FDA Validation - Does one have to conduct validation for Computer Servers?
Does one have to conduct validation for Computer Servers to meet FDA regualtion? If yes, anyone have any ideas how we should go about it?
|

9th March 2006, 11:26 AM
|
 |
Consultant
Registration Date: May 2005
Location: Maryland
Age: 44
|
|
Posts: 472
Thanks Given to Others: 36
Thanked 42 Times in 28 Posts
Karma Power: 43
|
|
Quote:
|
Originally Posted by SilverHawk
Does one have to conduct validation for Computer Servers to meet FDA regualtion? If yes, anyone have any ideas how we should go about it?
|
In short the answer is yes! I would review 21CFR11 to get a general understanding of the issues. I would highly recommend you find a consultant who is familar with the necessary protocols if you find after reading Part 11 that you do not have the expertise.
__________________
Steven Walfish
When in doubt, ask your company statistician!
|

10th March 2006, 07:29 PM
|
|
Courtesy Access
Registration Date: Jan 2004
Location: Malaysia/Kuala Lumpur
|
|
Posts: 308
Thanks Given to Others: 10
Thanked 26 Times in 17 Posts
Karma Power: 34
|
|
Noted with thanks./
|

7th August 2007, 11:46 AM
|
|
wehttam
Registration Date: Feb 2005
Location: IRELAND
Age: 43
|
|
Posts: 20
Thanks Given to Others: 0
Thanked 10 Times in 4 Posts
Karma Power: 21 Karma: 52 
|
|
Re: FDA Validation - Does one have to conduct validation for Computer Servers?
You only have to validate software for its intended purpose.
|

8th August 2007, 04:35 AM
|
 |
Super Moderator
Registration Date: Dec 2005
Location: India
Age: 46
|
|
Posts: 2,645
Thanks Given to Others: 674
Thanked 802 Times in 579 Posts
Karma Power: 176
|
|
Re: FDA Validation - Does one have to conduct validation for Computer Servers?
Quote:
Originally Posted by WEHTTAM
You only have to validate software for its intended purpose.
|
There is lot of information on this subject. I have attached a weblink from the FDA website - http://google2.fda.gov/search?client...are+validation
__________________
I have no special talents. I am only passionately curious - Albert Einstein
|

13th August 2007, 04:18 PM
|
 |
Involved in Discussions
Registration Date: Nov 2003
Location: Grand Rapids, MI
|
|
Posts: 32
Thanks Given to Others: 2
Thanked 82 Times in 15 Posts
Karma Power: 35
|
|
Re: FDA Validation - Does one have to conduct validation for Computer Servers?
You validate software.
You qualify the hardware. I would suggest that you check out the GAMP organization at http://www.ispe.org.
They have all kinds of templates suitable for validation in a pharma or medical device environment.
If you just need hardware qualification templates, I might have some available.
|
|
Thank You to pbojsen for your informative Post and/or Attachment!
|
|

13th August 2007, 05:31 PM
|
 |
Consultant
Registration Date: May 2005
Location: Maryland
Age: 44
|
|
Posts: 472
Thanks Given to Others: 36
Thanked 42 Times in 28 Posts
Karma Power: 43
|
|
Re: FDA Validation - Does one have to conduct validation for Computer Servers?
Quote:
Originally Posted by pbojsen
You validate software.
You qualify the hardware. I would suggest that you check out the GAMP organization at http://www.ispe.org.
They have all kinds of templates suitable for validation in a pharma or medical device environment.
If you just need hardware qualification templates, I might have some available.
|
Thank you, you used the correct term. After reading your reply it occured to me that I should have said YES to qualify
__________________
Steven Walfish
When in doubt, ask your company statistician!
|

17th November 2007, 01:28 PM
|
|
E-Mails Invalid or Rejected
Registration Date: Nov 2007
Location: Bangalore, India
|
|
Posts: 2
Thanks Given to Others: 1
Thanked 0 Times in 0 Posts
Karma Power: 0 Karma: 10 
|
|
Re: FDA Validation - Does one have to conduct validation for Computer Servers?
Regulations require Computerized Systems to be validated when they support
GxP Processes or hold GxP data which essentially means that the data center, DR Center, Datacenter support systems like UPS, HVAC etc, Servers, Network and applications need to be qualified. Typically one needs to follow IQ/OQ/PQ Methodology to qualify the above systems.
Thanks.
|
Lower Navigation Bar
|
|
|
|
Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
|
|
|
| Thread Tools |
Search this Thread |
|
|
|
| Display Modes |
Rate Thread Content |
Linear Mode
|
|
Posting Settings
|
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
HTML code is Off
|
|
|
|
|