The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


Related Topic Tags
thermal profile, vacuum chamber, vacuum oven, ovens
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 23rd January 2009, 09:03 AM
sledesma sledesma is offline
Inactive Registered Visitor

 
Registration Date: Jan 2009
Location: Argentina
 
Posts: 4
Thanks Given to Others: 3
Thanked 0 Times in 0 Posts
Karma Power: 18
Karma: 10
sledesma has less than 100 Karma points so far.
Thank You! Oven Qualification - Thermal profile during the OQ of a vacuum oven

Dear all,
I am writing to ask for some help in analysing the results obtained in a thermal profile during the OQ of a vacuum oven. It was carried out without appling vacuum, only thermal profile, with the set point at 105 centigrades (analog controler). The oven is 16 litres of capacity, and the test was carried out using 12 calibrated thermocouples during 2 hs (stability). The specification of the termal range is 103-107 C because the oven is used for API's loss on drying determination. 3 of the thermocouples have a medium temperature of 120 centigrades, they were near the top, 2 of them that where near the glass door have a medium temperature of 100 C and the other 7 have a medium temperature of 110 C. I want to know if the equipment have a problem or may be the test was not properly designed. What do you think? Is it common to have these differences in temperature profile in this kind of equipment? Thank you very much. Regards from Argentina!

Last edited by Al Rosen; 23rd January 2009 at 12:06 PM.

Sponsored Links
  #2  
Old 23rd January 2009, 02:45 PM
Kevin H's Avatar
Kevin H Kevin H is offline
Lab Manager

 
Registration Date: Jun 2004
Location: Western Pennsylvania, USA
Age: 60
 
Posts: 489
Thanks Given to Others: 19
Thanked 50 Times in 40 Posts
Karma Power: 87
Karma: 1203
Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.Kevin H is appreciated, and has over 1100 Karma points.
Re: Oven Qualification

Sledesma, I'm not surprised at the temperature variation in the thermocouples at that low temperature. I'm guessing that the oven is heated electrically - and that you're relying primarily on radiation for heat transfer. (Actually with air in the furnace you'll get some heat transfer via convection as well). Radiation only becomes effective at high temperatures - say 1800 degrees F or 980 degrees C. I used to heat treat precipitation hardening stainless steel tensile specimens at 900 F to verify capability. I had to be at +/- 15 F through out the furnace - a furnace without atmosphere, just air, that was heated via electrical resistance. After doing a temperature uniformity study we had to restrict the acceotable working volume to about 1/3 of the actural furnace working volume to meet those requirements.
At the solutioning temperatures, about 1750 F if I remember correctly temperature uniformity was fine. Our final solution to the temperature issue was to purchase a furnace with a builtin fan to circulate the air iresulting in transfer via forced convection n the furnace so that were more uniform and had better heat transfer, not relying just on radiation and natural convection.

You'll probably see even worse uniformity if you test under vacuum, which I believe you should do, as conformity tests should reflect working conditions. And yes, doors, even when insulated are usually cooler. I'm not a furnace designer, just a user, so I can't help with your thermal uniformity problems. - One possibility would be to do what we did, i.e., restrict the usable volume of your oven to that which meets the temperature variation requirements of your test.

The uniformity that you're looking for at the low temperatures you're measuring will be extremely challenging to obtain in a vacuum oven. Best of luck on getting them.
__________________
" We have met the enemy and he is us." Pogo
Thanks to Kevin H for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 23rd January 2009, 03:43 PM
MIREGMGR MIREGMGR is offline
Appreciated Information Resource

 
Registration Date: Aug 2008
Location: SE Michigan, USA
 
Posts: 2,843
Thanks Given to Others: 1,036
Thanked 2,028 Times in 1,269 Posts
Karma Power: 350
Karma: 23494
MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.
MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.MIREGMGR is appreciated, and has over 1700 Karma points.
Re: Oven Qualification

With only radiation as a heat transfer mechanism, what is the meaning of a reading from a thermocouple located away from the part? What is it measuring?

If the thermocouple is conductively coupled to an oven enclosure element, it's measuring the temperature of that element...which may have practically nothing to do with the spatial evenness of the radiation pattern, if in fact that would result in the desired part temperature profile.

If your heating mechanism is solely radiation and you're using thermocouples to measure temperature, they should be conductively attached to either a test part or a specially engineered object that's located where the production parts will go, and has appropriate characteristics to heat in the same pattern as the known actual heating pattern of the production parts. The thermocouples also should be locally radiatively shielded, so they measure the test part temperature via only conduction, without consideration of their own radiative heating characteristics.
Thanks to MIREGMGR for your informative Post and/or Attachment!
  #4  
Old 26th January 2009, 07:02 AM
sledesma sledesma is offline
Inactive Registered Visitor

 
Registration Date: Jan 2009
Location: Argentina
 
Posts: 4
Thanks Given to Others: 3
Thanked 0 Times in 0 Posts
Karma Power: 18
Karma: 10
sledesma has less than 100 Karma points so far.
Re: Oven Qualification

Yes Kevin H, this oven works with electrical resistences, and the test was carried out without applying vacuum, but the chamber is hermetically closed. So you said that in that condition the primary heat transfer is because of radiation. And so that the homogeneity will be poor at that temperature. Thank you very much for your contribution! I think the problem is being clarified!
  #5  
Old 26th January 2009, 07:03 AM
sledesma sledesma is offline
Inactive Registered Visitor

 
Registration Date: Jan 2009
Location: Argentina
 
Posts: 4
Thanks Given to Others: 3
Thanked 0 Times in 0 Posts
Karma Power: 18
Karma: 10
sledesma has less than 100 Karma points so far.
Re: Oven Qualification

Thank you very much Miregmgr! Now I will only concentrate in the results of the TC that are near the working surface!
  #6  
Old 26th January 2009, 08:00 AM
BradM's Avatar
BradM BradM is offline
Forum Administrator

 
Registration Date: Sep 2006
Location: Arlington,Texas
 
Posts: 5,343
Thanks Given to Others: 1,302
Thanked 1,816 Times in 1,226 Posts
Karma Power: 400
Karma: 16664
BradM is appreciated, and has over 1700 Karma points.
BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.BradM is appreciated, and has over 1700 Karma points.
Re: Oven Qualification - Thermal profile during the OQ of a vacuum oven

Hello there!

First, it sounds to me that your temperature specification is way too tight. Even with high quality Type T thermocouples you have +/-.75 C tolerance on each wire. So, you can have a total of 1.5 C deviation. That is not including any error within the recording device and other uncertainties along the way.

BTW... what kind of specifications were provided with the oven? If the oven was not designed for your particular uniformity, it may not ever be able to achieve it. Ovens have to have very high air exchanges per minute to achieve that type of uniformity.

What kind of oven is this? Is it a recirculating-air type oven, or is it constant temperature? If it is recirculating, there is a discharge and a return air. Generally the discharge side will be the hottest. Also, heat rises, so the top of an oven will be hotter than the bottom.

If this is a constant temperature vacuum oven and your process will involve vacuum, you really should consider running the profile with the vacuum on.

I hope something here helps.
__________________
Now, 75 years later in an abundant society where people have laptops, cell phones, iPods and minds like empty rooms, I still plod along with books. ”
—Harper Lee
Thanks to BradM for your informative Post and/or Attachment!
  #7  
Old 26th January 2009, 08:43 AM
sledesma sledesma is offline
Inactive Registered Visitor

 
Registration Date: Jan 2009
Location: Argentina
 
Posts: 4
Thanks Given to Others: 3
Thanked 0 Times in 0 Posts
Karma Power: 18
Karma: 10
sledesma has less than 100 Karma points so far.
Re: Oven Qualification - Thermal profile during the OQ of a vacuum oven

Thank you Brad! Your comments are really helpful!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - General Metrology, Measurement Device, Calibration and Test Laboratories > ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Oven Validation Protocol (procedure) needed TrucLe ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 19 31st December 2012 09:21 AM
Bench Top Oven Calibration Standards neoth ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 8 4th December 2009 09:32 PM
Oven Calibration Range: How many points? Charles Wathen ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 10 8th October 2008 10:51 PM
Need Help to Verify an Infrared Cure Oven Adriane ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 9 7th December 2005 10:38 PM
Leak Detection - Oven Seals Welty ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 1 13th August 2002 03:41 PM



The time now is 04:59 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


   


Marc Timothy Smith - Elsmar.com
8466 LeSourdsville-West Chester Road, Olde West Chester, Ohio 45069-1929
513 341-6272