The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


Related Topic Tags
calibration (general topics), measurement and monitoring and test devices, thermometers and thermostats, temperature (general), temperature gages and probes
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 15th June 2006, 03:27 PM
Garner Davis - 2011 Garner Davis - 2011 is offline
Email Address Invalid or Rejected by Recipient System

 
Registration Date: Apr 2005
Location: USA
 
Posts: 2
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 0
Karma: 10
Garner Davis - 2011 has less than 100 Karma points so far.
Please Help! Calibration of Thermometers and Temperature Sensitive Devices

Hello, all.

I've searched this forum for some ideas on calibrating thermometers, and I've found some good information. I used to work for a large company, and our system was to purchase the expensive NIST-traceable thermometer and use it to check the accuracy of other thermometers, as well as ovens, water baths, etc.

Now I work for a much smaller company, and spending that type of money is not an option. I've seen an idea in one thread that seemed to make sense for thermometers, which involved placing all the devices in the same environment for a given amount of time, taking a reading, and as long as they fell within acceptable tolerances for the process, everything was OK. Repeat for different temperatures as necessary.

Now, I want to extend that idea to include the devices such as the water baths, ovens, etc. If I select one of the thermometers to use as a "golden" standard, would it be acceptable to use it to check the accuracy of the other devices. My personal thought is that this should be OK, as long as it's defined within the quality system, but I welcome feedback from the group if there's something I might have missed.

Thanks for your time;

Garner Davis

Sponsored Links
  #2  
Old 15th June 2006, 03:44 PM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Super Moderator

 
Registration Date: Jun 2002
Location: Lawn Guyland
Age: 62
 
Posts: 3,547
Thanks Given to Others: 71
Thanked 717 Times in 442 Posts
Karma Power: 413
Karma: 7090
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
Re: Calibration of thermometers and temperature-sensitive devices....

Quote:
In Reply to Parent Post by Garner Davis

Hello, all.

I've searched this forum for some ideas on calibrating thermometers, and I've found some good information. I used to work for a large company, and our system was to purchase the expensive NIST-traceable thermometer and use it to check the accuracy of other thermometers, as well as ovens, water baths, etc.

Now I work for a much smaller company, and spending that type of money is not an option. I've seen an idea in one thread that seemed to make sense for thermometers, which involved placing all the devices in the same environment for a given amount of time, taking a reading, and as long as they fell within acceptable tolerances for the process, everything was OK. Repeat for different temperatures as necessary.

Now, I want to extend that idea to include the devices such as the water baths, ovens, etc. If I select one of the thermometers to use as a "golden" standard, would it be acceptable to use it to check the accuracy of the other devices. My personal thought is that this should be OK, as long as it's defined within the quality system, but I welcome feedback from the group if there's something I might have missed.

Thanks for your time;

Garner Davis
This is not an acceptable way to calibrate your devices. You should get one device calibrated with traceability to NIST and use it exclusively for calibrating all the others.
__________________
Al
Sponsored Links

  #3  
Old 15th June 2006, 03:52 PM
Craig H. Craig H. is offline
Inactive Registered Visitor

 
Registration Date: Jan 2002
Location: Mars
 
Posts: 2,054
Thanks Given to Others: 323
Thanked 310 Times in 239 Posts
Karma Power: 262
Karma: 5264
Craig H. is appreciated, and has over 1700 Karma points.
Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.Craig H. is appreciated, and has over 1700 Karma points.
Re: Calibration of thermometers and temperature-sensitive devices....

Quote:
In Reply to Parent Post by Al Rosen

This is not an acceptable way to calibrate your devices. You should get one device calibrated with traceability to NIST and use it exclusively for calibrating all the others.

Garner,

I agree with Al. Would it be possible to send one of your existing thermometers off to be checked? That could become your reference thermometer.
  #4  
Old 15th June 2006, 04:44 PM
Garner Davis - 2011 Garner Davis - 2011 is offline
Email Address Invalid or Rejected by Recipient System

 
Registration Date: Apr 2005
Location: USA
 
Posts: 2
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 0
Karma: 10
Garner Davis - 2011 has less than 100 Karma points so far.
Re: Calibration of thermometers and temperature-sensitive devices

Thank you for the responses; I'll investigate what the cost would be to do that.

I remember that the cost of a NIST-traceable thermometer from Fisher Scientific and the like seemed to run around $1K; if we're looking at about the same cost for sending one off, then I'm not sure that's feasible for us at this juncture.

Thanks again for the responses.

Garner
  #5  
Old 15th June 2006, 04:47 PM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Super Moderator

 
Registration Date: Jun 2002
Location: Lawn Guyland
Age: 62
 
Posts: 3,547
Thanks Given to Others: 71
Thanked 717 Times in 442 Posts
Karma Power: 413
Karma: 7090
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
Re: Calibration of thermometers and temperature-sensitive devices

Quote:
In Reply to Parent Post by Garner Davis

Thank you for the responses; I'll investigate what the cost would be to do that.

I remember that the cost of a NIST-traceable thermometer from Fisher Scientific and the like seemed to run around $1K; if we're looking at about the same cost for sending one off, then I'm not sure that's feasible for us at this juncture.

Thanks again for the responses.

Garner
The cost to calibrate it is not anywhere near that much.
__________________
Al
  #6  
Old 15th June 2006, 05:01 PM
pilchard's Avatar
pilchard pilchard is offline
No Badges

 
Registration Date: Oct 2003
Location: Lafayette Louisiana
 
Posts: 80
Thanks Given to Others: 0
Thanked 5 Times in 4 Posts
Karma Power: 47
Karma: 237
pilchard is appreciated, and has over 200 Karma points.pilchard is appreciated, and has over 200 Karma points.pilchard is appreciated, and has over 200 Karma points.
Money Re: Calibration of thermometers and temperature-sensitive devices

Ice is 32°F or 0°C

A properly prepared ice-point bath can be used for thermometer calibration. Poor preparation and use of the bath will result in errors. With care, one can use common laboratory beakers and stirring equipment to produce an ice-point bath that is within ± 0.01 °C (0.02 °F) of the true ice point.

Procedure
The following table details preparing and using ice-point baths.

Preparing and Using Ice-point Baths
Step Action
1 The water must be uniformly cooled to and maintained at the ice-point temperature.
¨ This can be done only by using slush made of ice shavings or crushed ice particles.
¨ Constantly stir the ice and water slush to keep it distributed uniformly.
¨ The slush must contain only enough water to wet the ice particles and allow the mixture to be slightly fluid.
¨ Water will have to be removed periodically and ice added as the bath is used.
¨ For best accuracy, use distilled water and clean clear ice.

2 Do not allow clumps of ice or water pockets to form.
¨ When ice particles clump together, the water between them freezes and the temperature drops.
¨ If a pocket of water forms, the temperature in it will rise.
¨ A water and ice cube mixture will not cool the water between the cubes sufficiently.

3 Do not allow a thermometer or temperature sensor in the ice-point bath to touch any part of the bath or any other thermometer or sensor. At least 25 mm (1 in.) of slush shall always be below the lowest point of any thermometer.

4 Suspend each thermometer from above and hold at the proper immersion depth. It will be easier to read the liquid column of a glass thermometer if the container is almost full.

5 Constantly stir the bath in order to uniformly distribute the ice particles and water.

6 Allow the thermometer or sensor to stabilize at the temperature of the bath.

7 ¨ Read the thermometer, wait for about a minute, and read again.
¨ Record only those readings where the indicated temperature is repeated within the desired accuracy.

NOTE: You may stop stirring momentarily while reading a glass thermometer.
8 Record the indicated value.

9 Find the error by subtracting the ice point (i.e., 0 °C [32 °F]) from the indicated value.


The correction factor has the same numerical value as the error but with opposite sign.



EXAMPLE: Let the true temperature be 0 ° and the indicated temperature be -0.40 °; the thermometer reads low with an error of -0.40 °; the correction factor to apply at that point is + 0.40 °.
__________________
A parasitic government and populous will kill its host - Just Watch
Thank You to pilchard for your informative Post and/or Attachment!
  #7  
Old 15th June 2006, 05:06 PM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Super Moderator

 
Registration Date: Jun 2002
Location: Lawn Guyland
Age: 62
 
Posts: 3,547
Thanks Given to Others: 71
Thanked 717 Times in 442 Posts
Karma Power: 413
Karma: 7090
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
Re: Calibration of thermometers and temperature-sensitive devices

Now you have a device calibrated at one point. What about the rest of the range?
__________________
Al
  #8  
Old 15th June 2006, 05:16 PM
pilchard's Avatar
pilchard pilchard is offline
No Badges

 
Registration Date: Oct 2003
Location: Lafayette Louisiana
 
Posts: 80
Thanks Given to Others: 0
Thanked 5 Times in 4 Posts
Karma Power: 47
Karma: 237
pilchard is appreciated, and has over 200 Karma points.pilchard is appreciated, and has over 200 Karma points.pilchard is appreciated, and has over 200 Karma points.
Truce Re: Calibration of thermometers and temperature-sensitive devices

Quote:
In Reply to Parent Post by Al Rosen

Now you have a device calibrated at one point. What about the rest of the range?
You buy a NIST Thermometer!

You use a water boiling point calculate at you altitude and barometric pressure and use the same procedure.

I would rather use Gumbo at least if it dont work you can feed the auditor!
__________________
A parasitic government and populous will kill its host - Just Watch

Last edited by pilchard; 15th June 2006 at 05:22 PM.
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Temperature indication In Vitro Medical Devices USA Paul22 US Medical Devices (21 CFR part 820) 2 23rd December 2005 03:24 AM
Temperature and Humidity Device Calibration - Came back as Out Of Calibration pancho-2007 ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 8 4th March 2005 01:45 PM
Level of accuracy required for the calibration of glass mercury thermometers Dave T ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 6 19th March 2000 04:19 PM
Calibration Supplier Selection - Calibration & Temperature & Work Instructions Jerry Eldred ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 1 22nd December 1999 12:11 AM