15th September 2009, 01:20 PM
|
|
Inactive Registered Visitor
Registration Date: Sep 2009
|
|
Posts: 3
Thanks Given to Others: 0
Thanked 2 Times in 1 Post
Karma Power: 16 Karma: 20 
|
|
IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device
We are in the process of doing a gap analysis for getting a CE Mark for our IVD medical device which is currently cleared in the US. One question that came up is which electrical safety standard we need to be following. For the US, we comply with 60601. For Europe it seems like there is a separate standard for IVDs - 61010. Can anyone confirm this?
Thanks.
|

15th September 2009, 03:20 PM
|
 |
Metrologist-Auditor
Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 57
|
|
Posts: 2,100
Thanks Given to Others: 1
Thanked 625 Times in 444 Posts
Karma Power: 265
|
|
|
Re: IEC 60601 vs IEC 61010
It appears that ISO/IEC 60101 may be more for medical devices, while ISO/IEC/UL/CSA 61010 is for laboratory equipment.
Best bet is to find a CB Scheme recognized lab and talk to them. Examples in North America are SGS, UL, CSA, ITS, and I suspect others as well.
Hope this helps.
|
|
Thanks to Hershal for your informative Post and/or Attachment!
|
|

22nd October 2009, 10:35 PM
|
|
Appreciated Information Resource
Registration Date: Oct 2009
Location: Japan / Ise
|
|
Posts: 629
Thanks Given to Others: 35
Thanked 474 Times in 324 Posts
Karma Power: 86
|
|
|
Re: IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device
According to the scope, IEC 60601-1 is only for medical devices with patient contact, this may include some IVD devices where the sample is taken using the same device (such as home use blood sugar level testers).
However, for the vast majority of IVDs the patient is well removed, and hence IEC 60601-1 is not applicable. The correct standard is IEC 61010-1, and any particulars that apply to your device.
There is persistent and incorrect use of IEC 60601-1 for anything that is electrical and medical because of the title of the standard. The scope itself which refers to patient contact are buried inside the standard.
The FDA complicates this, because IEC 60601-1 is listed as a recognized consensus standard, while IEC 61010-1 is not. Thus, many feel more comfortable using IEC 60601-1 in the US even though it is not the correct standard.
In Europe, you will definately have to use the correct standard: IEC 61010-1 and any particulars that are applicable.
|
|
Thanks to Peter Selvey for your informative Post and/or Attachment!
|
|

22nd October 2009, 11:20 PM
|
 |
Addicted to standards
Registration Date: Mar 2006
Location: São Paulo, SP, Brazil
|
|
Posts: 1,985
Thanks Given to Others: 286
Thanked 1,482 Times in 800 Posts
Karma Power: 249
|
|
|
Re: IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device
Quote:
|
According to the scope, IEC 60601-1 is only for medical devices with patient contact
|
Just to correct the information a little, IEC 60601-1 is for electrical medical devices, let´s say it this way, which have patient contact (applied parts by the definition of the standard), or transfer energy to or from the patient or detect this transfer (which means the equipment really does not have to contact the patient).
Still, IVDs devices, as you say, generally do not fit this scope.
__________________
I'm a moderator on the medical device forums, so if you need help with something, feel free to ask!
|
|
Thanks to Marcelo Antunes for your informative Post and/or Attachment!
|
|

18th February 2010, 01:06 AM
|
|
Involved in Discussions
Registration Date: Apr 2008
Location: Singapore
|
|
Posts: 10
Thanks Given to Others: 1
Thanked 2 Times in 1 Post
Karma Power: 22 Karma: 20 
|
|
|
Product Safety in Medical Device
Hi All
I am trying to compile the Product Safety standard and testing requirements for Medical Devices. Would appreciate knowing if any of you have any presentation or reference materials on the above.
Thanks
|

27th September 2012, 03:41 AM
|
|
Involved in Discussions
Registration Date: Mar 2011
Location: Germany
|
|
Posts: 15
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 10 Karma: 10 
|
|
|
Re: IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device
Hello All,
I am working on a IVD medical device. The device is an implantable pressure measurement. The implant communicates with the reader device via RFID means. I would like to know what standards will apply to the reader device antenna. The antenna is in contact with skin. Is IEC 60601 an appropriate one or should I also explore IEC 61010.
Thank you,
Ram
|
Lower Navigation Bar
|
|
|
Do you find this discussion thread helpful and informational?
|
Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
|
|
|
| Thread Tools |
Search this Thread |
|
|
|
| Display Modes |
Rate Thread Content |
Linear Mode
|
|
Forum Posting Settings
|
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts
HTML code is Off
|
|
|
|
|