The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 31st October 2005, 10:00 AM
mshell's Avatar
mshell mshell is offline
Involved - Posts

Registration Date: Aug 2003
Location: United States/Georgia
Age: 39
 
Posts: 401
Thanks Given to Others: 0
Thanked 7 Times in 7 Posts
Karma Power: 50
Karma: 815
mshell is appreciated, and has over 700 Karma points.mshell is appreciated, and has over 700 Karma points.mshell is appreciated, and has over 700 Karma points.mshell is appreciated, and has over 700 Karma points.mshell is appreciated, and has over 700 Karma points.mshell is appreciated, and has over 700 Karma points.mshell is appreciated, and has over 700 Karma points.
Please Help! Illuminance Meter - Extending Manufacturer's Recommended Calibration Frequency

We currently have 2 illuminance meters being calibrated annually by the manufacturer. I have reviewed the records of calibration for the last 3 years and found that the meters have been in calibration each time we have sent them out. The manufacturer certificate (traceable to NIST) requires annual calibration. I would like to extend the calibration period to every 2 years based on minimal use and previous test results. Can I do this without losing the NIST traceability?
__________________
Mshell
Reply With Quote

Sponsored Links
  #2  
Old 31st October 2005, 10:05 AM
Jim Wynne's Avatar
Jim Wynne Jim Wynne is offline
Courtesy Access

Registration Date: Jan 2005
Location: Southeast Wisconsin
Age: 57
 
Posts: 9,211
Thanks Given to Others: 755
Thanked 2,293 Times in 1,547 Posts
Karma Power: 611
Karma: 20380
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Default

Quote:
Originally Posted by mshell

Can I do this without losing the NIST traceability?
Yes. Calibration intervals are not related to traceability, unless whomever is doing the calibration is destroying their records after 12 months, which is highly unlikely. You determine the proper interval for your application, not the manufacturer or calibration source.
__________________
Some men are born mediocre, some men achieve mediocrity, and some men have mediocrity thrust upon them.-- Joseph Heller
Reply With Quote
Sponsored Links

  #3  
Old 31st October 2005, 10:41 AM
Jerry Eldred's Avatar
Jerry Eldred Jerry Eldred is offline
Forum Moderator

Registration Date: Dec 1999
Location: Texas
 
Posts: 411
Thanks Given to Others: 0
Thanked 118 Times in 70 Posts
Karma Power: 89
Karma: 2042
Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.
Default

The general rule is that you "may" extend interval when in-tolerance history says the probability is they will remain in-tolerance for the new longer interval, and you "must" decrease the interval when out-of-tolerance history so dictates.

Depending on what you must comply with, it may or may not be reasonable to double the interval. I ran the NCSLI RP1 Method A3 Interval Tester, and at 95% in-tolerance target, the longest new recommended interval was 14.4months (I varied % in-tolerance all the way to 1% and the confidence factor all the way down to 1% as well, and still only came up with 14.4 months as the maximum recommended interval. This is based on the National Conference of standards Laboratories Recommended Practices.

However, on the more practical side, if you really want to adjust the interval to 24 months, I would recommend reviewing the certificates, and perhaps calling the lab that cal'd them for you, and ask to determine if the units were ever adjusted or optimized by them during any of those calibrations. If they have never been adjusted or optimized, that would indicate that the units may have remained in-tolerance for the three years, and 24 month interval may be fully appropriate. Again, depending on what you must comply with (ISO, FDA, FAA, etc.), it may or may not be important for you to document that. The calibration interval is, after all is said and done, the statistically probable amount of time (or usage) within which an instrument may be expected (to a defined degree of confidence) to remain within specified tolerances.

As the first repy indicated, it will not magically go out of tolerance on the due date. What does happen is that from the instant the calibration label is applied (kind of like magic - sorry, it's Monday morning) the is a statistical increase in the probability that it will not remain in specs., which increases until the time at which it will likely be out of tolerance.

The more history you acquire, the higher the confidence in how long it will remain in tolerance. Method A3 uses the average interval and the maximum interval in statistically determining new intervals. IMPORTANT NOTE: When adjusting intervals, use the actual amount of time between calibrations (old cal date to new cal date), not the assigned interval (12 months). If you went significantly over 12 months, during any of the previous calibrations on your meters, that could increase the allowable new interval (per that recommended practice).

Also, as was previously posted, the customer assigns the interval (according to IEC/ISO17025). The manufacturer's interval is not binding, it is their recommendation. What that is based on depends on the manufacturer. You might even talk with someone at the manufacturer and ask them what it is based on, and what the risks might be with a two year interval.

Could be at each calibration that they do a periodic maintenance on the optics that could degrade (surface contamination buildup, for example) if left for two years.

It may well be fully acceptable to go two years (and in my opinion, there is reasonable likelihood that is the case). But I feel duty bound to share the risk potential.

There are two opposing forces in any circumstance where the interval is adjusted:

1. COST: if the interval is too short, too much is spent on unneeded calibrations.

2. RISK/COST: if the interval is too long, there is ALWAYS statistical increase in the risk of undesired out-of-tolerance. I added COST to the second statement, because that is an often overlooked potential impact of too long an interval. The undesired out-of-tolerance may result in "inaccurate" decision based on an out-of-tolerance reading.

Both of the above must always be factored into setting of initial intervals, and adjusting them. How the instrument is used and stored impacts how well it retains tolerance. How it is used also determines it's potential impact.
Reply With Quote
  #4  
Old 1st November 2005, 09:55 AM
Graeme's Avatar
Graeme Graeme is offline
$ Contributor

Registration Date: Sep 2000
Location: Lilburn, GA, USA
Age: 60
 
Posts: 401
Thanks Given to Others: 0
Thanked 8 Times in 7 Posts
Karma Power: 63
Karma: 894
Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.
Send a message via AIM to Graeme Send a message via MSN to Graeme Send a message via Yahoo to Graeme
Default

To add to the excellent information from Jerry and JSW05 --

Another part of the risk/cost consideration on extending intervals is this: if one of the meters is found to be out of tolerance, you are required to evaluate the impact of the OOT condition on the product and, if necessary, recall all items that meter was used on back to the last time it was known to be in tolerance. That can be a very substantial cost.

An "insurance policy" that can help with this is the practice of making regular measurements of a product item (or other appropriate stable artifact) that is reserved for use as a check standard. At specified intervals (depending on your use of the meter) the check standard is measured in exactly the same way as routine product measurements, and the results are plotted on a control chart. Examination of the control chart will indicate potential problems and action may be taken much sooner than waiting for the next calibration results. The "last known good date" is immediately before the control chart starts to indicate a problem. Naturally, a precondition for this is that the meaurment process is in a state of statistical control.

While agreeing that the manufacturer's recommended interval (if any) is only a guideline, if you are in a regulated industry there are other considerations as well. In many cases the regulating agency mandates that the manufacturer-suggested interval shall be used unless other conditions are met. For the agency I am most familiar with (FAA), any change of the interval must be justified on the basis of statistical analysis of actual data and a defined end-of-interval reliability goal. That means, using Jerry's calculations as an example, you could only justify extending the interval to 14 months. (It also means that any of the reactive methods listed in RP-1 can not be used.)
__________________
Graeme C. Payne
ASQ Sr. Member; CQE; CCT

"Does it matter if the measurement result is wrong?
If it does, then calibrate the instrument.
If it doesn't matter, they why are you making the measurement?"
(P. G. Stein, 2000)
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Calibration of a frequency measurement device/frequency source Ruebenn ISO 17025 - Calibration, Measurement Gages and Test Laboratories 2 25th March 2009 01:48 AM
What is the recommended calibration frequency for gage blocks? mcowan Calibration Frequency (Interval) 2 28th January 2009 12:50 PM
Extending Surface Plate Calibration Frequency mhirand20 ISO 17025 - Calibration, Measurement Gages and Test Laboratories 3 23rd January 2009 05:03 PM
Calibration - Manufacturer's Calibration Certificate Date edelsal ISO 9001 - Quality Management Systems Standard 8 11th January 2009 10:36 PM
Plug Gage Calibration - Manufacturer's Certification vs. Calibration Laboratory Dawn ISO 17025 - Calibration, Measurement Gages and Test Laboratories 2 24th January 2000 07:25 PM



The time now is 03:41 AM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts