The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > ISO 14971 - Medical Device Risk Management
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


Related Topic Tags
class ii medical device, fmeca, medical device(s), iso 14971 - medical device risk management, medical devices (general), risk management and analysis
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 3rd December 2008, 12:18 AM
tantan tantan is offline
Involved in Discussions

 
Registration Date: Nov 2006
Location: uk
 
Posts: 19
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 28
Karma: 10
tantan has less than 100 Karma points so far.
Please Help! Best Risk Analysis Approach (ISO 14971) - Class IIa Medical Devices

Hi,

I just wanted to get some feedback from any of you there familiar with the application of risk management (and more specifically risk analysis). We are a specifications developer for Class IIa medical devices. I am new to the company and have noticed that they have only previously done an analysis (FMECA) but with no real management process. Ive read in some articles that FMEA is better for Class III or safety critical parts of a device, and that different analysis methods should be used at different stages of design control. Any feedback on what Ive read or what others are doing would be muchly appreciated!

Cheers

Tan :-)

Sponsored Links
  #2  
Old 3rd December 2008, 09:29 AM
Coury Ferguson's Avatar
Coury Ferguson Coury Ferguson is offline
Moderator here to help

 
Registration Date: Jan 2006
Location: SoCal
Age: 55
 
Posts: 4,322
Thanks Given to Others: 941
Thanked 620 Times in 510 Posts
Karma Power: 473
Karma: 5820
Coury Ferguson is appreciated, and has over 1700 Karma points.
Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.Coury Ferguson is appreciated, and has over 1700 Karma points.
Re: Best Risk Analysis Approach (ISO 14971) - Class IIa medical devices

Has anyone performed a Risk Analysis on the Class II Medical device? Any comments/suggestions?
__________________
"The one thing I want to leave my children is an honorable name." "It is hard to fail, but it is worse never to have tried to succeed." Theodore Roosevelt Chicago, IL, April 10, 1899
Sponsored Links

  #3  
Old 3rd December 2008, 01:47 PM
MIREGMGR MIREGMGR is offline
Appreciated Information Resource

 
Registration Date: Aug 2008
Location: SE Michigan, USA
 
Posts: 2,843
Thanks Given to Others: 1,036
Thanked 2,031 Times in 1,271 Posts
Karma Power: 351
Karma: 23509
MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.
MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.
Re: Best Risk Analysis Approach (ISO 14971) - Class IIa medical devices

We always focus more on whether we have input from the right mix of medical procedural experts and device-type-technology historical- and theoretical-failure-potential experts, than we do the formal failure analysis technique and format.
  #4  
Old 3rd December 2008, 10:28 PM
gholland gholland is offline
Involved in Discussions

 
Registration Date: May 2008
Location: Denver, CO
 
Posts: 95
Thanks Given to Others: 15
Thanked 53 Times in 40 Posts
Karma Power: 31
Karma: 614
gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.
Re: Best Risk Analysis Approach (ISO 14971) - Class IIa medical devices

Here are the basics of what we do as a Class II/IIa medical device company. As context we are recovering from a 483 letter and our risk analysis procedure is now <very> thorough. It may be overkill but I know it has the blessing of FDA:

- Develop your Top Level Hazards (TLH). These are the risks to the user, patient, and caregiver. Develop your severity rankings for these TLHs here. The severities do not change and are carried through the entire analysis.

- From your TLHs you develop your basic fault events (BFEs). Basically 'how did my TLH happen?'. Develop your Likelihood of failure here. The multiplication of the severity and the likelihood gives your Risk level. See ISO 14971 for examples.

- Develop a fault tree analysis from this information. Your failure modes leading to your BFEs leading to your TLHs. You can then develop your Risk Control Measures (RCMs) from this information. These RCMs are what you would track and Verify during testing. Basically do your RCMs work as intended to reduce the risk of the BFE? Your RCMs are what you use to reduce the Likelihood number in your Risk analysis. If your RCMs are not effective you are not justified in reducing your Likelihood number, but if you can prove they are effective then you're OK.

- This completes your fault tree. This is a top down approach, i.e. what is my failure and how did I get there? From here you would then go the other way for dFMEA and pFMEA. For your dFMEA you would take each component and ask yourself "if this fails does it lead to a TLH?". If the answer is no then move on to the next component. For you pFMEA you ask yourself "if this process fails can it lead to a TLH?". If no then move on. You would develop more risk control measures from your FMEA if you have missed anything in the FTA.

An example of a TLH would be 'Patient receives a non-sterile infusion' or something similar. A BFE would be 'Packaging failure due to xxx causes sterile barrier breach'. A RCM would be 'Packaging tested to ISTA 2A requirements'. You would then be able to establish traceability from your packaging test to your risk analysis.

Thank You to gholland for your informative Post and/or Attachment!
  #5  
Old 22nd January 2009, 12:19 AM
Deepoz Deepoz is offline
Shy Poster (1 to 5 Posts)

 
Registration Date: Jan 2009
Location: Newcastle, Australia
 
Posts: 4
Thanks Given to Others: 3
Thanked 15 Times in 1 Post
Karma Power: 18
Karma: 85
Deepoz has less than 100 Karma points so far.
Read This! Re: Best Risk Analysis Approach (ISO 14971) - Class IIa medical devices

Hi
We manufacture custom-made sterile Procedure packs, some of which are class IIb due to components used within. I have developed a risk management checklist as per the standards which lists hazards and control measures implemented. We generate similar checklist for all products individually. Since there are no design issues involved there are only a few hazards identified but I have attached a copy of one of our checklists if it helps. It is an excel sheet with links to our component database which makes life easy for generating any new product checklist. It has been audited and accepted by Australian regulator (TGA).
Cheers
Attached Files: 1. Scan for viruses before using, 2. Please report any 'bad' files by Reporting this post, 3. Use at your Own Risk.
File Type: pdf FG00084_V1_Risk Checklist.pdf (35.7 KB, 1457 views)
Thank You to Deepoz for your informative Post and/or Attachment!
  #6  
Old 22nd January 2009, 01:56 AM
tantan tantan is offline
Involved in Discussions

 
Registration Date: Nov 2006
Location: uk
 
Posts: 19
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 28
Karma: 10
tantan has less than 100 Karma points so far.
Re: Best Risk Analysis Approach (ISO 14971) - Class IIa medical devices

Thanks for the attachment! That looks more like the direction I want our risk analysis to head in
  #7  
Old 22nd January 2009, 04:05 AM
Ajit Basrur's Avatar
Ajit Basrur Ajit Basrur is offline
Forum Administrator

 
Registration Date: Dec 2005
Location: Massachusetts, USA
Age: 49
 
Posts: 5,123
Thanks Given to Others: 1,471
Thanked 2,214 Times in 1,427 Posts
Karma Power: 400
Karma: 13713
Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.
Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.Ajit Basrur is appreciated, and has over 1700 Karma points.
Re: Best Risk Analysis Approach (ISO 14971) - Class IIa medical devices

Quote:
In Reply to Parent Post by Deepoz View Post

Hi
We manufacture custom-made sterile Procedure packs, some of which are class IIb due to components used within. I have developed a risk management checklist as per the standards which lists hazards and control measures implemented. We generate similar checklist for all products individually. Since there are no design issues involved there are only a few hazards identified but I have attached a copy of one of our checklists if it helps. It is an excel sheet with links to our component database which makes life easy for generating any new product checklist. It has been audited and accepted by Australian regulator (TGA).
Cheers
Great attachment and welcome to the Cove

Hope to see your active participation
__________________
Well done is better than well said - Benjamin Franklin
Thanks to Ajit Basrur for your informative Post and/or Attachment!
  #8  
Old 22nd January 2009, 04:20 PM
Deepoz Deepoz is offline
Shy Poster (1 to 5 Posts)

 
Registration Date: Jan 2009
Location: Newcastle, Australia
 
Posts: 4
Thanks Given to Others: 3
Thanked 15 Times in 1 Post
Karma Power: 18
Karma: 85
Deepoz has less than 100 Karma points so far.
Re: Best Risk Analysis Approach (ISO 14971) - Class IIa medical devices

no worries
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > ISO 14971 - Medical Device Risk Management

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Application of Risk Management - ISO 14971:2007(E) Medical Devices WBubel - 2010 ISO 14971 - Medical Device Risk Management 69 31st October 2012 12:42 PM
ISO 14971 Class II Medical Devices - Product Realization & Risk Management qremsg26 ISO 14971 - Medical Device Risk Management 5 26th February 2011 07:53 AM
ISO 14971:2007 Risk Management - Class I A Sterile Medical Device romit singh ISO 14971 - Medical Device Risk Management 4 29th March 2010 04:36 PM
ISO 14971:2007 Application of Risk Management to Medical Devices vaibhav_jami ISO 14971 - Medical Device Risk Management 9 23rd April 2009 08:07 AM
Applicability of Risk Management for Medical Devices - ISO 14971:2000 issmileplz ISO 14971 - Medical Device Risk Management 20 12th March 2009 02:01 PM



The time now is 12:08 AM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


   


Marc Timothy Smith - Elsmar.com
8466 LeSourdsville-West Chester Road, Olde West Chester, Ohio 45069-1929
513 341-6272