3rd May 2007, 06:26 AM
|
 |
Involved in Discussions
Registration Date: Aug 2000
Location: usa
Age: 61
|
|
Posts: 279
Thanks Given to Others: 217
Thanked 16 Times in 15 Posts
Karma Power: 80 Karma: 95 
|
|
Late calibration (Calibration Past Due Date) - What do I do now?
I forgot to send in a piece of equipment that needed it's yearly calibration. I sent it in a month after it was due. Should I write myself a corrective action? Any thoughts would be appreciated.
|

3rd May 2007, 06:31 AM
|
 |
Get Involved!!!
Registration Date: Dec 2004
Location: UK, Hampshire
|
|
Posts: 550
Thanks Given to Others: 124
Thanked 110 Times in 82 Posts
Karma Power: 91
|
|
|
Re: What do I do now?
I think it would be worth it, to avoid it happening again. You obviously don't have a "Mistake Proofing" type system in place that ensures that all equipment is calibrated as and when requried without fail.
You need to look into how you record calibration and how an effective reminder system can be implemented. Was the equipment used outside of it's calibrated state? If so that could lead to non-confirmance of some kind, either waste of OK products not being detected correctly or even worse, non conformaning product going out of the door! Definate need for corrective action to fix the problem, IMO.
__________________
All things are subject to interpretation whichever interpretation prevails at a given time is a function of power and not truth. - Friedrich Nietzsche
|
|
Thanks to Manix for your informative Post and/or Attachment!
|
|

3rd May 2007, 06:52 AM
|
 |
Involved in Discussions
Registration Date: Oct 2006
Location: São Paulo, Brazil
Age: 45
|
|
Posts: 665
Thanks Given to Others: 283
Thanked 298 Times in 205 Posts
Karma Power: 97
|
|
|
Re: What do I do now?
Quote:
In Reply to Parent Post by dbulak
I forgot to send in a piece of equipment that needed it's yearly calibration. I sent it in a month after it was due. Should I write myself a corrective action? Any thoughts would be appreciated.
|
7.6 States "In addition, the organization shall assess and record the validity of the previous measuring results when the equipment is found not to conform to requirements. The organization shall take appropriate action on the equipment and any product affected."
Therefore, IMO, yes you should write up a Non conformity report and take corrective action. You also should "back track" the measurements made with this instrument to see if you approved a "bad" product... Here at our factory, each machine (CNC lathes and mechanical lathes), has an inspection form which has the periodic measurement checks and also the instrument number that was utilized. This way I can easily check which parts where measured with which instrument.
__________________
Nothing is difficult, you just didn´t understand it yet...
|

3rd May 2007, 08:38 AM
|
 |
Forum Moderator
Registration Date: Feb 2005
Location: Michigan, USA
|
|
Posts: 7,035
Thanks Given to Others: 4,022
Thanked 2,875 Times in 1,969 Posts
Karma Power: 776
|
|
Re: Late calibration - What do I do now?
Quote:
In Reply to Parent Post by dbulak
I forgot to send in a piece of equipment that needed it's yearly calibration. I sent it in a month after it was due. Should I write myself a corrective action? Any thoughts would be appreciated.
|
Not at all! Firstly, why bother? One item isn't a systemmic issue and now you've done it, you're unlikely to forget again...........(unless it's the only item you have to send out)
What you should do, is allow yourself (in your calibration procedure) to write an extension to the calibration anniversary. You see, most items get calibrated on a time base, like annually, but don't require calibration on the same basis. Take a look at the data from previous calibrations and see how close the equipment's results are compared to the specification for the equipment and the product/process parameters (it's used to measure).
Once you know this, you can go ahead and write the 'extention' statement with the supporting data.
BTW - effective calibration systems allow recall periods. based on data, not timing of the recall.......
Hope this helps.
__________________
'Cause you know sometimes words have two meanings.....
|
|
Thanks to AndyN for your informative Post and/or Attachment!
|
|

3rd May 2007, 08:42 AM
|
 |
Quality Manager
Registration Date: Sep 2003
Location: Illinois
|
|
Posts: 10,453
Thanks Given to Others: 460
Thanked 2,636 Times in 1,722 Posts
Karma Power: 1124
|
|
|
Re: What do I do now?
Quote:
In Reply to Parent Post by tedschmitt
7.6 States "In addition, the organization shall assess and record the validity of the previous measuring results when the equipment is found not to conform to requirements. The organization shall take appropriate action on the equipment and any product affected."
Therefore, IMO, yes you should write up a Non conformity report and take corrective action. You also should "back track" the measurements made with this instrument to see if you approved a "bad" product... Here at our factory, each machine (CNC lathes and mechanical lathes), has an inspection form which has the periodic measurement checks and also the instrument number that was utilized. This way I can easily check which parts where measured with which instrument.
|
Let's be realistic here. My experience is many instruments are still in calibration when sent to the shop or lab for the required recalibration - the whole point is to set up the schedule so the equipment does NOT go out of calibration triggering the backtracking described.
This is an opportunity to revisit your calibration schedules and adjust accordingly. Do not look at it as a major nonconformance unless the equipment WAS out of calibration and triggered the backtracking and then the nonconformance should address the matter of ALL calibration schedules and the process to assure the schedule is followed.
__________________
"Few minds wear out; more rust out"
Inscribed over the entrance of Louis Pasteur School, Chicago
Christian Nestell Bovee (1820-1904) in Thoughts, Feelings and Fancies, 1857
|
|
Thanks to Wes Bucey for your informative Post and/or Attachment!
|
|

3rd May 2007, 08:44 AM
|
|
Tater Salad
Registration Date: Dec 2006
Location: CT
|
|
Posts: 379
Thanks Given to Others: 89
Thanked 159 Times in 108 Posts
Karma Power: 65
|
|
|
Re: Late calibration - What do I do now?
As Manix suggested I do think it would be appropriate, mainly because your process should not allow you to "forget" to calibrate an item. One thing I'm interested in however, if you did indeed forget to calibrate by the due date, was the instrument still in use?
Your system should not only ensure calibrations are done in a timely manner, but also that instruments which are due for calibration are immediately recalled from production or otherwise identified as unusable.
It shouldn't be too difficult to make this process more mistake proof, but anyhow, yes CA is appropriate with a focus on correcting your process so that items can not be simply "forgotten" and items overdue for calibration are removed from use.
|
|
Thanks to Benjamin28 for your informative Post and/or Attachment!
|
|

3rd May 2007, 08:46 AM
|
 |
Involved in Discussions
Registration Date: Oct 2006
Location: São Paulo, Brazil
Age: 45
|
|
Posts: 665
Thanks Given to Others: 283
Thanked 298 Times in 205 Posts
Karma Power: 97
|
|
|
Re: What do I do now?
Quote:
In Reply to Parent Post by Wes Bucey
Let's be realistic here. My experience is many instruments are still in calibration when sent to the shop or lab for the required recalibration - the whole point is to set up the schedule so the equipment does NOT go out of calibration triggering the backtracking described.
This is an opportunity to revisit your calibration schedules and adjust accordingly. Do not look at it as a major nonconformance unless the equipment WAS out of calibration and triggered the backtracking and then the nonconformance should address the matter of ALL calibration schedules and the process to assure the schedule is followed.
|
Wes,
I agree 100% with you... but, my answer was based on the info that was given that the instrument was sent to calibration one month after the calibration was due...
__________________
Nothing is difficult, you just didn´t understand it yet...
|

3rd May 2007, 08:49 AM
|
|
Tater Salad
Registration Date: Dec 2006
Location: CT
|
|
Posts: 379
Thanks Given to Others: 89
Thanked 159 Times in 108 Posts
Karma Power: 65
|
|
|
Re: Late calibration - What do I do now?
Hmm...
I don't agree, I don't see this as a question of whether the instrument is or was in tolerance, or whether the calibration frequency is appropriate, or opening the question of adjusting frequencies.
I see this as a human error which needs to be accounted for in the calibration process so that the organization can consistently meet the planned recall and calibration dates it assigns to its instruments.
|
|
Thanks to Benjamin28 for your informative Post and/or Attachment!
|
|
Lower Navigation Bar
|
|
|
Do you find this discussion thread helpful and informational?
|
Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
|
|
|
| Thread Tools |
Search this Thread |
|
|
|
| Display Modes |
Rate Thread Content |
Linear Mode
|
|
Forum Posting Settings
|
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
HTML code is Off
|
|
|
|
|