Go Back   The Elsmar Cove Forum > National and International Business Standards > FDA (United States Food and Drug Administration) > Medical Devices (21 CFR part 820)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 10th September 2007, 10:28 AM
temujin's Avatar
temujin temujin is offline
Involved in Discussions

Registration Date: May 2007
Location: Europa
 
Posts: 162
Thanks Given to Others: 44
Thanked 40 Times in 30 Posts
Karma Power: 25
Karma: 566
temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.
Please Help! Material Change - New 10993 Testing and Submitting a New 510(k)

Hi,

We are considering substituting one material in our cardiac probes, and I am a little confused with regard to the text in Deciding When to Submit a 510(k) for a Change to an Existing Device

The question is if we need to submit a new 510(k) when new 10993-testing is necessary.
C2.2.1
(...)
A 510(k) may not be needed if the manufacturer has satisfactory results from the testing indicated by ISO 10993-1 in its files for the material in question or if such results are available to the manufacturer, e.g., are available in the open published literature or have been provided to the 510(k) holder by the material supplier.
(...)
If I understand the text right, if the material supplier can prove that the material has been tested with respect to the necessary 10993 standards we do not need to submit a new 510(k).

However, if we decide to run the tests ourselves (or through an extern test laboratory) we need to submit a new 510(k).

Which would mean that in order to avoid a new 510(k) we could simply carry out the new testing through the material supplier.?? This sounds strange to me....?

Any suggestions as to how I should interpret this ?

best regards
t.
Reply With Quote

Sponsored Links
  #2  
Old 10th September 2007, 12:51 PM
Steve McQuality's Avatar
Steve McQuality Steve McQuality is offline
Quality Engineer

Registration Date: Mar 2006
Location: State College, Pennsylvania
Age: 48
 
Posts: 104
Thanks Given to Others: 12
Thanked 19 Times in 11 Posts
Karma Power: 19
Karma: 146
Steve McQuality is appreciated, and has over 100 Karma points.Steve McQuality is appreciated, and has over 100 Karma points.
Default Re: Material Change and 10993

Quote:
Originally Posted by temujin View Post

Hi,
C2.2.1
(...)
A 510(k) may not be needed if the manufacturer has satisfactory results from the testing indicated by ISO 10993-1 in its files for the material in question or if such results are available to the manufacturer, e.g., are available in the open published literature or have been provided to the 510(k) holder by the material supplier.
(...)
If I understand the text right, if the material supplier can prove that the material has been tested with respect to the necessary 10993 standards we do not need to submit a new 510(k).

However, if we decide to run the tests ourselves (or through an extern test laboratory) we need to submit a new 510(k).

Which would mean that in order to avoid a new 510(k) we could simply carry out the new testing through the material supplier.?? This sounds strange to me....?

Any suggestions as to how I should interpret this ?

best regards
t.
Hi t.

This is just my opinion from interpretting the C2.2.1 excerpt you've quoted, but I read this as either obtaining satisfactory results from ISO 10993-1 testing performed OR obtaining the information from your supplier. It seems to me that you are being provided the option.

However, if the results of the ISO 10993-1 are different from the original material then you must submit a new 510(k).

...just my opinion regarding MY interpretation of the standard you've quoted.

-Steve
Reply With Quote
Sponsored Links

  #3  
Old 11th September 2007, 02:25 AM
Gert Sorensen's Avatar
Gert Sorensen Gert Sorensen is offline
QA Systems Coordinator

Registration Date: Apr 2006
Location: Denmark
Age: 42
 
Posts: 720
Thanks Given to Others: 327
Thanked 224 Times in 134 Posts
Karma Power: 73
Karma: 2221
Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.Gert Sorensen is appreciated, and has over 1700 Karma points.
Default Re: Material Change - New 10993 Testing and Submitting a New 510(k)

Quote:
Originally Posted by temujin View Post

C2.2.1
(...)
A 510(k) may not be needed if the manufacturer has satisfactory results from the testing indicated by ISO 10993-1 in its files for the material in question or if such results are available to the manufacturer, e.g., are available in the open published literature or have been provided to the 510(k) holder by the material supplier.
(...)
If I understand the text right, if the material supplier can prove that the material has been tested with respect to the necessary 10993 standards we do not need to submit a new 510(k).

However, if we decide to run the tests ourselves (or through an extern test laboratory) we need to submit a new 510(k).

Which would mean that in order to avoid a new 510(k) we could simply carry out the new testing through the material supplier.?? This sounds strange to me....?
The new 10993, has the intention of reducing testing on animals, and it seems that the FDA has accepted this to some extent. IMO what this means in this case is that if the new material has been tested (and approved) and you have access to these data, either from literature, supplier or laboratory etc, then you can use these data as the rationale for your substition, and you will not need a new 510K.

