The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Measurement Uncertainty (MU)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 28th April 2008, 12:28 PM
JoeQNovice JoeQNovice is offline
Involved in Discussions

Registration Date: Jan 2008
Location: Connecticut
 
Posts: 28
Thanks Given to Others: 26
Thanked 0 Times in 0 Posts
Karma Power: 8
Karma: 21
JoeQNovice has less than 100 Karma points so far.
Please Help! Measurement Uncertainty Budget for Comparative Dimensional Measurements

Unique(?) situation.
Have to calculate an uncertainty budget for diameter. Gage is a custom-made (by Mahr-Federal) comparator. Standards are gage blocks sent to NIST for calibration. Subjects are gage blocks. Gage blocks and Standards (Masters) are of varying size - Let's say 10 different sizes. They are also of differing materials (ceramic, tungsten carbide and steel).
We measure gage blocks for customers. They are mostly steel, some ceramic and carbides come in.

The comaprator is set up with a Standard/Master. So to measure a 1/4 gage block, a 1/4 in Master is put into the gage and it is zero-ed out. Then the subject 1/4 inch block is measured. NIST says the uncertainty is .000002. When we measure the NIST Master we get .000007 on average.

Room is class 10000 with a constant 68 degrees +/1 .5. Masters are always at this temp. Subjects are allowed to settle to temp before measurement.

To calculate uncertainy, we took the Standards (masters) and had both (2) technicians measure all 10 Standard/Masters 3 times each and record the readings. The Standard deviation was calculated for all of the readings - I assume this is our Type A first step. Again, our averages were about .000005 worse than what NIST got. How do we deal with the differences in measurement we get from what NIST gets - or do we ignore what NIST gets and just factor our own readings?

Do I have to factor thermal coefficients for all three materials used ?

Can I measure a 1/4 inch tungsten carbide subject with a 1/4 inch ceramic master?

How can I factor gage uncertainty for a custom gage that we calibrate in house?

Should I be measuring something other than the NIST Masters because the NIST Masters are what we set up the comparator to?

Does anyone have an idea of what factors traditionally are considered in addition to standard deviation (type A) factors?

In it over my head and looking for guidance - thanks!
Reply With Quote

Sponsored Links
  #2  
Old 29th April 2008, 05:15 PM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 53
 
Posts: 1,621
Thanks Given to Others: 1
Thanked 310 Times in 224 Posts
Karma Power: 186
Karma: 6526
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Default Re: Uncertainty Budget for comparative dimensional measurements

Great questions!

That your MU is different than NIST, even on your Masters is not an issue.....I would question it if you did not. You have a good setup from the description, but remember NIST is comparing to a still higher standard with a still lower MU, so don't stress about that.

The measurements you do and MU you get from the calibration of the other gage blocks will likely be higher than what you get from your Masters, also OK and normal.

Now, if way out of line, say 1000 times more for the same class and material block, then something is not quite right.

Now, if you are using the Masters to setup to cal the ceraminc or tungsten, then yes, you must take the material difference into account. Both ceramic and tungsten have different coefficients of expansion than steel.

That gage unit itself must also be calibrated which you do in-house using your Masters. This calibration must have the uncertainty calculated. For this calibration, the NIST uncertainty is expressed at either k=2 or k=3, not sure which NIST has reported to you, but that k factor becomes your divisor to return that uncertainty value to standard uncertainty for calculation. This is a Type B with normal distribution.

The calibration of your gage will of course be documented at k=2, so to determine the uncertainty of the calibration of the other blocks you divide by two to return to standard uncertainty. The uncertainty of the calibration of your gage is a Type B with normal distribution.

Now, even though the Masters were considered during the calibration of the gage, if also used to setup the gage for calibration of other blocks, then the uncertainty of the Masters must be considered a second time. This is a Type B contribution and is normal distribution.

Even though the temp is well controlled in your room, when using the Masters, you must consider the difference in temp between what NIST reports and your actual. This is a Type B with rectangular distribution. The coeffieicient of expansion is used to arrive at the actual difference based on temp.

You may be able to lower your uncertainty by going to a laser interferometer, but the cost likely does not make that worthwhile, unless you really need that.

All in all, what you are reporting seems to be reasonable numbers for your uncertainty. you may find the uncertainty going up a bit when all the Type B's are included, but that is also normal.

Hope this helps.
Reply With Quote
Thank You to Hershal for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 29th April 2008, 07:50 PM
JoeQNovice JoeQNovice is offline
Involved in Discussions

Registration Date: Jan 2008
Location: Connecticut
 
Posts: 28
Thanks Given to Others: 26
Thanked 0 Times in 0 Posts
Karma Power: 8
Karma: 21
JoeQNovice has less than 100 Karma points so far.
Question Re: Uncertainty Budget for comparative dimensional measurements

I thank you for your clear explaination, but have one further clarification - if you will - Because I have Masters of varying sizes and because the effects of the expansion on the individual sizes due to the thermal coefficient seems more significant for smaller blocks than larger ones, do I have to calculate and list results for each of the ten blocks and stratify my uncertainties to show the effects on each? I notice a difference in the NIST reported uncertainties for each size... It's important that I accurately report my uncertainty for dimensionals because we are considering an A2LA certification with dimensional measurement in the Scope.

Thanks again!
Reply With Quote
  #4  
Old 1st May 2008, 01:45 PM
jfgunn jfgunn is offline
Lab Manager

Registration Date: May 2004
Location: Cleveland, Ohio
Age: 33
 
Posts: 74
Thanks Given to Others: 1
Thanked 50 Times in 30 Posts
Karma Power: 29
Karma: 274
jfgunn is appreciated, and has over 200 Karma points.jfgunn is appreciated, and has over 200 Karma points.jfgunn is appreciated, and has over 200 Karma points.
Default Re: Measurement Uncertainty Budget for Comparative Dimensional Measurements

Another point to consider: Contact Deformation on the part vs contact deformation of the item you are testing.....

Go to the following website where you can calculate how much different materials will deform when contacted by different geometery measuring heads at differnet forces.

Say for example that you have a steel master that is really hard and you are measuring a piece of rubber. The peice of rubber will deform some which must be accounted for in the measurement. Clearly, Rubber is an extreme example.

http://emtoolbox.nist.gov/Elastic/Photo.asp
Reply With Quote
Thanks to jfgunn for your informative Post and/or Attachment!
  #5  
Old 5th May 2008, 11:36 AM
Mchurch Mchurch is offline
Inactive Registered Visitor

Registration Date: Mar 2008
Location: Dallas Texas
 
Posts: 11
Thanks Given to Others: 1
Thanked 2 Times in 2 Posts
Karma Power: 7
Karma: 10
Mchurch has less than 100 Karma points so far.
Default Re: Measurement Uncertainty Budget for Comparative Dimensional Measurements

Disclamer: My opinion will follow, I am not the expert, have done some digging recently and this is what I have found.

Another couple of thoughts to consider;
Uncertainty of temperature;
"Room is class 10000 with a constant 68 degrees +/1 .5. Masters are always at this temp. Subjects are allowed to settle to temp before measurement."

-There are different types of enviromental controls, I however am limited financially in my lab, in that I have a pulsed air conditioning unit. I have relatively the same controls 20 ±1 C. Meaning that my ac unit kicks on, Cools the lab to lower end around 19, turns off, lab warms up to near 21, and ac kicks on again. Due to the nature of the system, And the gages being of different materials the master and the uut can truely be 2 full degrees apart.
There are other systems that allow a constant flow of 20 degrees into the work enviroment though.

Uncertainty of the tempurature monitor system. If using a standard Thermohumidigraph or something of that nature your are introducing an unknown around ±1 F to the mix. Meaning the lab could actually be swinging ±2.5 F. A contact type of thermometer/thermocouple could greatly reduce this error if measuring the parts themselves.

Unknown of the Coefficient of Thermal Expansion. There are alot of charts out there that will give you a CTE for a type of material. They won't all agree, And I haven't found any resource that lists an uncertainty associated with these numbers. Here is a good ppt on the issue:
http: //emtoolbox.nist.gov/Temperature/Temperature%20and%20Dimensional%20Measurement.ppt
With that, I assume an unknown of the my CTE equal to ±10% of the CTE's value. So if I'm using 11.5 uIn/In/Deg C I have an Uncertainty of 2.3uIn/In/Deg C in that number.

So I have addopted the philosophy of
Eu=3^√(Ucte(Ut))^3+((Dif)(Ut)(CTE))^3+(CTE(Ut))^3

Where:
Ucte-Uncertainty of Coefficient of thermal expansion, 10 percent of both blocks summed together
Ut- Uncertainty of temperature monitoring system
Dif- Difference of temperature of the gages
CTE- Coefficient that will be used for both blocks summed together.

I use a rectangular distribution with a divisor of 1.732 for this number.

Hope I have been of some help, Or at least made you think about it enough to say: "That guy knows nothing!"

Last edited by Mchurch; 5th May 2008 at 02:34 PM.
Reply With Quote
Thanks to Mchurch for your informative Post and/or Attachment!
  #6  
Old 5th May 2008, 12:24 PM
Mchurch Mchurch is offline
Inactive Registered Visitor

Registration Date: Mar 2008
Location: Dallas Texas
 
Posts: 11
Thanks Given to Others: 1
Thanked 2 Times in 2 Posts
Karma Power: 7
Karma: 10
Mchurch has less than 100 Karma points so far.
Default Re: Uncertainty Budget for comparative dimensional measurements

Quote:
Originally Posted by JoeQNovice View Post

I thank you for your clear explaination, but have one further clarification - if you will - Because I have Masters of varying sizes and because the effects of the expansion on the individual sizes due to the thermal coefficient seems more significant for smaller blocks than larger ones, do I have to calculate and list results for each of the ten blocks and stratify my uncertainties to show the effects on each? I notice a difference in the NIST reported uncertainties for each size... It's important that I accurately report my uncertainty for dimensionals because we are considering an A2LA certification with dimensional measurement in the Scope.

Thanks again!
With the masters of varying size, You don't have to have a BMC for every size. A base uncertainty with and adder for longer measurements is the standard practice. Something like 5uIn + 2.1uInL Where L is the length of the unit you are going to be measuring. Work up your uncertainties at a couple different points throughout your measurement range(3-5 is usually acceptable), Find out how much your uncertainty increases per in, and express in that manner.
Reply With Quote
Thanks to Mchurch for your informative Post and/or Attachment!
  #7  
Old 6th May 2008, 10:23 AM
JoeQNovice JoeQNovice is offline
Involved in Discussions

Registration Date: Jan 2008
Location: Connecticut
 
Posts: 28
Thanks Given to Others: 26
Thanked 0 Times in 0 Posts
Karma Power: 8
Karma: 21
JoeQNovice has less than 100 Karma points so far.
Question Re: Measurement Uncertainty Budget for Comparative Dimensional Measurements

Just a quick question to satisfy a dispute the technicians are having in trying to calculate the standard deviation: I use a NIST master to set up the comparator. It has an uncertainty of .000002. When I measure it it reads an average .000005. Now I am taking my three sets of readings on the master for repeatability and get an average of .000006. Is my standard deviation (that is to become my Type A part of my budget) based on my deviation from my set-up reading of .000005 or NISTs reading of .000002?
Reply With Quote
  #8  
Old 6th May 2008, 10:42 AM
Mchurch Mchurch is offline
Inactive Registered Visitor

Registration Date: Mar 2008
Location: Dallas Texas
 
Posts: 11
Thanks Given to Others: 1
Thanked 2 Times in 2 Posts
Karma Power: 7
Karma: 10
Mchurch has less than 100 Karma points so far.
Default Re: Measurement Uncertainty Budget for Comparative Dimensional Measurements

I would take the standard deviation for Type A as a repeatability study from your technicians reading it. Ie all techs read it 3 times and crunch the numbers to find a std deviation between your readings.

Nist read the master at +2uIn and your reading at +5 uIn. That is not part of your type A, There is a good possibility that it was +2 at NIST and it is +5 in your lab for a variety of reasons.

So to clearly answer the question, Type A in your budget for the standard deviation will be the deviation from your set up, Not NIST's.
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Measurement Uncertainty (MU)

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Measurement Uncertainty Budget (MU) of Ultrasonic Thickness gage? ScottBP Measurement Uncertainty (MU) 4 8th June 2009 03:58 PM
Contributing Factors in a Supermicrometer Measurement Uncertainty Budget? Mchurch Measurement Uncertainty (MU) 8 5th June 2008 09:15 AM
Measurement Uncertainty for Adjustable Thread Rings but No Budget Made CalRich Measurement Uncertainty (MU) 4 26th March 2007 12:31 PM
Measurement Uncertainty (MU) Budget for a 0-1 inch Micrometer rich.dittmanjr Measurement Uncertainty (MU) 3 10th May 2005 08:23 AM
QUERIES On ISO 17025 AUDIT - Measurement Uncertainty Budget, etc. neo_inst Measurement Uncertainty (MU) 1 20th November 2002 03:48 PM



The time now is 08:18 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts