The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 4th July 2008, 02:11 PM
jleond jleond is offline
Getting Involved (6 to 9 Posts)

Registration Date: Dec 2006
Location: Hermosillo, Sonora-Mexico
Age: 46
 
Posts: 9
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 12
Karma: 10
jleond has less than 100 Karma points so far.
Question Can I use the CMM after the Calibration has expired?

The CMM calibration expires 7-19-08 and the service provider will start the new 2008-09 calibration on the week of 7-28-08. My question is: how long can the CMM be used to measure parts after the due date? I heard that we can use it (legally) 30 days after due date. Is this true? Is there a standard that states this or something related to this?

Thanks in advance for your help.
Reply With Quote

Sponsored Links
  #2  
Old 4th July 2008, 02:17 PM
Coury Ferguson's Avatar
Coury Ferguson Coury Ferguson is offline
Moderator here to help

Registration Date: Jan 2006
Location: SoCal
Age: 51
 
Posts: 3,839
Thanks Given to Others: 657
Thanked 460 Times in 395 Posts
Karma Power: 191
Karma: 4966
Coury Ferguson is appreciated, and has over 1700 Karma points.
Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.
Send a message via Skype™ to Coury Ferguson
Default Re: can I use the cmm after calibration expires?

Quote:
Originally Posted by jleond View Post

The CMM calibration expires 7-19-08 and the service provider will start the new 2008-09 calibration on the week of 7-28-08. My question is: how long can the CMM be used to measure parts after the due date? I heard that we can use it (legally) 30 days after due date. Is this true? Is there a standard that states this or something related to this?

Thanks in advance for your help.
In my opinion, I don't see a problem with it. I can't address the 30 days after rule.

A Suggestion: Change you Calibration Procedures, stickers and such to only reflect Month/Year only (example: 07/2008). This will help you in the long run.
__________________
"The one thing I want to leave my children is an honorable name." "It is hard to fail, but it is worse never to have tried to succeed." Theodore Roosevelt Chicago, IL, April 10, 1899
Reply With Quote
Sponsored Links

  #3  
Old 4th July 2008, 02:19 PM
BradM's Avatar
BradM BradM is offline
Super Moderator

Registration Date: Sep 2006
Location: Arlington,Texas
 
Posts: 3,384
Thanks Given to Others: 814
Thanked 850 Times in 607 Posts
Karma Power: 274
Karma: 9659
BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.
BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.
Default Re: can I use the cmm after calibration expires?

Hello there!

In general, calibration intervals are much more of an Art than a Science. The perfect calibration interval is that which find error 50-75% of the unacceptable point, at least that is my opinion.

To your point, it would all depend on what your quality management system states. Do you have procedures that states when calibration is past due?

Without anything documented in your system or anything in writing, it does not look good to use any instrument beyond its due date.

The best thing for you do do is establish a calibration procedure for when instruments are recalled for calibration (like month/year), using an instrument past its due date, tagging, etc. Like in your example, I would make the due date 07/08. That way you have the month to get the work done.

Hopefully that helps.
__________________
Brad
My idea of housework is to sweep the room with a glance.
Reply With Quote
  #4  
Old 4th July 2008, 02:23 PM
BradM's Avatar
BradM BradM is offline
Super Moderator

Registration Date: Sep 2006
Location: Arlington,Texas
 
Posts: 3,384
Thanks Given to Others: 814
Thanked 850 Times in 607 Posts
Karma Power: 274
Karma: 9659
BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.
BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.
Default Re: can I use the cmm after calibration expires?

Quote:
Originally Posted by Coury Ferguson View Post

In my opinion, I don't see a problem with it. I can't address the 30 days after rule.
Actually, I agree with you. But... there are many auditors who are not as processed-oriented as they should. They would see 7/19/2008, look at their watch and see 7/22/2008, the absence of any procedural exception, and they start writing.

The best thing to do is write up some exception/ some report stating that your organization says it's OK to use, IMHO.

As far as Rule of Thumbs (30 days, etc.), I have seen them all over the place, and again depends on the industry and the people involved.
__________________
Brad
My idea of housework is to sweep the room with a glance.
Reply With Quote
  #5  
Old 4th July 2008, 02:27 PM
Coury Ferguson's Avatar
Coury Ferguson Coury Ferguson is offline
Moderator here to help

Registration Date: Jan 2006
Location: SoCal
Age: 51
 
Posts: 3,839
Thanks Given to Others: 657
Thanked 460 Times in 395 Posts
Karma Power: 191
Karma: 4966
Coury Ferguson is appreciated, and has over 1700 Karma points.
Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.
Send a message via Skype™ to Coury Ferguson
Default Re: can I use the cmm after calibration expires?

Quote:
Originally Posted by BradM View Post

Actually, I agree with you. But... there are many auditors who are not as processed-oriented as they should. They would see 7/19/2008, look at their watch and see 7/22/2008, the absence of any procedural exception, and they start writing.

The best thing to do is write up some exception/ some report stating that your organization says it's OK to use, IMHO.

As far as Rule of Thumbs (30 days, etc.), I have seen them all over the place, and again depends on the industry and the people involved.
If an Auditor is choosing to be that picky, they need to re-evaluate their techniques. We are talking a couple of days, still within the same month of calibration due date.

I would challenge any Auditor that would write this up as an NC.
__________________
"The one thing I want to leave my children is an honorable name." "It is hard to fail, but it is worse never to have tried to succeed." Theodore Roosevelt Chicago, IL, April 10, 1899
Reply With Quote
  #6  
Old 4th July 2008, 02:35 PM
Stijloor's Avatar
Stijloor Stijloor is offline
Cross Forum Moderator

Registration Date: May 2003
Location: Charlotte, North Carolina.
Age: 61
 
Posts: 6,626
Thanks Given to Others: 1,318
Thanked 2,233 Times in 1,596 Posts
Karma Power: 377
Karma: 10939
Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.
Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.Stijloor is appreciated, and has over 1700 Karma points.
Default Re: can I use the cmm after calibration expires?

Quote:
Originally Posted by Coury Ferguson View Post

If an Auditor is choosing to be that picky, they need to re-evaluate their techniques. We are talking a couple of days, still within the same month of calibration due date.

I would challenge any Auditor that would write this up as an NC.
If I were the auditor, I would look at other instances where devices exceed the stated calibration due date (systemic issue). If this is indeed the case, I would write an NC.

Now, having said that, if you allow deviations from an established requirement, even if it's only one device, where does one "draw the line?"

What do you think?

Stijloor.
Reply With Quote
  #7  
Old 4th July 2008, 02:42 PM
Coury Ferguson's Avatar
Coury Ferguson Coury Ferguson is offline
Moderator here to help

Registration Date: Jan 2006
Location: SoCal
Age: 51
 
Posts: 3,839
Thanks Given to Others: 657
Thanked 460 Times in 395 Posts
Karma Power: 191
Karma: 4966
Coury Ferguson is appreciated, and has over 1700 Karma points.
Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.
Send a message via Skype™ to Coury Ferguson
Default Re: can I use the cmm after calibration expires?

Quote:
Originally Posted by Stijloor View Post

If I were the auditor, I would look at other instances where devices exceed the stated calibration due date (systemic issue). If this is indeed the case, I would write an NC.

Now, having said that, if you allow deviations from an established requirement, even if it's only one device, where does one "draw the line?"

What do you think?

Stijloor.
Alright I see you point. I don't totally agree with, but I understand where this is coming from. (added: if it is a systematic problem then I would identify it as a Major NC, but I think we are talking about one CMM)

My question back is: Is there not a master diametrical sphere to check the accuracy of the CMM? And this diametrical sphere is calibrated to be accurate within millionths or even tenths.

That is why I suggested only identifying the stickers with Month/Year would help in the long run.
__________________
"The one thing I want to leave my children is an honorable name." "It is hard to fail, but it is worse never to have tried to succeed." Theodore Roosevelt Chicago, IL, April 10, 1899

Last edited by Coury Ferguson; 4th July 2008 at 02:49 PM. Reason: Added in blue font
Reply With Quote
  #8  
Old 4th July 2008, 02:43 PM
AndyN's Avatar
AndyN AndyN is offline
Forum Moderator

Registration Date: Feb 2005
Location: Michigan, USA
 
Posts: 3,880
Thanks Given to Others: 1,786
Thanked 1,259 Times in 902 Posts
Karma Power: 297
Karma: 10030
AndyN is appreciated, and has over 1700 Karma points.AndyN is appreciated, and has over 1700 Karma points.AndyN is appreciated, and has over 1700 Karma points.
AndyN is appreciated, and has over 1700 Karma points.AndyN is appreciated, and has over 1700 Karma points.AndyN is appreciated, and has over 1700 Karma points.AndyN is appreciated, and has over 1700 Karma points.AndyN is appreciated, and has over 1700 Karma points.AndyN is appreciated, and has over 1700 Karma points.AndyN is appreciated, and has over 1700 Karma points.AndyN is appreciated, and has over 1700 Karma points.
Let Me Help You Re: Can I use the CMM after the Calibration has expired?

Firstly, there's nothing 'legal' about it - literally or metaphorically. You should have past data that shows for each calibration, what the 'as found' condition was. Compared to the specification limits and the tolerance of the features measured on the cmm, you should be able to write yourself an 'extension' to the recall, based on DATA and the decision you made, based on that.

If your external auditor gives you grief, roll up the deviation, coat the end in Vaseline, and.........

Of course, make sure your extension form is a) mentioned in your procedures etc and b) is a controlled document.

Go for it!!
__________________
'Cause you know sometimes words have two meanings.....
Reply With Quote
Thanks to AndyN for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)

Bookmarks

Tags
calibration, calibration past due date, cmm


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Calibration of ring gages with a CMM merrick65 ISO 17025 - Calibration, Measurement Gages and Test Laboratories 3 23rd July 2008 11:23 PM
CMM Qualifying Sphere Calibration KosmoQA ISO 17025 - Calibration, Measurement Gages and Test Laboratories 9 26th March 2006 09:22 PM
Understanding Calibration Results on CMM Reports lee01 SPC Monitoring and Statistical Analysis Techniques 2 2nd January 2003 04:35 PM
Understanding calibration results on CMM reports lee01 ISO 17025 - Calibration, Measurement Gages and Test Laboratories 4 30th December 2002 11:56 AM
CMM Calibration Certification - Small CMM Calibration Service qts4u ISO 17025 - Calibration, Measurement Gages and Test Laboratories 5 26th August 1998 09:33 AM



The time now is 11:47 PM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts