CE Mark Myths - The Medical Device Directive's Seven Major Myths
From Quality Digest:
The Medical Device Directive's Seven Major Myths
Mystified by the European Union's Medical Device Directive? The following seven points touch on some of the directive's persistent myths and how it will affect manufacturers:
Myth No. 1: CE marking is a trade barrier. Fact: The European Union views the CE Mark much like a passport. It is a manufacturer's claim that its product meets all applicable directives in force at that time. Member states must allow CE-marked products free market access.
Myth No. 2: The June 14, 1998, deadline was extended. Fact: There was no extension, and the current deadline was enforced.
Myth No. 3: All products placed on the market prior to the June 14, 1998, deadline will be exempt from CE marking until 2001. Fact: This is still only a proposal. Even if it is accepted, companies are finding sales negatively affected by the CE Mark's absence.
Myth No. 4: Countries in the EU can impose additional requirements within the directive. Fact: Countries must give products legally displaying the CE Mark free access to their markets. Member states may only restrict products that endanger the public.
Myth No. 5: ISO 9000 is sufficient to meet the directive's Annex II. Fact: In essence, EN 46001 is ISO 9001 plus specific medical requirements. Those additional quality items are found in EN 46001 and Annex II.
Myth No. 6: A manufacturer needs only a Declaration of Conformity for Class I Devices. Fact: Items required by the MDD's Annex VII include the technical file, post-market vigilance, risk analysis, and essential requirements for health and safety.
Myth No. 7: All 15 recognized European languages must appear on a product's label to meet EU standards. Fact: It depends on what the product is and where it is sold. Some member countries require their languages on the labels of any products marketed within their borders. The audience the product is meant for also must be considered.
For more information, call the MDD Hot Line at telephone (800) 810-9000.
(Excerpted from European Community Quarterly Review, Volume VI, Issue 2, Technology International Inc., 1998.)
|

8th October 1998, 07:55 PM
|
|
Courtesy Access
Registration Date: Dec 1998
Location: South Central Massachusetts
|
|
Posts: 788
Thanks Given to Others: 0
Thanked 15 Times in 12 Posts
Karma Power: 63 Karma: 151  
|
|
Nice Post...once again lets hear it for Marc !!!!!!
|

13th October 1998, 09:26 PM
|
 |
dWizard
Registration Date: Nov 1998
Location: Tullahoma, TN
Age: 53
|
|
Posts: 475
Thanks Given to Others: 0
Thanked 12 Times in 9 Posts
Karma Power: 63
|
|
Agreed, good post. I would like to add another common one (or at least one I have encountered often).
To obtain a CE Mark for a medical device, your firm must be audited by a notified body. This is not true. That varies with the device and class.
Best Regards,
Don
|

28th July 2006, 08:39 AM
|
 |
Your Elsmar Cove Host
Registration Date: Jan 1996
Location: West Chester, Ohio - USA
Age: 59
|
|
Posts: 15,857
Thanks Given to Others: 1,895
Thanked 1,566 Times in 1,018 Posts
Karma Power: 605
|
|
Re: CE Mark Myths - The Medical Device Directive's Seven Major Myths
Quote:
|
Originally Posted by Don Winton
To obtain a CE Mark for a medical device, your firm must be audited by a notified body. This is not true. That varies with the device and class.
|
Is this still true? Can anyone explain or expand upon this?
__________________
A Search is a terrible thing to waste!
One Test is Worth 1000 Expert Opinions - The plural of anecdote is not data.
We can't solve problems by using the same kind of thinking we used when we created them. - Unknown
|

28th July 2006, 08:51 AM
|
 |
Forum Moderator
Registration Date: Jun 2002
Location: Lawn Guyland
Age: 59
|
|
Posts: 3,101
Thanks Given to Others: 48
Thanked 390 Times in 272 Posts
Karma Power: 192
|
|
Re: CE Mark Myths - The Medical Device Directive's Seven Major Myths
Quote:
|
Originally Posted by Marc
Is this still true? Can anyone explain or expand upon this?
|
I believe non sterilized class I devices do not require a notified body. (I know that's a short answer, but it's all I have time for right now.)
__________________
Al
|

14th August 2006, 02:16 PM
|
|
Registered User
Registration Date: Sep 2005
Location: Houston (ex-pat Brit)
|
|
Posts: 531
Thanks Given to Others: 37
Thanked 222 Times in 159 Posts
Karma Power: 47
|
|
Re: CE Mark Myths - The Medical Device Directive's Seven Major Myths
Wonderful article, even though it is is seriously old, it is still mostly valid. I've added some of my own thoughts here.
Myth No. 1: CE marking is a trade barrier. Fact: The European Union views the CE Mark much like a passport. It is a manufacturer's claim that its product meets all applicable directives in force at that time. Member states must allow CE-marked products free market access.
Mostly true. Each country has the right to invoke the safeguard clause for any device they don't like. There's a few other wierd exceptions as well.
Myth No. 2: The June 14, 1998, deadline was extended. Fact: There was no extension, and the current deadline was enforced.
Oh so true, and may be true once more for 2005/50 (reclassification of hip/knee and shoulder implants) when the deadline for that is reached.
Myth No. 3: All products placed on the market prior to the June 14, 1998, deadline will be exempt from CE marking until 2001. Fact: This is still only a proposal. Even if it is accepted, companies are finding sales negatively affected by the CE Mark's absence.
It was an exciting time.
CE Marking was accepted very fast, and many companies that were "in the right" from a regulatory perspective, nevertheless got totally hammered by their customers who only wanted CE marked product.
Myth No. 4: Countries in the EU can impose additional requirements within the directive. Fact: Countries must give products legally displaying the CE Mark free access to their markets. Member states may only restrict products that endanger the public.
Mostly true - however many countries bolted on all sorts of interesting "local colour", when the Directive got transposed into their individual national legislations. Usually only us geeks need to worry about the differences, but every now and again...
There is talk that future versions of the European regulations will not be in the same Directive format, to avoid this very thing - one set of fixed rules for all.
Myth No. 5: ISO 9000 is sufficient to meet the directive's Annex II. Fact: In essence, EN 46001 is ISO 9001 plus specific medical requirements. Those additional quality items are found in EN 46001 and Annex II.
Mostly true - for some (very few) companies ISO9001 may be accepted, or perhaps a system that met the QSR. The European regulatory procedures etc are non-negotiable
Myth No. 6: A manufacturer needs only a Declaration of Conformity for Class I Devices. Fact: Items required by the MDD's Annex VII include the technical file, post-market vigilance, risk analysis, and essential requirements for health and safety.
Still true. Note of course that there are additional requirements if the Class I device is sterile or has a measuring function (or both).
Myth No. 7: All 15 recognized European languages must appear on a product's label to meet EU standards. Fact: It depends on what the product is and where it is sold. Some member countries require their languages on the labels of any products marketed within their borders. The audience the product is meant for also must be considered.
Still true, though I think the number of possible languages is approaching 20...
To answer Al's question, for "straight" Class I devices not only do you not need a Notified Body, you don't even need a quality system!
Annex VII has no such requirement, except for devices that are either sterile or have a measuring function in which case Annex V must also be followed. (The new revision of the MDD will allow Annex II also, a well-intentioned goof is why this is not an option currently).
Of course implementing a ISO13485:2003 system is recommended.
To go a little further, devices for clinical investigation, and custom-made medical devices, are outside the usual system and do not need Notified Body intervention. (Annex VIII).
Article 12 is a bit of a nightmare, but you could also potentially create a procedure pack or system without a Notified Body being involved.
|

16th August 2006, 01:11 AM
|
 |
Quality Manager
Registration Date: Sep 2003
Location: Illinois
|
|
Posts: 7,534
Thanks Given to Others: 182
Thanked 1,127 Times in 762 Posts
Karma Power: 397
|
|
Re: CE Mark Myths - The Medical Device Directive's Seven Major Myths
In regard to medical devices - sometimes governments may have a different idea which class a product fits in than what the manufacturer thinks or uses in marketing. In such cases, governments willingly throw previously "grandfathered" products into the pool, requiring 3rd party certification to ISO13485 before allowing sale of previously allowed products.
The fairness of such a procedure is easy to visulaize if you consider whether the original formula for Coca Cola (with real cocaine) should be continued for sale merely because it was once allowed 100 years ago. (Or whether some of the machines in the Quack Medicine museum should be allowed to be sold because they were once allowed.)
__________________
"Few minds wear out; more rust out"
Inscribed over the entrance of Louis Pasteur School, Chicago
Christian Nestell Bovee (1820-1904) in Thoughts, Feelings and Fancies, 1857
|

16th August 2006, 05:32 AM
|
|
Registered User
Registration Date: Sep 2005
Location: Houston (ex-pat Brit)
|
|
Posts: 531
Thanks Given to Others: 37
Thanked 222 Times in 159 Posts
Karma Power: 47
|
|
Re: CE Mark Myths - The Medical Device Directive's Seven Major Myths
Hi Wes, good points.
One of the big developments within Europe at the moment is the introduction of Directive 2005/50, making hip, knee and shoulder joint replacement implants highest-risk (Class III, from medium-high risk IIb). That has been done precisely because the grandfather products are seen by the authorities as NOT being adequately trustworthy.
A pseudo-myth related to your post above is that because a device has US FDA 510(k) approval it will automatically be accepted for EC certification.
These things tend to vary from one Notified Body to another, but essentially grandfathering, unsupported statements of equivalency, or approvals from countries outside the EU won't be blindly accepted.
Each device should stand on its own merits, in terms of evidence for safety, performance, and clinical benefit.
Of course for most descendant or copycat products, there IS data available to make those justifications possible - either already held by the applicant, or within the public domain. (Of course that's not the automatically the same thing as the data being presented clearly within the technical file!  )
|
Lower Navigation Bar
|
|
|
|
Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
|
|
|
| Thread Tools |
Search this Thread |
|
|
|
| Display Modes |
Rate Thread Content |
Linear Mode
|
|
Posting Settings
|
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
HTML code is Off
|
|
|
|
|