However, you will of course need to update your DMR to reflect the changes. You will also need to make sure that any testing done is actually sufficient. In 10993 the responsibility for choosing the correct testing lies solely with the manufacturer. So if your new material has not been previously tested with regards to, say intracutaneous reaction or long term use, then it is your responsibility to have it tested for this if this test applies to your product.

Additionally you will of course also need to test the material for its intended use in your device.
__________________
Beetle Bailey Rules!
Reply With Quote
Thanks to Gert Sorensen for your informative Post and/or Attachment!
  #4  
Old 11th September 2007, 06:26 AM
Ajit Basrur's Avatar
Ajit Basrur Ajit Basrur is offline
Super Moderator

Registration Date: Dec 2005
Location: India
Age: 46
 
Posts: 2,645
Thanks Given to Others: 674
Thanked 802 Times in 579 Posts
Karma Power: 176
Karma: 5443
Ajit Basrur is appreciated, and has over 1700 Karma points.
Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.
Default Re: Material Change - New 10993 Testing and Submitting a New 510(k)

There are different criteria applicaqble for the type of the device. For eg. - change in material in case of implant, "Manufacturers should submit a new 510(k) for a change in implant material where the material contacts tissue (including bone tissue) or body fluid. Examples of devices for which changes in material type would normally require a new 510(k) are total joints or their components. On the other hand, changes in materials of an implant that are not intended to contact body tissues or fluids are not likely to require a 510(k) submission. Examples of such changes in material type are changes in the interior materials of an implantable electric stimulator (e.g., a single chamber cardiac pacemaker) which are sealed from ingress of body fluids or tissues."

Sometimes change in materials may affect the performance standards in which case, the manufacturer has to decide whether a 510 (K0 is required.

Pl refer http://www.fda.gov/cdrh/ode/510kmod.html#page16

Thus IMO, you can not come out with a generic statement as mentioned in the OP.
__________________
I have no special talents. I am only passionately curious - Albert Einstein
Reply With Quote
Thank You to Ajit Basrur for your informative Post and/or Attachment!
  #5  
Old 11th September 2007, 07:34 AM
temujin's Avatar
temujin temujin is offline
Involved in Discussions

Registration Date: May 2007
Location: Europa
 
Posts: 162
Thanks Given to Others: 44
Thanked 40 Times in 30 Posts
Karma Power: 25
Karma: 566
temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.temujin is appreciated, and has over 500 Karma points.
Default Re: Material Change - New 10993 Testing and Submitting a New 510(k)

Thanks to all so far,

I have used the Flow Charts in the FDA´s guidance document. The device is not an implant,but it will be in contact with body tissue/blood in vivo.

This takes me to section 2.2.1-" Considering that the material is likely to contact in vivo body tissues or fluids and the requirements of ISO 10993-1, is additional testing required"

I find it hard to interpret what they mean.
The material we introduce is new to us, so testing it according to the relevant 10993 standards is something we must do.

But even at this point, I´m not sure if we need to submit a new 510(k).

If I interpret the text I quoted in my first post to be such that as
<< We don´t need to submit a new 510(k) as long as we have the respective 10993 test reports - either from the supplier or made by us in an external laboratory >> Then why is this question even in FDA´s flowchart.

Wouldn´t just everybody arriving at this point just carry out the tests and document it without submitting a new 510(k) if this is a possiblility.

t.
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > National and International Business Standards > FDA (United States Food and Drug Administration) > Medical Devices (21 CFR part 820)

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Submitting a 510(k)-the cuff individually or the system wangyang Medical Devices (21 CFR part 820) 2 5th September 2009 10:19 AM
ISO 10993-5:2009 vs. ISO 10993-5:2003 differences - Cytotoxicity testing waziz ISO 13485 - Medical Devices - Quality Management Systems 2 21st June 2009 03:13 AM
Establishing a biocompability test lab for testing EN 10993 series ersincetin CE Marking (Conformité Européene) / CB Scheme 1 7th February 2009 02:21 PM
Is testing per ISO 10993 - 5 required for resin discoloration ? Ajit Basrur ISO 13485 - Medical Devices - Quality Management Systems 3 11th June 2008 01:04 PM
Biocompatibiltiy testing of materials (ISO 10993-1) vs. Testing of final product? Asherlee ISO 13485 - Medical Devices - Quality Management Systems 5 24th May 2006 05:19 AM



The time now is 08:53 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts