The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 2 votes, 5.00 average. Display Modes
  #1  
Old 26th October 2005, 06:41 PM
Jerry Eldred's Avatar
Jerry Eldred Jerry Eldred is offline
Forum Moderator

Registration Date: Dec 1999
Location: Texas
 
Posts: 411
Thanks Given to Others: 0
Thanked 118 Times in 70 Posts
Karma Power: 89
Karma: 2042
Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.
BIG Smile Adjustment of Calibration Intervals (Calibration Frequency)

I am in the middle of a thorough evaluation of the intervals for 86,000 calibrations for my company. I'm curious to benchmark some of the criteria used by everyone out there. There doesn't seem to be a consensus or best practice as to some of the base statistical parameters. Please post your inputs or thoughts on the below parameters:

1. Reliability Target (percent in-tolerance): 80% to 95% seems to be the norm. What does everyone think?

2. Confidence Factor (K factor for making cal interval adjustments): I've seen everywhere from 20% to 99.9999%.

3. Maximum Interval (electronic test equipment, mostly): 3 - 5 years seems to be the norm.

4. Minimum Interval (not so critical): Some special cases probably once a day; but for most instruments, I've seen 1 - 3 months as the norm.

5. Renewal (optimize or adjust to nominal): Only when out of tolerance, at 70% of tolerance, at 95% of tolerance, always, are the most common.

6. Interval Evaluation/Adjustment Method: Common answers are NCSLI RP1 Method A3 based on % in-tolerance, statistical evaluation of data, other NCSLI RP1 methods.

OTHER OPTIONS:
1. Fixed interval: Pick a convenient interval and leave there.
2. The Old "1-2-3" Method: I don't like, but a common method (or variant) is every 3rd calibration (or other low number), calculate % IN-Tolerance for each instrument. Based on % IN-Tolerance, either bump up by 20%, leave the same, bump down by 20%, or 50%. This (in my opinion) is an over-reactive method that over-adjusts, often leaving either too short an interval or too long an interval.

Looking forward to your inputs. -Jerry
Reply With Quote

Sponsored Links
  #2  
Old 27th October 2005, 01:08 AM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 53
 
Posts: 1,621
Thanks Given to Others: 1
Thanked 310 Times in 224 Posts
Karma Power: 186
Karma: 6526
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Default

Hi Jerry,

Reliability target is ok.

Confidence is typically k=2 to approximate 95%.....actually k=2 is 95.5%.....99% is k=3.....k=2 is easier, and more appropriate for a non-NMI calibration.

For interval I suggest using RP-1 from NCSLI.....it has been my experience that electronic is better suited to a max of 12 months rather than a max of 36 months.....remember electronic instruments are more likely to drift than a set of calipers or a gage block.....I could see gage blocks going to 36 or even 60 months IF they are kept in a tightly controlled environment and rarely used.

For min, 1-3 months is typical.

If adjustment is required, always shoot for nominal.....gives the greatest chance of staying in tolerance longer.

RP-1 is a good method.....if you have access to the Navy or AF interval information, that is better.

For some things, a set interval, based on storage or use, is ok. Gage blocks are an example. Oscilloscopes (older ones) may have MIPs (Multiple Interval Parameters).

Just my thoughts. Hope this helps.

Hershal
Reply With Quote
Thanks to Hershal for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 27th October 2005, 03:48 PM
JerryStem's Avatar
JerryStem JerryStem is offline
Involved - Posts

Registration Date: Apr 2001
Location: W Chester OH USA
 
Posts: 269
Thanks Given to Others: 0
Thanked 2 Times in 2 Posts
Karma Power: 41
Karma: 36
JerryStem has less than 100 Karma points so far.
Default

Set my gage blocks at 36 months, little use. A2LA had no problem with that.

Jerry
Reply With Quote
Thanks to JerryStem for your informative Post and/or Attachment!
  #4  
Old 27th October 2005, 04:01 PM
Randy Stewart's Avatar
Randy Stewart Randy Stewart is offline
VINI, VIDI, DORMIVI

Registration Date: Feb 2002
Location: Walled Lake, Michigan
 
Posts: 1,153
Thanks Given to Others: 20
Thanked 79 Times in 51 Posts
Karma Power: 84
Karma: 1481
Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.Randy Stewart is appreciated, and has over 1300 Karma points.
Default

Back in the day when I was a cal tech, we used the 10% figure for increase in interval. As long as you call out the cal frequency units and keep it the same we never had an issue.
If you call out a 12 month interval for a thread plug then the 10% increase would be 1.2 months. Just keep the measuring units the same.
__________________
"What this country needs is more unemployed politicians." Edward Langley
Reply With Quote
Thanks to Randy Stewart for your informative Post and/or Attachment!
  #5  
Old 28th October 2005, 12:11 PM
Graeme's Avatar
Graeme Graeme is offline
$ Contributor

Registration Date: Sep 2000
Location: Lilburn, GA, USA
Age: 60
 
Posts: 401
Thanks Given to Others: 0
Thanked 8 Times in 7 Posts
Karma Power: 63
Karma: 894
Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.
Send a message via AIM to Graeme Send a message via MSN to Graeme Send a message via Yahoo to Graeme
I Say... My thoughts ...

Jerry,

Like you, virtually all of my workload is electronic test & measuring equipment. Here's what I do. It's based on experience in the Government (Department of the Navy), a regulated industry where safety is paramount (aviation) and other experience acquired from a number of sources.

1. I use 95% as the reliability target - the probability of a single item being in-tolerance at the end of the calibration recall period. In another industry I might go to 90% but I'm personally not comfortable with anything lower than that. Anthing higher than 95% is really not cost-effective because it takes way too long to collect "enough" data.

2. I use 95% as the confidence factor - the probability of making a correct in/out tolerance decision. Some would like to see 100% here, but the only way to get that is if it never leaves the cal lab, in which case it is perfectly reliable but totally useless to the end user.

3. I use a maximum interval of 60 months (5 years). The only product I have ever seen a longer inteval on is waveguide directional couplers (72 months) but about the only way they change is if they are physically damaged.

4. The minimum interval used right now is 1 month, but that is a special case. The "normal" minimum is 3 months. Some infrequently used equipment may be labeled "Calibrate before each use."

5. Adjustment (renewal) is performed if the measurement is more than +/- 75% of the calibration limits, or at a different value if specified in the calibration procedure. If the TAR is less than 4:1 I may specify a different adjustment limit in the procedure.

6. The lab supervisor recently purchased a commercial interval analysis program and that is being used now. It does some fancy statistical analysis of the data. Before that, I used a spreadsheet I developed that is based on method A3 but uses the binomial distribution to estimate an interval. (I have attached a copy of the spreadsheet.)

Other - fixed interval: I've never used that ... it seems somewhat foolish to me because you never have the opportunity to improve reliability by reducing an interval, or save money while maintaining the same reliability by increasing the interval.

Other - 1-2-3: I have never used this type of method but I have seen it used in many places. I will observe that RP-1 specifically states that this and other reactive methods (A1, A2 and A3 if used without statistical methods) never settle to a stable value. As you said, the interval is always either too long or too short.

Additional thoughts:

I never try to do interval analysis on a single item or even on a small number of similar items. It tales so long to get a statistically significant sample size that it often exceeds the useful lifetime of the equipment. (And that's almost exactly what RP-1 says as well.)

I try to group things by model where possible and have all of them on the same interval. For example, my client has about 350 Fluke 87 digital multimeters which is a large sample. However, there are also a lot of 3-1/2 digit multimeters from various manufacturers, and while there are over 100 total there are fewer than 10 from any single manufacturer. Since they are similar in type, function and accuracy all of those are treated together as a group for interval analysis.

I am very conservative in extending calibration intervals - I don't take the numbers at face value. I suggest doing no more than doubling the current interval even if the numbers say it could be more. One analysis showed that a particular model of instrument, that had been on a 12 month interval, could be extended to something like 9-1/2 years -- it was actually changed to 24 months, and will never be more than 60 months.

After changing an interval, it is best to wait for at least the time (old interval + new interval) before analyzing that group again. An axiom in RP-1 is that all items in the group must have been on the same interval, so that amount of time allows something that was calibrated the day before the change to complete the old interval and one cycle of the new interval.

For ease of calculation, all time intervals should be in the same units. That is why I alsways use months. It is a convenient size, but also years are not appropriate for a lot of things and weeks or days rapidly ballon into numbers that are hard to grasp. (How long is 1,461 days? How long is 208 weeks?) If calculations result in a fractional part on a result, it is always rounded down to an integer. (Both of those numbers equate to 48 months.)

This lab's definition of "calibration" specifically excludes anything that the equipment manufacturer may call calibration but is really setup, normalization, standardization or error-correction storage that is performed by the operator before use. The definition also recognizes that a large number of "calibration" procedures in equipment manuals are really factory adjustment procedures. According to NCSL RP-3, a fundamental assumption for writing a calibration procedure is that the item to be calibrated is in good working order and ready for use in every respect -- and the purpose of calibration is to verify that status by using traceable standards. On the other hand, many manufacturer procedures are adjustment or alignment procedures intended for use after the item has been manufactured or repaired, and is therefore in an unknown state.

Increasing and decreasing calibration intervals can be justified with different amounts of data. Increasing an interval requires a large sample size in order to get statistically significant results. Decreasing an interval can be justified with much less data. In the case of the product that is on a 1 month interval - it had been at three months, but between the three units there had been five failures in the previous 12 calibrations. Even without doing the math it is easy to see that this is way less than 95% reliability, and they would have to pass all of the next 100 calibrations to have a chance of meeting it. There was not a chance of that happening, so it was an easy decision.
Attached Files: 1. Scan for viruses before using, 2. Please report any 'bad' files by Reporting the post it is in, 3. Use at your Own Risk.
File Type: xls A3 testing.xls (20.0 KB, 853 views)
__________________
Graeme C. Payne
ASQ Sr. Member; CQE; CCT

"Does it matter if the measurement result is wrong?
If it does, then calibrate the instrument.
If it doesn't matter, they why are you making the measurement?"
(P. G. Stein, 2000)
Reply With Quote
Thank You to Graeme for your informative Post and/or Attachment!
  #6  
Old 31st October 2005, 10:46 PM
Charles Wathen Charles Wathen is offline
Involved - Posts

Registration Date: Feb 2002
Location: Temecula, CA
Age: 54
 
Posts: 86
Thanks Given to Others: 0
Thanked 6 Times in 5 Posts
Karma Power: 33
Karma: 17
Charles Wathen has less than 100 Karma points so far.
Default

As usual, Graeme is right on. I went through the same dilemma 4 years ago, until I stumbled upon a software program called Interval Analysis by Dr. Howard Castrup. You can get the free software here: http://www.isgmax.com/Freeware/A3Test.exe
main page:
http://www.isgmax.com/home_l.htm

It's basically the same as Graeme's spreadsheet program.

I perform this analysis once a year by grouping most of the MTE in groups. We have ~13,000 instruments in our database. Once I have the analysis completed, I send the results (via email) to our entire Engineering staff for review and comments. Most of the Engineers recommend that I do this more often, but I tried to explain to them that it's based on history, and this can take some time.

Since doing interval analysis, auditors have been impressed that this is even done. As Graeme mentioned, intervals less than one month are removed from our calibration system and tagged with a do not use label. The MTE that falls into this category will need to be replaced, redesigned, or tolerances opened so it does not fail in future analysis.
Reply With Quote
Thanks to Charles Wathen for your informative Post and/or Attachment!
  #7  
Old 9th October 2008, 05:28 PM
normzone normzone is offline
Registered Visitor

Registration Date: Jun 2004
Location: San Diego CA USA
 
Posts: 37
Thanks Given to Others: 30
Thanked 2 Times in 2 Posts
Karma Power: 23
Karma: 17
normzone has less than 100 Karma points so far.
Question Re: Adjustment of Calibration Intervals (Calibration Frequency)

---" I am very conservative in extending calibration intervals - I don't take the numbers at face value. I suggest doing no more than doubling the current interval even if the numbers say it could be more. One analysis showed that a particular model of instrument, that had been on a 12 month interval, could be extended to something like 9-1/2 years -- it was actually changed to 24 months, and will never be more than 60 months.

After changing an interval, it is best to wait for at least the time (old interval + new interval) before analyzing that group again. An axiom in RP-1 is that all items in the group must have been on the same interval, so that amount of time allows something that was calibrated the day before the change to complete the old interval and one cycle of the new interval. "---

What kind of language would you use in a calibration procedure ( AS9100 ) to reference a caveat allowing extending calibration interval for tools demonstrating consistant calibration status or little use?


Last edited by normzone; 9th October 2008 at 05:29 PM. Reason: format
Reply With Quote
Thanks to normzone for your informative Post and/or Attachment!
  #8  
Old 17th October 2008, 10:21 AM
bobdoering's Avatar
bobdoering bobdoering is offline
Stop X-bar/R Madness

Registration Date: May 2002
Location: LaGrange, OH
 
Posts: 2,089
Thanks Given to Others: 680
Thanked 563 Times in 469 Posts
Blog Entries: 13
Karma Power: 150
Karma: 3915
bobdoering is appreciated, and has over 1700 Karma points.
bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.
Send a message via Yahoo to bobdoering
Default Re: Adjustment of Calibration Intervals (Calibration Frequency)

Quote:
Originally Posted by Hershal View Post

If adjustment is required, always shoot for nominal.....gives the greatest chance of staying in tolerance longer.
If your most significant variation is drift, and, for example, the drift is upward, wouldn't it be better to adjust down to below the nominal to allow for the most drift? That would give the greatest chance of staying in tolerance longer for that case.

Something to ponder.
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)

Bookmarks

Tags
calibration, calibration frequency, calibration interval


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Calibration Intervals (Frequency) derived from Variables Data rdragons Calibration Frequency (Interval) 30 12th August 2008 12:45 AM
Basis for Calibration Method and Intervals (Frequency) - FAA approved repair station turbines Federal Aviation Administration (FAA) 10 4th February 2008 10:45 PM
Calibration Intervals (Frequency) for Stereoscopes, Microscopes, etc. freelovefest Inspection and Test, Sampling and Related Topics 2 12th October 2007 08:55 PM
Calibration Interval (Frequency) Adjustment using Hilliard type 2 algorithm A C Diggs Calibration Frequency (Interval) 11 7th January 2005 03:19 PM
Determination of Calibration Intervals and Frequency Analysis Marc Calibration Frequency (Interval) 6 21st March 2000 05:50 PM



The time now is 05:11 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts