Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 28th May 2007, 10:11 AM
Wayne's Avatar
Wayne Wayne is offline
Gage Crib Worldwide

Registration Date: Aug 2005
Location: Grand Rapids, MI USA
Age: 58
 
Posts: 170
Thanks Given to Others: 14
Thanked 65 Times in 36 Posts
Karma Power: 46
Karma: 1298
Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.Wayne is appreciated, and has over 1100 Karma points.
Send a message via Skype™ to Wayne
Caution Never put an Expiration (Due) Date on a Calibration Certificate

I have some fundamental problems with the concept of a certification having an Expiration Data, or a Recalibration Date.

My problems begin with the basic fact that a certification is only valid at the moment the instrument is being checked. By the time the certification has been typed the instrument may already be out of calibration. Some may call this an absurd statement, but I contend that it is absolutely true.

Let me give some examples:

1. A Lab Tech finishes his calibration work on an instrument, leaves the tool on his bench and goes to his computer to transfer his measurement data to the certificate. While he is away from his workbench a co-worker accidentally knocks the instrument onto the floor. For reasons know only to the co-worker, he picks-up the instrument and replaces it on the workbench and does not report the incident. The fall, in this situation has put the instrument out of wack and invalidated the calibration as it was being typed.

2. A properly calibrated instrument has been packaged for shipping back to the owner of the tool. The shipment is by overnight courier. During the return trip the instrument passes through several extreme temperature situations: Heated by sun baked delivery truck; Frozen in the belly of a high flying airplane; Heated again in another truck turned oven by the sun. The result is that critical components of the instrument, which are held in tension and locked in place, are loosened by thermal expansion and contraction allowing the tool is no longer at the certified set size.

3. Upon arrival at the owner’s factory, the Receiving Clerk, in the process of unpackaging the instrument, handles it roughly by whatever untoward method you can imagine. He never reports the possible damage to the tool because it looks just fine. Regardless, the instrument has been rendered inaccurate by his actions.

4. A properly calibrated thread plug gage is placed into service with the instructions to the user to measure every 100th part. The operator, being a good employee in his eyes, figures that if one part measured in a hundred is good, every part measured is lots better. A week later the poor thread plug gage is, while unnoticeable to visual inspection worn to a nub. It is returned to the crib and stored for future use even though it has been worn beyond its limits.

Let me say that these situations may or may not cause a tool to fall out of calibration, and I would hope that many instruments can handle some abuse without going out of wack. Let me also say that the opportunity for an instrument’s settings to be silently altered are numerous and beyond total control. It is for these reasons that placing an Expiration Date on a certification, in my opinion, should be avoided like the plague.

Yes, we all know that an Expiration Date is for reference only. Yes, we all know that all instruments should be validated for function frequently. Or do we? To many not as deeply involved in the calibration of instruments and not familiar with the fragility of measurement instruments, an Expiration Date gives the false impression that all will be fine until that date is met. It might even be argued that an Expiration Date is a guarantee of performance up to that date, which is something that no calibration laboratory would be willing to assure. For these reasons I contend that no certification should ever have an Expiration Date. A Recalibration Date is nearly as bad. It implies the same thing as an Expiration Date with the subtitle differences being lost on the masses.

My preference would be to have no time based calibration cycle stated on a certification. Calibration cycles are best determined based on usage. But, customers have become trained to use time based systems, and expectant of such data being found on the certification. Calibration Laboratories, wanting to meet customer requirements, to keep the customer happy, have complied and placed such specious data on he certification.

The solution, on the certification any recalibration date should be identified as ‘suggested’. Further caveats should also be included in the fine print as to proper determination of calibration cycle or some such other disclaimer against the absolute value of any given suggested, time based, recalibration date.
Reply With Quote
Thank You to Wayne for your informative Post and/or Attachment!

Sponsored Links
  #2  
Old 28th May 2007, 10:28 AM
silentrunning's Avatar
silentrunning silentrunning is offline
Lab Technician

Registration Date: Apr 2007
Location: Sarasota, Florida
 
Posts: 135
Thanks Given to Others: 112
Thanked 33 Times in 26 Posts
Karma Power: 16
Karma: 182
silentrunning is appreciated, and has over 100 Karma points.silentrunning is appreciated, and has over 100 Karma points.
Default Re: Never put an Expiration (Due) Date on a Calibration Certificate

I agree with everything you have said here. Since there is no "perfect" way to do calibration at a reasonable cost, I use it as a fence. Once a device is proven to be in calibration everthing checked prior to that calibration is considered acceptable. As you stated, a gage can go out at any point during the inspection process. It is not feasable to calibrate a gage after each measurement so we keep track of what product is checked by which tool. At the end of a production run the inspection tools are rechecked for accuracy. Not recalibrated! We send our inspection tools out every 6 months for outside calibration to verify our in house measurements. When they are proven accurate we approve the job. The extra 10 or 15 minutes this takes for each job has paid huge dividends over the last few years.

Doug
__________________
W4JDZ
Reply With Quote
Thank You to silentrunning for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 28th May 2007, 12:40 PM
Tim Folkerts's Avatar
Tim Folkerts Tim Folkerts is offline
Forum Moderator

Registration Date: Sep 2003
Location: Kansas, USA
Age: 46
 
Posts: 900
Thanks Given to Others: 27
Thanked 249 Times in 150 Posts
Karma Power: 123
Karma: 3974
Tim Folkerts is appreciated, and has over 1700 Karma points.
Tim Folkerts is appreciated, and has over 1700 Karma points.Tim Folkerts is appreciated, and has over 1700 Karma points.Tim Folkerts is appreciated, and has over 1700 Karma points.Tim Folkerts is appreciated, and has over 1700 Karma points.Tim Folkerts is appreciated, and has over 1700 Karma points.Tim Folkerts is appreciated, and has over 1700 Karma points.Tim Folkerts is appreciated, and has over 1700 Karma points.Tim Folkerts is appreciated, and has over 1700 Karma points.
Default Re: Never put an Expiration (Due) Date on a Calibration Certificate

I would tend to treat an expiration date as a maximum time, not necessarily a guarantee.

For example, the egg salad in my fridge has an expiration date a week from now. But if I leave it out of the fridge overnight, I'd throw it out - even if it is still a week to the stated expiration.

So an instrument - if shipped and stored and used as expected - should be good up until the expiration data. Eventually the calibration will fail - due to normal use, corrosion, evaporation, etc. The due data (IMHO) is an effort to estimate these sorts of expectations.

If not shipped and stored and used as expected, then it could well go out of calibration earlier. As you say, it is hard (or impossible) to know if it has been handled appropriately. That is where in-house checks help to ensure it is still operating as expected.


Tim F
__________________
To wonder is to begin to understand.
Reply With Quote
Thank You to Tim Folkerts for your informative Post and/or Attachment!
  #4  
Old 28th May 2007, 02:20 PM
Jerry Eldred's Avatar
Jerry Eldred Jerry Eldred is offline
Forum Moderator

Registration Date: Dec 1999
Location: Texas
 
Posts: 411
Thanks Given to Others: 0
Thanked 118 Times in 70 Posts
Karma Power: 89
Karma: 2042
Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.Jerry Eldred is appreciated, and has over 1700 Karma points.
Default Re: Never put an Expiration (Due) Date on a Calibration Certificate

Let me give a reply that sounds like I am talking out both sides of my mouth.

On the one hand, a Certification and due date are not in any way a guarantee of how long the instrument will stay within spec. When an instrument is calibrated/certified, it is returned to the owner, and if anything happens between the cal date and due date, the lab is not liable for it's remaining within tolerance.

On the other hand, the due date (or let me shift gears and use the term "interval" for the remainder of my reply) is based on some valid reasoning. The calibration interval is the amount of time to a defined statistical confidence level that an instrument can be expected to remain within tolerance. This means that the assignement of a due date is based on a predicted interval of time that the instrument will "probably" remain in tolerance. 95% is a common confidence, based on statistical probability.

There are other ways to assign or adjust intervals (but I won't go into all that now). So the assignment of a due date on a certificate means who ever assigned it is stating (if using the statistical method), that the unit will have a 95% probability (using my example above) of remaining in tolerance until that date.

So there is valid thought that goes into assigning the due date.

I would have to go on record that a certification (with some exceptions) must have a due date on it. This is a part of what a certificate of calibration is - a prediction of how long an instrument will probably remain in tolerance.
Reply With Quote
Thank You to Jerry Eldred for your informative Post and/or Attachment!
  #5  
Old 28th May 2007, 02:44 PM
Jim Wynne's Avatar
Jim Wynne Jim Wynne is offline
Courtesy Access

Registration Date: Jan 2005
Location: Southeast Wisconsin
Age: 57
 
Posts: 9,211
Thanks Given to Others: 755
Thanked 2,293 Times in 1,547 Posts
Karma Power: 611
Karma: 20380
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Default Re: Never put an Expiration (Due) Date on a Calibration Certificate

Quote:
Originally Posted by Jerry Eldred View Post

I would have to go on record that a certification (with some exceptions) must have a due date on it. This is a part of what a certificate of calibration is - a prediction of how long an instrument will probably remain in tolerance.
I agree, and thanks to Wayne for raising an interesting topic.

We need to not forget that part of the information we receive from calibration is the "as found" condition, and we should be using it wisely. If, for example, we find that 6-inch calipers in similar use consistently need adjustment when being calibrated, we need to (A) shorten the interval, and (B) investigate the situation. Are the calipers being abused? Are they worn such that maintaining set points is impossible?

The problem, as I see it, is one of education; people need to know that the due date on the label represents a maximum and not an absolute value. This seems obvious, and I think most people understand that there is risk in assuming that devices are properly calibrated when the state of calibration isn't intuitive. But you can't get people to use measurement equipment prudently by masking significant information.
__________________
Some men are born mediocre, some men achieve mediocrity, and some men have mediocrity thrust upon them.-- Joseph Heller
Reply With Quote
Thanks to Jim Wynne for your informative Post and/or Attachment!
  #6  
Old 28th May 2007, 05:52 PM
Wes Bucey's Avatar
Wes Bucey Wes Bucey is offline
Quality Manager

Registration Date: Sep 2003
Location: Illinois
 
Posts: 7,534
Thanks Given to Others: 182
Thanked 1,127 Times in 762 Posts
Karma Power: 397
Karma: 11090
Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.
Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.Wes Bucey is appreciated, and has over 1700 Karma points.
Default Re: Never put an Expiration (Due) Date on a Calibration Certificate

For my part, I had only the date of calibration/certification put on the label or tag, never a due date. To my mind, the interval between calibration/certification dates doesn't even start until the gage/instrument is put into service. Some gages might lay in a drawer in a temperature-controlled lab for 6 months before going into use. I kept track of usage in a database keyed to serial numbers of each gage or instrument. If the gage got excessive use, it got checked more frequently in-house to confirm it was still within tolerance.

I have had some gages go out of tolerance very quickly. I heard an apocryphal story about employees using a 2-inch micrometer as a nut cracker!
__________________
"Few minds wear out; more rust out"
Inscribed over the entrance of Louis Pasteur School, Chicago
Christian Nestell Bovee (1820-1904) in Thoughts, Feelings and Fancies, 1857
Reply With Quote
  #7  
Old 28th May 2007, 06:10 PM
Zuggy's Avatar
Zuggy Zuggy is offline
Registered Visitor

Registration Date: Feb 2005
Location: Canada
 
Posts: 35
Thanks Given to Others: 6
Thanked 3 Times in 3 Posts
Karma Power: 20
Karma: 25
Zuggy has less than 100 Karma points so far.
Default Re: Never put an Expiration (Due) Date on a Calibration Certificate

Quote:
Originally Posted by Jerry Eldred View Post

Let me give a reply that sounds like I am talking out both sides of my mouth.

On the one hand, a Certification and due date are not in any way a guarantee of how long the instrument will stay within spec. When an instrument is calibrated/certified, it is returned to the owner, and if anything happens between the cal date and due date, the lab is not liable for it's remaining within tolerance.

On the other hand, the due date (or let me shift gears and use the term "interval" for the remainder of my reply) is based on some valid reasoning. The calibration interval is the amount of time to a defined statistical confidence level that an instrument can be expected to remain within tolerance. This means that the assignment of a due date is based on a predicted interval of time that the instrument will "probably" remain in tolerance. 95% is a common confidence, based on statistical probability.

There are other ways to assign or adjust intervals (but I won't go into all that now). So the assignment of a due date on a certificate means who ever assigned it is stating (if using the statistical method), that the unit will have a 95% probability (using my example above) of remaining in tolerance until that date.

So there is valid thought that goes into assigning the due date.

I would have to go on record that a certification (with some exceptions) must have a due date on it. This is a part of what a certificate of calibration is - a prediction of how long an instrument will probably remain in tolerance.
I would also agree to this statement. Use to work for a Machining/Millwright company. Our standard practice was to check tools such as I/S and O/S mic's before use. Furthermore, we (for the most part) had two sets of standards for the QA system. I agree that a calibration sticker or certificate is no promise that the tool is within spec, however it does prove to the customer that your QA program is important enough to ensure best practice.


JMT Tim
Reply With Quote
  #8  
Old 28th May 2007, 09:20 PM
Marc's Avatar
Marc Marc is online now
Your Elsmar Cove Host

Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
 
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,566 Times in 1,018 Posts
Blog Entries: 4
Karma Power: 605
Karma: 11559
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.Marc is appreciated, and has over 1700 Karma points.
Send a message via AIM to Marc Send a message via Skype™ to Marc
I Say... What about a 'Due Date' on calibration stickers?

A 'Due Date' on a calibration certificate is one thing. What about a 'Due Date' on calibration stickers which most companies put on individual measurement instruments?
__________________
A Search is a terrible thing to waste!
One Test is Worth 1000 Expert Opinions - The plural of anecdote is not data.
We can't solve problems by using the same kind of thinking we used when we created them. - Unknown
Reply With Quote
Thank You to Marc for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)

Bookmarks

Tags
calibration, calibration certificate, calibration sticker, due date, expiration date


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
How to get the Calibration Due Date? Calibration Certificates harinandu2003 Calibration Frequency (Interval) 11 3rd June 2009 03:16 AM
Done date & due date on calibration stickers mhirand20 ISO 17025 - Calibration, Measurement Gages and Test Laboratories 12 3rd December 2008 01:14 AM
Calibration grace period after due date GreatManBehinD Calibration Frequency (Interval) 5 31st October 2008 01:24 PM
Late calibration (Calibration Past Due Date) - What do I do now? dbulak ISO 17025 - Calibration, Measurement Gages and Test Laboratories 32 25th May 2007 10:18 AM
Calibration Due Date - How can I word my procedure? SCOTT SNYDER ISO 17025 - Calibration, Measurement Gages and Test Laboratories 38 5th April 2002 10:40 AM



The time now is 06:38 AM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts