The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > National and International Business Standards > FDA (United States Food and Drug Administration) > Medical Devices (21 CFR part 820)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 10th January 2008, 04:42 PM
jscholen jscholen is offline
Involved in Discussions

Registration Date: Jan 2008
Location: Atlanta
 
Posts: 50
Thanks Given to Others: 16
Thanked 14 Times in 9 Posts
Karma Power: 10
Karma: 65
jscholen has less than 100 Karma points so far.
Question Process Validation needed with IDE Approval?

This is clearly a gray area from my perspective and I am looking for some input:

Since IDE(Investigational Device Exemption) approvals exempt a company from all CFR but 820.30 Design Controls and GDP, does this give a company freedom not to validate a process even if it is manual and you are doing 100% verification of a build but sampling on the testing(using the device is destructive.)??
Reply With Quote

Sponsored Links
  #2  
Old 10th January 2008, 09:10 PM
Ajit Basrur's Avatar
Ajit Basrur Ajit Basrur is offline
Super Moderator

Registration Date: Dec 2005
Location: India
Age: 46
 
Posts: 2,647
Thanks Given to Others: 674
Thanked 802 Times in 579 Posts
Karma Power: 176
Karma: 5443
Ajit Basrur is appreciated, and has over 1700 Karma points.
Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.
Default Re: Process validation needed with IDE Approval?

Quote:
Originally Posted by jscholen View Post

This is clearly a gray area from my perspective and I am looking for some input:

Since IDE(Investigational Device Exemption) approvals exempt a company from all CFR but 820.30 Design Controls and GDP, does this give a company freedom not to validate a process even if it is manual and you are doing 100% verification of a build but sampling on the testing(using the device is destructive.)??
Hi jscholen,

For a particular process to be validated / verified, I always would like to refer the GHTF document - Quality Management Systems - Process Validation Guidance
__________________
I have no special talents. I am only passionately curious - Albert Einstein
Reply With Quote
Thanks to Ajit Basrur for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 11th January 2008, 12:27 PM
yodon yodon is offline
Registered Visitor

Registration Date: Jun 2005
Location: Plano, TX
Age: 50
 
Posts: 186
Thanks Given to Others: 119
Thanked 83 Times in 63 Posts
Karma Power: 31
Karma: 491
yodon is appreciated, and has over 400 Karma points.yodon is appreciated, and has over 400 Karma points.yodon is appreciated, and has over 400 Karma points.yodon is appreciated, and has over 400 Karma points.yodon is appreciated, and has over 400 Karma points.
Default Re: Process validation needed with IDE Approval?

You're right in that you are somewhat exempt from manufacturing controls, I believe validation is determined on a case-by-case basis. For example, we are developing a device that will go out as IDE initially and we know (I believe through discussions with the FDA) that the sterilization process will have to be validated. We will also have to show biologic compatibility. So clearly, FDA has authority to exercise some judgment in this area. So I think you'd be best served by discussing what FDA will expect with your FDA rep.
Reply With Quote
  #4  
Old 14th January 2008, 11:19 AM
sanjoh sanjoh is offline
Inactive Registered Visitor

Registration Date: Mar 2003
Location: southeast
 
Posts: 12
Thanks Given to Others: 0
Thanked 4 Times in 2 Posts
Karma Power: 31
Karma: 169
sanjoh is appreciated, and has over 100 Karma points.sanjoh is appreciated, and has over 100 Karma points.
Default Re: Process validation needed with IDE Approval?

Quote:
Originally Posted by jscholen View Post

This is clearly a gray area from my perspective and I am looking for some input:

Since IDE(Investigational Device Exemption) approvals exempt a company from all CFR but 820.30 Design Controls and GDP, does this give a company freedom not to validate a process even if it is manual and you are doing 100% verification of a build but sampling on the testing(using the device is destructive.)??
Most Companies hope to support a PMA submission with data from the IDE. As part of the submission, your manufacturing processes will be audited. In my experience, having the processes developed and validated as soon as possible can help to avoid filing a PMA supplement because of a process change.

With your example, the fact that the process is manual is not relevant. If the output of the process is 100% verifiable as you say, then validation is not required.

I'm not sure what you mean by "sampling on the testing". Are you taking a sample to test?

Sounds like you may need to define your process outputs.
Reply With Quote
  #5  
Old 14th January 2008, 11:33 AM
jscholen jscholen is offline
Involved in Discussions

Registration Date: Jan 2008
Location: Atlanta
 
Posts: 50
Thanks Given to Others: 16
Thanked 14 Times in 9 Posts
Karma Power: 10
Karma: 65
jscholen has less than 100 Karma points so far.
Default Re: Process validation needed with IDE Approval?

Quote:
Originally Posted by sanjoh View Post


I'm not sure what you mean by "sampling on the testing".
820.75(a) stated that if "the process is not fully verified by subsequent inspection and test"...

since we are not able to do 100% testing since the testing is destructive, we sample test, but the device is 100% verified for correct assembly.

This is my latest view...since the device is investigational and exempt from validations except as noted by yodon, the device must be adequatley tested/inspected to meet user expectations, not specifications.....

what does everyone think of that statement?
Reply With Quote
  #6  
Old 14th January 2008, 12:27 PM
sanjoh sanjoh is offline
Inactive Registered Visitor

Registration Date: Mar 2003
Location: southeast
 
Posts: 12
Thanks Given to Others: 0
Thanked 4 Times in 2 Posts
Karma Power: 31
Karma: 169
sanjoh is appreciated, and has over 100 Karma points.sanjoh is appreciated, and has over 100 Karma points.
Default Re: Process validation needed with IDE Approval?

Quote:
Originally Posted by jscholen View Post

820.75(a) stated that if "the process is not fully verified by subsequent inspection and test"...

since we are not able to do 100% testing since the testing is destructive, we sample test, but the device is 100% verified for correct assembly.

This is my latest view...since the device is investigational and exempt from validations except as noted by yodon, the device must be adequatley tested/inspected to meet user expectations, not specifications.....

what does everyone think of that statement?
It sounds as if you've made up your mind and are looking for others to support your conclusion.

I don't agree with the approach as it ignores the long term implications.

I suggest you document your conclusion and be prepared to defend it with FDA.
Reply With Quote
  #7  
Old 14th January 2008, 12:37 PM
jscholen jscholen is offline
Involved in Discussions

Registration Date: Jan 2008
Location: Atlanta
 
Posts: 50
Thanks Given to Others: 16
Thanked 14 Times in 9 Posts
Karma Power: 10
Karma: 65
jscholen has less than 100 Karma points so far.
Default Re: Process validation needed with IDE Approval?

I'm too new for support, just looking for advice...

Since you don't agree with the statement, why does the FDA give companies IDE approval for a device with exemption from all but 820.30? I am assuming that a device specification is not fully verifiable in this situation, but the user expectations is....
Reply With Quote
  #8  
Old 16th January 2008, 03:18 PM
JohnM's Avatar
JohnM JohnM is offline
Involved in Discussions

Registration Date: Nov 2006
Location: New York
 
Posts: 20
Thanks Given to Others: 4
Thanked 24 Times in 3 Posts
Karma Power: 16
Karma: 190
JohnM is appreciated, and has over 100 Karma points.JohnM is appreciated, and has over 100 Karma points.
Default Re: Process Validation needed with IDE Approval?

I would argue that some validation is part of design control and risk mitigation which is also part of design control. I would not be quick to dismiss validation.

Regards,

John
__________________
John Minier
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > National and International Business Standards > FDA (United States Food and Drug Administration) > Medical Devices (21 CFR part 820)

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Process Validation Format for Press (Stamping) Operations - Help needed atulpatil APQP and PPAP 11 22nd September 2009 08:54 AM
Process & Tooling Validation Help Needed francomc Documentation Control Systems, Procedures, Forms and Templates 4 20th May 2009 01:31 PM
When Validation Needed? We are changing to new process equipments mashiroma ISO 13485 - Medical Devices - Quality Management Systems 4 21st March 2008 12:48 AM
Tool Approval Process - Should we be using Cmk for tool validation DJN SPC Monitoring and Statistical Analysis Techniques 2 22nd July 2005 06:02 AM
TS 16949 - 7.5.2 - Process Validation - Criteria for review and approval of processes Morpheus ISO/TS 16949 - International Automotive Quality Systems Standard 8 8th August 2003 12:14 PM



The time now is 11:35 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts