is not in control. The range chart is in control. I am mystified as to why the chart would indicate that the process is not stable. Perhaps someone can enlighten me.
|

28th April 2005, 12:06 PM
|
 |
Courtesy Access
Registration Date: Jan 2005
Location: Southeast Wisconsin
Age: 57
|
|
Posts: 9,210
Thanks Given to Others: 755
Thanked 2,292 Times in 1,546 Posts
Karma Power: 611
|
|
Quote:
|
Originally Posted by DJN
I am mystified as to why the chart would indicate that the process is not stable.
|
We have no way of knowing why you have points beyond the control limits. What are you measuring? Do you see similar problems in the process data?
__________________
Some men are born mediocre, some men achieve mediocrity, and some men have mediocrity thrust upon them.-- Joseph Heller
|

28th April 2005, 12:34 PM
|
|
Involved - Posts
Registration Date: Mar 2005
Location: Pennsylvania
|
|
Posts: 65
Thanks Given to Others: 0
Thanked 3 Times in 2 Posts
Karma Power: 27
|
|
Yes - We cut wire to length using an automatic machine. The machine was very repeatable and having a product spec of +/- 1/2" didn't hurt either. We had CPK's in the 70 - 80 range. After a few years we backed off on our frequncy of measurements. Never had a problem with this number from the customer or the auditor (once they understood the process)
__________________
:cool:
|

28th April 2005, 12:38 PM
|
|
Involved - Posts
Registration Date: Jul 2003
Location: England
|
|
Posts: 67
Thanks Given to Others: 8
Thanked 3 Times in 3 Posts
Karma Power: 30 Karma: 180  
|
|
It does effect GR&Rpv
Quote:
|
Originally Posted by DJN
Is it possible to have a process that is too good. The reason I ask, is that I have conducted a stability study on a gauge. The Xbar chart shows points out of control. The mean is 36.007. The UCL is 36.014 and the LCL is 36.001 The rbar is .011 and the specification limits are 36.000±.2 I also get a CPk figure of 14.63!! Which I know is invalid if the process is not in control. The range chart is in control. I am mystified as to why the chart would indicate that the process is not stable. Perhaps someone can enlighten me.
|
I don't know about effecting stability but it does effect GR&Rpv.
For example I recently improved a process very significantly 1.8 to 6+. GR&Rpv on new process failed because I cant improve the gauge enough for the drop in the process range value, needless to say I applied common sense MSA Page 23, and told my customer regarding cost effectiveness of improvements MSA Page 46
Martin
Older not wiser
|

28th April 2005, 12:45 PM
|
 |
Involved - Posts
Registration Date: Feb 2005
Location: toronto, ontario, canada
|
|
Posts: 188
Thanks Given to Others: 8
Thanked 0 Times in 0 Posts
Karma Power: 27
|
|
The problem i see is in your measurement system. Instead of measuring to 1/1000th of an inch, try to round off to 1/100th of an inch. Your chart will be more stable and in-control.
Remember, out of control condition includes other unusual patterns (runs, cycles, etc), not just points beyond the control limits.
|

28th April 2005, 01:27 PM
|
 |
Heretical Statistician
Registration Date: Jan 2003
Location: New England
|
|
Posts: 1,134
Thanks Given to Others: 265
Thanked 516 Times in 340 Posts
Karma Power: 137
|
|
Did you do a gauge R&R study and plot the numbers on the control chart? Such that the range data is the range of the repeated measures of the same unit and the average values are the averages of the repeated measurements of each part?
If so - this would give you in control range values and a lot of 'out of control' averages...because you are calcualting the control limits from the average range which would be very tight if your gauge is reasonable. Then the averages are from PARTS wich vary greatly from each other - in comparison to the difference between repeated measures of the same parts...
Also, depending on your software or formula choice - if you calculated the traditional Cpk value (as opposed to the Ppk value) you would have used the within subgroup standard deviation which comes from the range of the repeated measures...again this woulg give you a very small standard deviation compared to your spec limits...
|

28th April 2005, 03:46 PM
|
 |
VINI, VIDI, DORMIVI
Registration Date: Feb 2002
Location: Walled Lake, Michigan
|
|
Posts: 1,153
Thanks Given to Others: 20
Thanked 79 Times in 51 Posts
Karma Power: 84
|
|
Quote:
|
The problem i see is in your measurement system.
|
I have to agree, it's like you are trying to measure a klick using a meter stick while gauging errors in centimeters. Reduce the accuracy of the gage or tighten up the UCL & LCL.
__________________
"What this country needs is more unemployed politicians." Edward Langley
|

29th April 2005, 03:15 AM
|
 |
Forum Administrator
Registration Date: May 2000
Location: Eskilstuna, Sweden
Age: 49
|
|
Posts: 3,771
Thanks Given to Others: 246
Thanked 244 Times in 172 Posts
Karma Power: 213
|
|
Considering the original question...
Is it possible to have a process that is too good? Cpk is over 14
I suppose it is possible: Let's assume you buy top end CNC equipment, and then use it for producing a low spec product you could easily make with an old lathe from 1870 (An extreme case, I know, but just to make an example)?
/Claes
|
Lower Navigation Bar
|
|
|
|
Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
|
|
|
| Thread Tools |
Search this Thread |
|
|
|
| Display Modes |
Rate Thread Content |
Linear Mode
|
|
Posting Settings
|
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
HTML code is Off
|
|
|
Similar Discussion Threads
|
| Discussion Thread Title |
Thread Starter |
Forum |
Replies |
Last Post or Poll Vote |
|
Is Cpk a good measure of capability? There are several shortcomings of Cpk
|
Tim Folkerts |
Capability, Accuracy and Stability - Process, Machine, etc. |
31 |
12th February 2009 07:39 PM |
|
Quality Control Index - We thought that Cpk or Ppk would be good choices
|
vskiller |
Capability, Accuracy and Stability - Process, Machine, etc. |
6 |
28th June 2007 12:57 PM |
|
Can CPK be a good manufacturing processor’s quality objective?
|
darkafar |
Capability, Accuracy and Stability - Process, Machine, etc. |
14 |
22nd March 2007 12:58 PM |
|
How much does Cpk > 1.33 represent in precentage of good pieces?
|
tattva |
Capability, Accuracy and Stability - Process, Machine, etc. |
17 |
14th November 2003 11:09 AM |
|
Cp and Cpk - Cpk is an index that should be applied to a stable process
|
Don Winton |
Capability, Accuracy and Stability - Process, Machine, etc. |
16 |
13th August 2001 06:50 PM |
|
|