The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO and IEC Medical Device Related Standards > ISO 13485 - Medical Devices - Quality Management Systems


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 5th June 2008, 11:33 PM
Adele Adele is offline
Involved in Discussions

Registration Date: Oct 2005
Location: New Zealand
 
Posts: 47
Thanks Given to Others: 31
Thanked 18 Times in 12 Posts
Karma Power: 18
Karma: 97
Adele has less than 100 Karma points so far.
Question Biocompatibility testing & Registration in China

Hi All,

We are currently going through re-registeration of a medcial device in China and have been asked to provide biocompatibility data.

No problem (or so we thought) and we supplied the biocompatibility test reports that have been used in a number of registerations around the world (including the USA).

However, the testing was done in 2000 and 2001 and the Chinese authorities haven't accepted this as the testing must be to ISO 10993-1: 2003 and the other more recent versions of the specific ISO 10993 standards.

I have tried finding out specifics of the regulation but all I can get is that it is a new regulatory requirement issued in June 2007 (but it isn't available on the English SFDA site).

Does anyone know the specifics of the regulation? Has anyone else come across this problem? How did they solve it?

Thanks!
Reply With Quote

Sponsored Links
  #2  
Old 6th June 2008, 03:56 AM
cyue2000 cyue2000 is offline
Shy Poster (1 to 5 Posts)

Registration Date: Jan 2007
Location: China
 
Posts: 3
Thanks Given to Others: 23
Thanked 4 Times in 1 Post
Karma Power: 12
Karma: 10
cyue2000 has less than 100 Karma points so far.
Default Re: Biocompatibility testing & Registration in China

Hi Adele,

Attached please find the Chinse version.

The new regulation says:
(1) The company can choose (a) to do all the tests based on GB/T-16886-ISO 10993, or (b) to provide a biological evaluation report to justify why some of the tests are exempted.
(2) For tests done outside China, the company should provide supporting documents on GLP qualification of the test lab.

There are two attachments, which are guidelines for biological evaluation process and the biological evaluation report.

To answer your question, I would suggest do a gap analysis first for ISO 10993:2003 and the old version, to see if additional tests are needed. If the answer is yes, then just do it (and better to do it in China). Also, I believe you should have plenty of clinical data available as you are doing re-registeration. All of the clinical data can be included in your biological evaluation report to support the safety of your device.

Best,
Attached Files: 1. Scan for viruses before using, 2. Please report any 'bad' files by Reporting the post it is in, 3. Use at your Own Risk.
File Type: pdf 医疗器械生物学评价和审查.pdf (128.6 KB, 296 views)
Reply With Quote
Thank You to cyue2000 for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 6th June 2008, 04:38 AM
harry harry is offline
Super Moderator

Registration Date: Sep 2005
Location: Johore/Malaysia
 
Posts: 2,944
Thanks Given to Others: 571
Thanked 1,031 Times in 746 Posts
Blog Entries: 1
Karma Power: 231
Karma: 7783
harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.
harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.
Default Re: Biocompatibility testing & Registration in China

Welcome Cyue2000 and thanks for that informative first post. We look forward to see your active participation.
Reply With Quote
Thanks to harry for your informative Post and/or Attachment!
  #4  
Old 8th June 2008, 09:08 PM
Adele Adele is offline
Involved in Discussions

Registration Date: Oct 2005
Location: New Zealand
 
Posts: 47
Thanks Given to Others: 31
Thanked 18 Times in 12 Posts
Karma Power: 18
Karma: 97
Adele has less than 100 Karma points so far.
Default Re: Biocompatibility testing & Registration in China

Thanks cyue2000.

Is GB/T-16886-ISO 10993 based on ISO 10993-1: 2003?

We can easily write a biological evaluation report to explain the differences. However, we have the additional problem of the lab that was used not having GLP qualification. Again, this testing was done back in 2000 and 2001 and testing since that date has all been done in a GLP facility but we are left with some historical materials that were tested in a non-GLP facility.

Are we still able to meet this requirements by writing a biological evaluation report? Or does the lack of GLP certification mean we have no choice but to retest?

Thanks once again.
Reply With Quote
  #5  
Old 10th June 2008, 06:15 AM
liuyy liuyy is offline
Involved in Discussions

Registration Date: Mar 2004
Location: BEIJING,CHINA
 
Posts: 23
Thanks Given to Others: 0
Thanked 5 Times in 3 Posts
Karma Power: 27
Karma: 172
liuyy is appreciated, and has over 100 Karma points.liuyy is appreciated, and has over 100 Karma points.
Default Re: Biocompatibility testing & Registration in China

Quote:
Originally Posted by Adele View Post

Thanks cyue2000.

Is GB/T-16886-ISO 10993 based on ISO 10993-1: 2003?

We can easily write a biological evaluation report to explain the differences. However, we have the additional problem of the lab that was used not having GLP qualification. Again, this testing was done back in 2000 and 2001 and testing since that date has all been done in a GLP facility but we are left with some historical materials that were tested in a non-GLP facility.

Are we still able to meet this requirements by writing a biological evaluation report? Or does the lack of GLP certification mean we have no choice but to retest?

Thanks once again.
The GB/T 16886.X is identical to ISO 10993-X.
Reply With Quote
  #6  
Old 10th June 2008, 06:30 AM
cyue2000 cyue2000 is offline
Shy Poster (1 to 5 Posts)

Registration Date: Jan 2007
Location: China
 
Posts: 3
Thanks Given to Others: 23
Thanked 4 Times in 1 Post
Karma Power: 12
Karma: 10
cyue2000 has less than 100 Karma points so far.
Default Re: Biocompatibility testing & Registration in China

Yes. GB/T 16886.1 is identical to ISO 10993.1 As a matter of fact, only ISO 10993.1 is mentioned in attament 2, not ISO 10993.1:2003. It means the current version of ISO 10993.1 is always used.

You may want to check Attachment 1, part 4 - "re-evaluation of biological safety of medical device" carefully. It says that if the manufactuer can claim that (1) to (5) in the list doesn't happen, then no re-evaluation is required for re-registration of the device. The list includes (1) change of raw materials - supplier or technical specification; (2) changes in formulation, process condition,primary package and sterilization; (3) any change happened to the final product during storage; (4) change of intended use; (5) any indication of adverse effects.
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO and IEC Medical Device Related Standards > ISO 13485 - Medical Devices - Quality Management Systems

Bookmarks

Tags
biocompatibility testing, biological evaluation process, biological evaluation report, certification, china, iso 10993, registration


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Biocompatibility of Implantable Component - What testing, if any? WisdomseekerSC ISO 13485 - Medical Devices - Quality Management Systems 5 19th October 2009 01:02 PM
Animals for biocompatibility testing? Tiffany ISO 13485 - Medical Devices - Quality Management Systems 10 1st May 2009 12:01 PM
Biocompatibility testing centre in Montreal QC ? vaibhav_jami ISO 13485 - Medical Devices - Quality Management Systems 2 3rd December 2008 10:52 AM
Predicate Device - Question on Biocompatibility and IEC 60601-1 Testing Requirements melissa66 IEC 60601 - Medical Electrical Products Collateral Standards 4 28th March 2008 07:36 PM
Biocompatibility Testing for Implantable Medical Devices according to ISO 10993 Tiffany ISO 13485 - Medical Devices - Quality Management Systems 4 12th March 2008 10:14 PM



The time now is 09:20 AM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts