The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 1 votes, 5.00 average. Display Modes
  #1  
Old 14th October 2005, 06:05 PM
joecalibratio's Avatar
joecalibratio joecalibratio is offline
Cal Supervisor

Registration Date: Oct 2005
Location: Snoqualmie, WA, USA
 
Posts: 11
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 17
Karma: 10
joecalibratio has less than 100 Karma points so far.
Let Me Help You Recommendations on calibration interval (frequency) ISO/IEC 17025:2005 para 5.10.4.4

I.A.W. ISO/IEC 17025:2005 section 5.10.4.4 "A calibration certificate (or label) shall not contain any recommendation on the calibration interval except where this has been agreed with the customer. This requirement may be superceded by legal regulations."

Does anyone know the reason behind this paragraph? I am under the understanding that the manufacturers recommendations are typically utilized for calibration intervals. This paragraph has some ambiguity with "agreed with the customer" and "superceded by legal regulations".

Does this mean we leave the Expiration date blank? I do not believe the majority of our customers would be receptive to this.

Joel
Reply With Quote

Sponsored Links
  #2  
Old 14th October 2005, 06:41 PM
Jerry Eldred's Avatar
Jerry Eldred Jerry Eldred is offline
Forum Moderator

Registration Date: Dec 1999
Location: Texas
 
Posts: 411
Thanks Given to Others: 0
Thanked 118 Times in 70 Posts
Karma Power: 89
Karma: 2042
Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.
BIG Smile

Although I'm not an expert on this, I will give it a shot. The "Certificate of Calibration" is essentially a statement of as found/as released quantitative information about how the instrument is performing at a given time. Establishing the interval is a separate issue.

For example, I am currently working on statistical evaluation of about 86,000 calibrations in my company's calibration recall database. This is based on NCSLI Recommended Practice method 3 (if I recall corrrectly - sorry, it's late on Friday afternoon). The interval is a statistical probability to a given confidence of how long an instrument may be expected to remain within tolerance. It may be lengthened when prior as received calibration history allows it, or may be shortened when prior history dictates that way.

The given users accuracy requirements, frequency of use, environmental factors, the "lemon factor" (if you have one bad one in an otherwise "good" population of instruments, it has a "lemon factor") and other factors can impact what the interval should be set to.

This is distinct from the Certificate of Calibration. Particularly if you send to the OEM or Third Party lab, since they don't have your internal data, it isn't fair for them to assign an interval. This may potentially create an undue increased risk of out-of-tolerance performance, and possible conflicts.

For (fictitious) example, suppose I have 100 Fluke 77 multimeters (no good or bad implications about Fluke 77; it's just a very common instrument). I've (fictitiously) been calibrating them for the last 10 years with an initial assigned interval of 12 months. After 10 years of history, I find that 5 of the meters have a higher out-of-tolerance history than the others. I research and find that these are used by a crew outdoors. They are frequently dropped, exposed to higher voltages, lay in the truck in the heat, and so forth. I've had to adjust that one little potentiometer in these five a few times, where I haven't ever adjusted any other in my entire population of Fluke 77's.

My recall system and calibration records state the assigned interval (which I statistically derived based on history - still, a fictitious example). However, I sent two of the bad meters back to Fluke, where they calibrated them twice. They sent back a Certificate of Calibration with them, but did not provide a due date, since it isn't appropriate for them to include that in the certificate. When I received them back from Fluke, even though I increased the intervals on all the remaining population to 30 months, these five are at 12 months, and 2 of them are at 6 months.

Hopefully this will shed some amount of light on the question. Have a good weekend. I'm going home.
Reply With Quote
Sponsored Links

  #3  
Old 14th October 2005, 06:49 PM
Marc's Avatar
Marc Marc is offline
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,859
Thanks Given to Others: 1,895
Thanked 1,569 Times in 1,020 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11574
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
Default

Thanks for the detailed reply. Excellent post. Excellent example. Thanks, Jerry!
__________________
A Search is a terrible thing to waste!
One Test is Worth 1000 Expert Opinions - The plural of anecdote is not data.
We can't solve problems by using the same kind of thinking we used when we created them. - Unknown
Reply With Quote
  #4  
Old 15th October 2005, 02:27 AM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 53
 
Posts: 1,621
Thanks Given to Others: 1
Thanked 310 Times in 224 Posts
Karma Power: 186
Karma: 6526
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Default

Quote:
Originally Posted by joecalibratio

I.A.W. ISO/IEC 17025:2005 section 5.10.4.4 "A calibration certificate (or label) shall not contain any recommendation on the calibration interval except where this has been agreed with the customer. This requirement may be superceded by legal regulations."

Does anyone know the reason behind this paragraph? I am under the understanding that the manufacturers recommendations are typically utilized for calibration intervals. This paragraph has some ambiguity with "agreed with the customer" and "superceded by legal regulations".

Does this mean we leave the Expiration date blank? I do not believe the majority of our customers would be receptive to this.

Joel
Actually, it does mean leave the due date blank, UNLESS there is an agreed-upon calibration cycle.

Why the clause? Users may simply use an annual cycle or may want to change the cycle based on some criteria or lack of criteria. So, this clause is for the protection of both the customer AND the cal lab. The cal lab is not suggesting or requiring some arbitrary date to generate business, and the customer is free to select their own date.

Now, most customers will ask their cal house for a recommendation. That's OK and desired.....I suggest it gets documented.

Hope this helps.

Hershal
Reply With Quote
  #5  
Old 15th October 2005, 02:28 AM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 53
 
Posts: 1,621
Thanks Given to Others: 1
Thanked 310 Times in 224 Posts
Karma Power: 186
Karma: 6526
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Default

One last post.....Jerry, RP-1.
Reply With Quote
  #6  
Old 23rd October 2005, 11:14 PM
Wayne's Avatar
Wayne Wayne is offline
Gage Crib Worldwide

Registration Date: Aug 2005
Location: Grand Rapids, MI USA
Age: 58
 
Posts: 170
Thanks Given to Others: 14
Thanked 65 Times in 36 Posts
Karma Power: 46
Karma: 1298
Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.
Send a message via Skype™ to Wayne
Default

The customer is the only one who can accurately specify the calibration cycle. The calibration lab is not in any way qualified to suggest the calibration cycle.

My experience is with hard gages. I frequently advise my customers that time is not the best unit of measure to use when determining calibration increment. When using hard gages, they have a tendency to wear or gaul with use. The operative variable here is ‘with use.’ With use brings its own variables to the table.

How many uses?
A hard gage that is used once a year, needs to have a greater interval than a gage used a thousand times a day.

Against what material is it used?
Some materials are more abrasive than others. Very soft materials being manufactured with dirty cutting fluid can pick-up worn particles of carbide suspended in the cutting fluid. These particles then become imbedded in the product and act as a lap against the gage causing rapid gage wear. Some materials are more gummy. They leave material welded to the gage surface thus causing the gage to grow.

How close to the mean is the product being checked?
If the product being measured is made close to the maximum material condition, the GO gage will have much more gage to product contact. With more contact comes more wear. If made closer to the minimum material condition the NOGO gage will be used more and thus experience more wear.

What type and amount of lubrication is used?
Cutting fluid is not lubricant. Lubricant does not effect the measurement properties of the gage. Keep the gage well lubricated and you will extend the life of the gage. Conversely, use no lubricant and, well, you know.

What are the work station conditions?
Kept clean. Gage not laying in a pile of chips. Machined product chips removed from the work piece before applying the gage. Chips between the gage and the work piece cause gage wear. Cutting tools kept sharp. Burrs left on the work piece get between the work piece and the gage cause wear.

Who is using the gage?
Popeye on spinach or Olive Oil? If more muscle is used the more abrasion can occur. More abrasion means wear.

Hard gages should be checked when first received from the supplier. If not used they should be checked again at one year to check for gage growth. After that the gage can be considered stable for long term storage. Calibration cycle should start at first use after calibration. Calibration cycle should be based on number of uses the gage sees. Using the number of uses provides a more accurate measure than time. Consideration should be given to the above variables when determining the specified calibration cycle. This is a good use for statistical evaluation.

In the final analysis, time is the almost the worst measure to use to determining calibration cycle for hard gages, but time is usually what is recommended by calibration labs.
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Calibration Frequency of CMM as per ISO/IEC 17025 dandrewsca Calibration Frequency (Interval) 4 24th October 2009 07:47 PM
ISO/IEC 17025:2005 Clause 4.1.5 and 4.1.6 Hershal ISO 17025 - Calibration, Measurement Gages and Test Laboratories 5 27th November 2008 11:44 PM
Appropriate Equipment for ISO/IEC 17025:2005 Clause 5.6.3.3 - Intermediate Check Dmokong ISO 17025 - Calibration, Measurement Gages and Test Laboratories 5 6th February 2008 06:13 AM
Do we need to revise our Quality Policy for ISO/IEC 17025:2005? amanbhai ISO 17025 - Calibration, Measurement Gages and Test Laboratories 2 10th April 2006 04:53 PM
ISO/IEC 17025 Revised: ISO/IEC 17025:2005 Sidney Vianna ISO 17025 - Calibration, Measurement Gages and Test Laboratories 16 27th July 2005 12:49 AM



The time now is 12:07 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts