The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 19th May 2009, 02:10 AM
sharduli_k sharduli_k is offline
Getting Involved (6 to 9 Posts)

Registration Date: May 2009
Location: India
 
Posts: 9
Thanks Given to Others: 2
Thanked 0 Times in 0 Posts
Karma Power: 3
Karma: 10
sharduli_k has less than 100 Karma points so far.
Please Help! What to do with Prior Analysis after instrument was Out Of Calibration (OOC)?

If an instrument is OOC from unknown days before and analyst could not encounter the error as it is much internal and critical to get,what to do with the samples that run on the instrument from some before days?
How one can get the time from when it was out of calibration?
Reply With Quote

Sponsored Links
  #2  
Old 19th May 2009, 07:46 AM
brahmaiah brahmaiah is offline
Quality veteran

Registration Date: Mar 2009
Location: india
 
Posts: 383
Thanks Given to Others: 10
Thanked 68 Times in 55 Posts
Blog Entries: 3
Karma Power: 18
Karma: 402
brahmaiah is appreciated, and has over 400 Karma points.brahmaiah is appreciated, and has over 400 Karma points.brahmaiah is appreciated, and has over 400 Karma points.brahmaiah is appreciated, and has over 400 Karma points.brahmaiah is appreciated, and has over 400 Karma points.
Default Re: What to do with Prior Analysis after instrument was Out Of Calibration (OOC)?

The foolproof method to correct the situation, is :
1)Containment action:
a)Stop production untill you get a correct instrument
b) inspect all the parts available in your custody.
c)Send an express message to your customer to hold all parts available with him
d)Rush to your customer premises and segregate those components.

2)Corrective action:
a)Calibrate the instrument in question and start production.

3)Prevention:
a)Calibrate more frequently
b)Check the instrument before each time it is used as we do in case of hardness testers and micrometers.
c)Introduce a 'Poka Yoke' if possible.

I HOPE THIS IS ENOUGH!.
V.J.Brahmaiah
Reply With Quote
Thank You to brahmaiah for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 19th May 2009, 08:26 AM
trainerbob trainerbob is offline
Involved in Discussions

Registration Date: Sep 2008
Location: Michigan, USA
 
Posts: 86
Thanks Given to Others: 27
Thanked 39 Times in 25 Posts
Karma Power: 15
Karma: 487
trainerbob is appreciated, and has over 400 Karma points.trainerbob is appreciated, and has over 400 Karma points.trainerbob is appreciated, and has over 400 Karma points.trainerbob is appreciated, and has over 400 Karma points.trainerbob is appreciated, and has over 400 Karma points.
Default Re: What to do with Prior Analysis after instrument was Out Of Calibration (OOC)?

You may never know exactly your intrument went out of calibration. What you must do is move on from this point. I agree with bramiah's procedure. It should take care of your situation. One of the biggest mistakes we know is not letting the customer know what is going on.
Reply With Quote
  #4  
Old 19th May 2009, 09:45 AM
Arena Arena is offline
Involved in Discussions

Registration Date: May 2008
Location: Puerto Rico
 
Posts: 13
Thanks Given to Others: 14
Thanked 3 Times in 3 Posts
Karma Power: 10
Karma: 184
Arena is appreciated, and has over 100 Karma points.Arena is appreciated, and has over 100 Karma points.
Default Re: What to do with Prior Analysis after instrument was Out Of Calibration (OOC)?

Quote:
Originally Posted by sharduli_k View Post

If an instrument is OOC from unknown days before and analyst could not encounter the error as it is much internal and critical to get,what to do with the samples that run on the instrument from some before days?
How one can get the time from when it was out of calibration?

In our organization, we developed a calendar for internal an external calibrations. It lists all measurement equipment used by production and its calibration frequency (weekly, monthly, quarterly, etc.) Also, each time an equipment is calibrated, we use a label with the calibration date and due date. It has worked for us so far.

Just make sure the equipment listed on the calibration calendar is up to date and try not to run out of labels!
Reply With Quote
Thanks to Arena for your informative Post and/or Attachment!
  #5  
Old 19th May 2009, 10:17 AM
harry harry is offline
Super Moderator

Registration Date: Sep 2005
Location: Johore/Malaysia
 
Posts: 2,942
Thanks Given to Others: 571
Thanked 1,031 Times in 746 Posts
Blog Entries: 1
Karma Power: 231
Karma: 7783
harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.
harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.harry is appreciated, and has over 1700 Karma points.
Default Re: What to do with Prior Analysis after instrument was Out Of Calibration (OOC)?

Quote:
Originally Posted by sharduli_k View Post

............ How one can get the time from when it was out of calibration?
What are the 'Risk' involved?

If the risk (or consequential loss or damage) is high, you may want to validate your instrument every day or even every shift. If it is low, you may choose to do it once a week/month or so. In this way, you may be able to contain problems - otherwise, you can't tell by your present practice.
Reply With Quote
  #6  
Old 20th May 2009, 04:09 AM
sharduli_k sharduli_k is offline
Getting Involved (6 to 9 Posts)

Registration Date: May 2009
Location: India
 
Posts: 9
Thanks Given to Others: 2
Thanked 0 Times in 0 Posts
Karma Power: 3
Karma: 10
sharduli_k has less than 100 Karma points so far.
Default Re: What to do with Prior Analysis after instrument was Out Of Calibration (OOC)?

Quote:
Originally Posted by brahmaiah View Post

The foolproof method to correct the situation, is :
1)Containment action:
a)Stop production untill you get a correct instrument
b) inspect all the parts available in your custody.
c)Send an express message to your customer to hold all parts available with him
d)Rush to your customer premises and segregate those components.

2)Corrective action:
a)Calibrate the instrument in question and start production.

3)Prevention:
a)Calibrate more frequently
b)Check the instrument before each time it is used as we do in case of hardness testers and micrometers.
c)Introduce a 'Poka Yoke' if possible.

I HOPE THIS IS ENOUGH!.
V.J.Brahmaiah

So after this, what about the reports released before? how to correct them?
Reply With Quote
  #7  
Old 20th May 2009, 04:13 AM
sharduli_k sharduli_k is offline
Getting Involved (6 to 9 Posts)

Registration Date: May 2009
Location: India
 
Posts: 9
Thanks Given to Others: 2
Thanked 0 Times in 0 Posts
Karma Power: 3
Karma: 10
sharduli_k has less than 100 Karma points so far.
Default Re: What to do with Prior Analysis after instrument was Out Of Calibration (OOC)?

Quote:
Originally Posted by trainerbob View Post

You may never know exactly your intrument went out of calibration. What you must do is move on from this point. I agree with bramiah's procedure. It should take care of your situation. One of the biggest mistakes we know is not letting the customer know what is going on.

I agree to you but we have to inform to customers at least when we have failed any batch or we have passed the batch on threshold values
Reply With Quote
  #8  
Old 20th May 2009, 04:25 AM
brahmaiah brahmaiah is offline
Quality veteran

Registration Date: Mar 2009
Location: india
 
Posts: 383
Thanks Given to Others: 10
Thanked 68 Times in 55 Posts
Blog Entries: 3
Karma Power: 18
Karma: 402
brahmaiah is appreciated, and has over 400 Karma points.brahmaiah is appreciated, and has over 400 Karma points.brahmaiah is appreciated, and has over 400 Karma points.brahmaiah is appreciated, and has over 400 Karma points.brahmaiah is appreciated, and has over 400 Karma points.
Default Re: What to do with Prior Analysis after instrument was Out Of Calibration (OOC)?

Quote:
Originally Posted by sharduli_k View Post

So after this, what about the reports released before? how to correct them?
You may release a revision to all related reports stating 'The inspection report No.XXX dt.DD MM YY is invalid due to use of defective measuring instrument.
-VJB
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Maintenance Instrument Calibration Requirements edelsal ISO 17025 - Calibration, Measurement Gages and Test Laboratories 6 11th September 2008 12:49 PM
Calibration of Noise Source Instrument Ruebenn ISO 17025 - Calibration, Measurement Gages and Test Laboratories 4 2nd September 2008 11:17 AM
Calibration of Measuring instrument, MRP gauges in machining harry ISO 17025 - Calibration, Measurement Gages and Test Laboratories 2 14th July 2008 12:20 PM
Calibration Laboratory Environment and Instrument Tolerances Verod ISO 17025 - Calibration, Measurement Gages and Test Laboratories 15 15th January 2008 03:06 PM
Instrument Calibration Requirements - Determining what affects quality matthew evans ISO 17025 - Calibration, Measurement Gages and Test Laboratories 1 5th May 2000 10:04 AM



The time now is 02:40 AM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts