Calibration Frequency - Evidential Breath Analizer
Hi, I have a question as well. I am reviewing past calibration records of a particular instrument.
One of the calibration date shows 03 March, 2006 on the cert. The next cert date shows 19 September, 2006. The recall interval is 6 mos.
My question is that isn't the next date should have shown 03 September, 2006 instead of the 19th of September?
Thank you for your input.
Regards,
Tungsten
|

23rd June 2009, 01:29 PM
|
 |
Courtesy Access
Registration Date: Jan 2005
Location: Southeast Wisconsin
Age: 57
|
|
Posts: 9,216
Thanks Given to Others: 755
Thanked 2,298 Times in 1,550 Posts
Karma Power: 612
|
|
Re: Calibration Frequency
Quote:
Originally Posted by Tungsten
Hi, I have a question as well. I am reviewing past calibration records of a particular instrument.
One of the calibration date shows 03 March, 2006 on the cert. The next cert date shows 19 September, 2006. The recall interval is 6 mos.
My question is that isn't the next date should have shown 03 September, 2006 instead of the 19th of September?
Thank you for your input.
Regards,
Tungsten
|
What are the requirements? If the documentation says that recall has to happen in the month that the instrument comes due, there's no problem. If the documentation says that the thing must be recalled on or before the end date, then there might be a minor issue. What's the statute of limitations?  Why worry about something that happened nearly three years ago?
__________________
Some men are born mediocre, some men achieve mediocrity, and some men have mediocrity thrust upon them.-- Joseph Heller
|
|
Thanks to Jim Wynne for your informative Post and/or Attachment!
|
|

23rd June 2009, 02:14 PM
|
|
Involved in Discussions
Registration Date: Jun 2009
Location: Ireland
|
|
Posts: 13
Thanks Given to Others: 12
Thanked 0 Times in 0 Posts
Karma Power: 2 Karma: 10 
|
|
Re: Audit NC (nonconformance) for not following guidelines on calibration!
Hi, Thanks for you all for the prompt reply. I have no cal procedure available. This particular equipment is a LION Intoxilizer 6000 IRL EBA (Evidential Breath Analizer) currently used by the Limerick (Ireland) Garda (Police) for evidential breath tests to determine the amount of alcohol in the breath (micro grammes in 100 ml of breath).
My job is to dig up the "dirt", and prove it to the court that there are (several) holes in the calibration intervals and the subject equipment was used from time to time in "out of calibration", otherwords it was ovedrue.
Regards and many thanks.
Tungsten
|

23rd June 2009, 02:34 PM
|
|
Involved in Discussions
Registration Date: Jun 2009
Location: Ireland
|
|
Posts: 13
Thanks Given to Others: 12
Thanked 0 Times in 0 Posts
Karma Power: 2 Karma: 10 
|
|
Re: Calibration Frequency
Quote:
Originally Posted by Jim Wynne
What are the requirements? If the documentation says that recall has to happen in the month that the instrument comes due, there's no problem. If the documentation says that the thing must be recalled on or before the end date, then there might be a minor issue. What's the statute of limitations?  Why worry about something that happened nearly three years ago?
|
Thanks Jim for your input. Yeah, why should we worry...about something that happened almost 3 years ago. I wish it was that easy. I am involved in a court case where my job is to find holes whether the instrument (a breath analyzer) used for breath tests used during out of calibration condition.
Regards,
Tungsten
|

23rd June 2009, 02:38 PM
|
 |
Stop X-bar/R Madness
Registration Date: May 2002
Location: LaGrange, OH
|
|
Posts: 2,091
Thanks Given to Others: 680
Thanked 563 Times in 469 Posts
Karma Power: 150
|
|
Re: Audit NC (nonconformance) for not following guidelines on calibration!
Quote:
Originally Posted by Tungsten
My job is to dig up the "dirt", and prove it to the court that there are (several) holes in the calibration intervals and the subject equipment was used from time to time in "out of calibration", otherwords it was overdue.
|
A weak defense might come from if that device - wherever it was used - ever failed a calibration. If so, how long was the calibration period? Because that would provide evidence that the period was potentially too short.
However, that can be easily countered if they can show that the device passed the previous calibration prior to the event in question, and a calibration after the event. That infers that it was in calibration throughout the period, whether it was late or not, leaving a need for another defense - such as operator training, documentation, etc.
Late calibration is only a risk that later calibration may find the device out of calibration, causing all data back to the previous calibration to be suspect. It is not evidence of out of calibration. A jury might not be able to follow that logic, though.
Of course, this is not legal advice - but it is a good place to start...
__________________
If we didn't care, we wouldn't share. That's why we're there.
|
|
Thanks to bobdoering for your informative Post and/or Attachment!
|
|

23rd June 2009, 02:58 PM
|
|
Involved in Discussions
Registration Date: Jun 2009
Location: Ireland
|
|
Posts: 13
Thanks Given to Others: 12
Thanked 0 Times in 0 Posts
Karma Power: 2 Karma: 10 
|
|
Re: Calibration Frequency
Hi, and thanks very much for your reply. So, otherwords if I stopped by the cops on 19th and they took a breath test with a device which was due for calibration on the 12th of the same month, there is no, or little argument from the point of legality?
|

23rd June 2009, 03:13 PM
|
 |
Stop X-bar/R Madness
Registration Date: May 2002
Location: LaGrange, OH
|
|
Posts: 2,091
Thanks Given to Others: 680
Thanked 563 Times in 469 Posts
Karma Power: 150
|
|
Re: Calibration Frequency
Quote:
Originally Posted by Tungsten
Hi, and thanks very much for your reply. So, otherwords if I stopped by the cops on 19th and they took a breath test with a device which was due for calibration on the 12th of the same month, there is no, or little argument from the point of legality?
|
IF the calibration on the 25th showed the device in calibration, no - there is little to stand on. (Again - not legal advice - but good gage advice)
__________________
If we didn't care, we wouldn't share. That's why we're there.
|
|
Thanks to bobdoering for your informative Post and/or Attachment!
|
|

23rd June 2009, 03:47 PM
|
|
Involved in Discussions
Registration Date: Jun 2009
Location: Ireland
|
|
Posts: 13
Thanks Given to Others: 12
Thanked 0 Times in 0 Posts
Karma Power: 2 Karma: 10 
|
|
Re: Calibration Frequency
This is getting interesting...  . Actually, the next calibration did show the equipment in calibration with a note on the service record that the No. 1 Ethanol cylinder was replaced (the record did not say whether it was faulty, empty or there was a leak). The purpose of the ethanol cylinder (there are two altogether) in the unit is to self-calibrate the device prior to the breath test, and post self-calibrate again at the end of the test. It called the "blank" and it should show zero value.
The other thing is what looks interesting that the test was done at 14.5C, .5C BELOW the OIML International Atmospheric Recommendation on Evidential Breath Analyzers. The guide clearly states that for non-portable EBA's the rated operating conditions must be between 15 deg.C and 35 deg.C.
0.5C does not make much of the difference on the value of the outcome, but in the point of legality (I think...) could be challenged.
Back to the cal intervals, I am wondering why there is calibration intervals at all? Few days here and there out of calibration, seems like does not matter.
Any thoughts would be appreciated. - Thank you!
|
Lower Navigation Bar
|
|
|
|
Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
|
|
|
| Thread Tools |
Search this Thread |
|
|
|
| Display Modes |
Rate Thread Content |
Linear Mode
|
|
Posting Settings
|
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
HTML code is Off
|
|
|
|
|