The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > National and International Business Standards > CE Marking (Conformité Européene) / CB Scheme


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 4th June 2007, 02:50 AM
Regulatory Regulatory is offline
Involved in Discussions

Registration Date: May 2007
Location: India/Tamil Nadu/Chennai
 
Posts: 17
Thanks Given to Others: 0
Thanked 1 Time in 1 Post
Karma Power: 11
Karma: 15
Regulatory has less than 100 Karma points so far.
Please Help! Preparing the Technical File for a 12 Lead ECG Monitor

Hi,

We are preparing the Technical File for the 12 Lead ECG Monitor.

Should we have to submit the Clinical Report, If yes can any guide us how, if No...how to submit?

In FDA if we follow the Guidance for Industry - Diagnostic ECG Guidance (EC 11) standard - its is enough, there is no need to submit the Clinical Reports.

Will the same can be followed for CE Marking procedures also....I tried to look at the MDD Directives but i couldn't find any information!

Can any one please help us!
Reply With Quote

Sponsored Links
  #2  
Old 4th June 2007, 09:29 AM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Forum Moderator

Registration Date: Jun 2002
Location: Lawn Guyland
Age: 59
 
Posts: 3,101
Thanks Given to Others: 48
Thanked 390 Times in 272 Posts
Karma Power: 192
Karma: 4968
Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
Default Re: Preparing the Technical File for a 12 Lead ECG Monitor

Quote:
Originally Posted by Regulatory View Post

Hi,

We are preparing the Technical File for the 12 Lead ECG Monitor.

Should we have to submit the Clinical Report, If yes can any guide us how, if No...how to submit?

In FDA if we follow the Guidance for Industry - Diagnostic ECG Guidance (EC 11) standard - its is enough, there is no need to submit the Clinical Reports.

Will the same can be followed for CE Marking procedures also....I tried to look at the MDD Directives but i couldn't find any information!

Can any one please help us!
You need to include this as part of your Technical File which your Notified Body will review.
__________________
Al
Reply With Quote
Sponsored Links

  #3  
Old 4th June 2007, 11:13 AM
Regulatory Regulatory is offline
Involved in Discussions

Registration Date: May 2007
Location: India/Tamil Nadu/Chennai
 
Posts: 17
Thanks Given to Others: 0
Thanked 1 Time in 1 Post
Karma Power: 11
Karma: 15
Regulatory has less than 100 Karma points so far.
Default Re: Preparing the Technical File for a 12 Lead ECG Monitor

Can you please be more specific.......!
Reply With Quote
  #4  
Old 4th June 2007, 11:37 AM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Forum Moderator

Registration Date: Jun 2002
Location: Lawn Guyland
Age: 59
 
Posts: 3,101
Thanks Given to Others: 48
Thanked 390 Times in 272 Posts
Karma Power: 192
Karma: 4968
Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
Default Re: Preparing the Technical File for a 12 Lead ECG Monitor

Quote:
Originally Posted by Regulatory View Post

Can you please be more specific.......!
First you need to read and understand the MDD. You will then need a Notified Body to which you will submit your Technical File for review. Your ISO 13485 Registrar may be able to help you.
__________________
Al
Reply With Quote
Thanks to Al Rosen for your informative Post and/or Attachment!
  #5  
Old 5th June 2007, 03:18 AM
Regulatory Regulatory is offline
Involved in Discussions

Registration Date: May 2007
Location: India/Tamil Nadu/Chennai
 
Posts: 17
Thanks Given to Others: 0
Thanked 1 Time in 1 Post
Karma Power: 11
Karma: 15
Regulatory has less than 100 Karma points so far.
Default Re: Preparing the Technical File for a 12 Lead ECG Monitor

First you need to read and understand the MDD. You will then need a Notified Body to which you will submit your Technical File for review. Your ISO 13485 Registrar may be able to help you.

Hi...Thanks for your inputs.

We are certified as ISO 13485:2003 company...we have also got the CE Marking certification for our new product.

I went through the complete MDD - 93/42/EEC Directive....It is general document.

