Determining Calibration Frequency for Ammeters & Voltmeters used on the line
howI can the calibration interval for the ameters& voltmeters ( allways used on the production line)
|

6th June 2007, 04:10 PM
|
|
Tater Salad
Registration Date: Dec 2006
Location: CT
|
|
Posts: 379
Thanks Given to Others: 89
Thanked 149 Times in 107 Posts
Karma Power: 37
|
|
Determining Calibration Frequency for Ammeters & Voltmeters used on the line
Hispeed,
There are quite a few threads which touch upon this subject in detail, I suggest using the search function and reading through these. This thread is a more in depth mathematical exploration into setting intervals based on instrument performance & trends, so hopefully you aren't reading only this thread and wondering what you got yourself into lol. This level of detail isn't required for common measuring tools on plant floors.
|

6th June 2007, 04:32 PM
|
 |
Your Elsmar Cove Host
Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
|
|
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,567 Times in 1,019 Posts
Karma Power: 605
|
|
Re: Determining Calibration Frequency for Ammeters & Voltmeters used on the line
__________________
A Search is a terrible thing to waste!
One Test is Worth 1000 Expert Opinions - The plural of anecdote is not data.
We can't solve problems by using the same kind of thinking we used when we created them. - Unknown
|
|
Thanks to Marc for your informative Post and/or Attachment!
|
|

6th June 2007, 04:35 PM
|
 |
$ Contributor
Registration Date: Oct 2006
Location: São Paulo, Brazil
Age: 41
|
|
Posts: 614
Thanks Given to Others: 250
Thanked 241 Times in 178 Posts
Karma Power: 57
|
|
Re: Determining Calibration Frequency for Ammeters & Voltmeters used on the line
Quote:
Originally Posted by hispeed
howI can the calibration interval for the ameters& voltmeters ( allways used on the production line)
|
We calibrate them every six months... ours are used daily on MIG welding machines...
__________________
Ted
Nothing is difficult, you just didn´t understand it yet...
|
|
Thanks to Ted Schmitt for your informative Post and/or Attachment!
|
|

6th June 2007, 06:14 PM
|
 |
Forum Moderator
Registration Date: Jun 2002
Location: Lawn Guyland
Age: 59
|
|
Posts: 3,101
Thanks Given to Others: 48
Thanked 390 Times in 272 Posts
Karma Power: 192
|
|
Re: Determining Calibration Frequency for Ammeters & Voltmeters used on the line
Quote:
Originally Posted by hispeed
howI can the calibration interval for the ameters& voltmeters ( allways used on the production line)
|
Start with the mfrs recommended calibration frequency. If the history of the as found condition is within spec for a specified number of cal cycles, you can reduce the frequency of calibrations. On the other hand, if the as found condition is out, you should increase the frequency.
__________________
Al
|
|
Thanks to Al Rosen for your informative Post and/or Attachment!
|
|

7th June 2007, 08:43 AM
|
|
Inactive Registered Visitor
Registration Date: Jun 2007
|
|
Posts: 4
Thanks Given to Others: 2
Thanked 0 Times in 0 Posts
Karma Power: 11 Karma: 10 
|
|
Re: Determining Calibration Frequency for Ammeters & Voltmeters used on the line
In fact, the MSA is a new topic to me ,so I can not understand the meanning of mfrs,as,and cal. So please explain
|

7th June 2007, 09:24 AM
|
|
Tater Salad
Registration Date: Dec 2006
Location: CT
|
|
Posts: 379
Thanks Given to Others: 89
Thanked 149 Times in 107 Posts
Karma Power: 37
|
|
Re: Determining Calibration Frequency for Ammeters & Voltmeters used on the line
Mfr was just an abbreviation for manufacturer. If you contact the company who makes the instrument (OEM) they should be able to supply you with a suggested calibration interval.
As supplied simply refers to the readings you obtain when you calibrate an item (prior to making any adjustments).
Cal. is just an abbreviation for calibration.
As you are new to MSA and possibly calibration? You may want to forego anything overly complicated until you can obtain further training. I would try to find out if there are any requirements for your specific industry regardingi these devices...if there are no specific interval requirements I would suggest following the manufacturers recommended calibration interval.
If you get the devices calibrated by an outside calibration laboratory ISO 17025 these folks may be able to suggest a reliable calibration interval which will provide you with acceptable confidence. Don't be afraid to utilize their expertise either, if you have questions about the calibrations, intervals, msa studies, get on the phone with them and see if they can assist you, most would be more than happy to provide you with the information you need.
|

7th June 2007, 10:20 AM
|
 |
Your Elsmar Cove Host
Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
|
|
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,567 Times in 1,019 Posts
Karma Power: 605
|
|
Re: Determining Calibration Frequency for Ammeters & Voltmeters used on the line
Quote:
Originally Posted by hispeed
In fact, the MSA is a new topic to me ,so I can not understand the meanning of mfrs,as,and cal. So please explain
|
You might want to do some 'homework'. Most terms are defined in The Elsmar Cove Wiki and in the Definitions, Acronyms, Abbreviations and Interpretation forum. Mfrs and Cal are notable exceptions because they are so common.
__________________
A Search is a terrible thing to waste!
One Test is Worth 1000 Expert Opinions - The plural of anecdote is not data.
We can't solve problems by using the same kind of thinking we used when we created them. - Unknown
|
Lower Navigation Bar
|
|
|
|
Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
|
|
|
| Thread Tools |
Search this Thread |
|
|
|
| Display Modes |
Rate Thread Content |
Linear Mode
|
|
Posting Settings
|
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
HTML code is Off
|
|
|
|
|