My Question is....Since ECG is proved concept should we need to submit the Technical File with the Clinical documents!

I hope my question is clear!
Reply With Quote
  #6  
Old 5th June 2007, 04:44 AM
chris1price chris1price is offline
Courtesy Access

Registration Date: Dec 2005
Location: Berks, UK
 
Posts: 129
Thanks Given to Others: 28
Thanked 98 Times in 59 Posts
Karma Power: 26
Karma: 417
chris1price is appreciated, and has over 400 Karma points.chris1price is appreciated, and has over 400 Karma points.chris1price is appreciated, and has over 400 Karma points.chris1price is appreciated, and has over 400 Karma points.chris1price is appreciated, and has over 400 Karma points.
Default Re: Preparing the Technical File for a 12 Lead ECG Monitor

Hi
Read through Annex X, section 1.1, this explains what you need to meet Section 1 and 3 of the ER. You either need clinical data or a suitable compliation of scientific literature. You can use information on predicate devices as part of the literature review.

What type of clinical data is required is a discussion you need to have with your Notified Body. Depending on the product classification and the history with the client, sometimes Notified Bodies will want this submitted in advance and other times it will only be examined during an audit.
Reply With Quote
Thank You to chris1price for your informative Post and/or Attachment!
  #7  
Old 5th June 2007, 09:35 AM
Al Rosen's Avatar
Al Rosen Al Rosen is offline
Forum Moderator

Registration Date: Jun 2002
Location: Lawn Guyland
Age: 59
 
Posts: 3,101
Thanks Given to Others: 48
Thanked 390 Times in 272 Posts
Karma Power: 192
Karma: 4968
Al Rosen is appreciated, and has over 1700 Karma points.
Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.Al Rosen is appreciated, and has over 1700 Karma points.
Send a message via AIM to Al Rosen
Default Re: Preparing the Technical File for a 12 Lead ECG Monitor

Quote:
Originally Posted by Regulatory View Post

First you need to read and understand the MDD. You will then need a Notified Body to which you will submit your Technical File for review. Your ISO 13485 Registrar may be able to help you.

Hi...Thanks for your inputs.

We are certified as ISO 13485:2003 company...we have also got the CE Marking certification for our new product.

I went through the complete MDD - 93/42/EEC Directive....It is general document.

My Question is....Since ECG is proved concept should we need to submit the Technical File with the Clinical documents!

I hope my question is clear!
Yes, you do. You also need to comply with IEC 60601-1 and any other appropriate harmonised standard. Again, this is something your Notified Body will tell you. I hope that I have been clear enough. You need a Notified Body!
__________________
Al
Reply With Quote
  #8  
Old 5th June 2007, 11:56 AM
Regulatory Regulatory is offline
Involved in Discussions

Registration Date: May 2007
Location: India/Tamil Nadu/Chennai
 
Posts: 17
Thanks Given to Others: 0
Thanked 1 Time in 1 Post
Karma Power: 11
Karma: 15
Regulatory has less than 100 Karma points so far.
Default Re: Preparing the Technical File for a 12 Lead ECG Monitor

Thanks for your inputs!

It's been great learning experience.....we have Notified body but the problem is they are very slow in response.
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > National and International Business Standards > CE Marking (Conformité Européene) / CB Scheme

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Technical File Format - MEDDEV 2.5.1 AOIRA CE Marking (Conformité Européene) / CB Scheme 2 28th October 2009 09:59 PM
CE Mark / TCF (Technical Construction File) File sample achih100 CE Marking (Conformité Européene) / CB Scheme 6 27th October 2009 02:51 PM
Technical File - Own Brand Labeling Requirements Rob Udo CE Marking (Conformité Européene) / CB Scheme 4 10th September 2009 10:19 AM
CE Marking Technical File questions azaplas ISO 13485 - Medical Devices - Quality Management Systems 2 21st February 2009 06:22 AM
Level of Technical Detail of Technical File Submission jscholen ISO 13485 - Medical Devices - Quality Management Systems 1 8th February 2008 01:36 PM



The time now is 06:33 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts