The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


Related Topic Tags
iec 60601 - medical electrical equipment, testing (general testing topics)
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 18th January 2010, 05:39 AM
nikolaos nikolaos is offline
Involved in Discussions

 
Registration Date: Dec 2008
Location: Italy
 
Posts: 64
Thanks Given to Others: 5
Thanked 6 Times in 6 Posts
Karma Power: 25
Karma: 40
nikolaos has less than 100 Karma points so far.
Please Help! Insulation Barrier Testing - According IEC 60601-1

Hello All,
According IEC 60601-1, testing an insulation barrier, both dielectric and spacing requirements are necessary or it is enough only the dielectric high voltage test? Is there any current limit during the dielectric high voltage test?
Thank you

Sponsored Links
  #2  
Old 18th January 2010, 04:07 PM
Jennifer Kirley's Avatar
Jennifer Kirley Jennifer Kirley is offline
Forum Moderator

 
Registration Date: Jan 2004
Location: Maine, USA
 
Posts: 5,207
Thanks Given to Others: 2,983
Thanked 2,836 Times in 1,627 Posts
Karma Power: 612
Karma: 27178
Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.
Jennifer Kirley is appreciated, and has over 1700 Karma points.Jennifer Kirley is appreciated, and has over 1700 Karma points.
Re: Insulation Barrier Testing - According IEC 60601-1

Does anyone have advice for nikolaos?
__________________
"If you only have a hammer, you tend to see every problem as a nail." Abraham Maslow
Thanks to Jennifer Kirley for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 19th January 2010, 08:31 AM
Marcelo Antunes's Avatar
Marcelo Antunes Marcelo Antunes is offline
Addicted to standards

 
Registration Date: Mar 2006
Location: São Paulo, SP, Brazil
 
Posts: 1,985
Thanks Given to Others: 286
Thanked 1,482 Times in 800 Posts
Karma Power: 249
Karma: 10982
Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.
Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.
Send a message via ICQ to Marcelo Antunes Send a message via MSN to Marcelo Antunes Send a message via Skype™ to Marcelo Antunes
Re: Insulation Barrier Testing - According IEC 60601-1

Hi, just a quick answer cause i´m offline for some days: you need both spacing and dielectric strenght.

And there´s no current limit for the dielectric test in IEC 60601.
__________________
I'm a moderator on the medical device forums, so if you need help with something, feel free to ask!
Thanks to Marcelo Antunes for your informative Post and/or Attachment!
  #4  
Old 19th January 2010, 07:11 PM
Peter Selvey Peter Selvey is offline
Appreciated Information Resource

 
Registration Date: Oct 2009
Location: Japan / Ise
 
Posts: 629
Thanks Given to Others: 35
Thanked 474 Times in 324 Posts
Karma Power: 86
Karma: 4719
Peter Selvey is appreciated, and has over 1700 Karma points.
Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.
Re: Insulation Barrier Testing - According IEC 60601-1

Dielectric strength and spacings are different requirements.

Dielectric strength is an ageing test for solid insulation, such as on wiring, in transformers and optocouplers.

Spacings are for two other reasons: one (clearance) for short term spikes in the order of 1000-3000V that occasionally appear on the mains, the other (creepage) is to have enough space to protect against long term dust/pollution accumulation.

So you really need to assess all three separately.

Current limit for dielectric strength: IEC 60601-1 does not specify a current limit for the HV tester, however other standards specify 100mA. This high current is theoretically important due to the effect of partial breakdown, but in practice modern insulating materials have dielectric strength well above the needs so it rarely has an impact.

Otherwise, because of EMC capacitors, you may need a HV current capacity of around 5mA.

The current pulled during a HV test is about 6 times the "neutral open" earth leakage current, e.g. if your equipment has 300uA @ 264V earth leakage in SFC (neutral open), you will see ~1.8mA leakage in the basic insulation test (@ 1,500V). It's a nice confirmation of your test set up to confirm this value. Double insulation test current is usually much lower (if not, something is wrong with your set up).

Yoroshiku,
(= regards in Japanese)
Peter
Thanks to Peter Selvey for your informative Post and/or Attachment!
  #5  
Old 20th January 2010, 02:39 AM
nikolaos nikolaos is offline
Involved in Discussions

 
Registration Date: Dec 2008
Location: Italy
 
Posts: 64
Thanks Given to Others: 5
Thanked 6 Times in 6 Posts
Karma Power: 25
Karma: 40
nikolaos has less than 100 Karma points so far.
Re: Insulation Barrier Testing - According IEC 60601-1

Thank you for all these information, indeed very clear and precise. Actually I have a discussion with an Italian Notify Body about Insulation Barrier Testing. I agree that both electrical (dielectric HV test) and mechanical (C&C) are both necessary to check the insulation, but the NB insist telling me that only dielectric test with 8mA of leakage current is enough. According them C&C check is not necessary. I work as a project engineer for a Medical manufacturing company and it is very useful for me to know what you think about this item, so any other opinions are very valuable.
Regards
Nikolaos
  #6  
Old 20th January 2010, 04:13 AM
Marcelo Antunes's Avatar
Marcelo Antunes Marcelo Antunes is offline
Addicted to standards

 
Registration Date: Mar 2006
Location: São Paulo, SP, Brazil
 
Posts: 1,985
Thanks Given to Others: 286
Thanked 1,482 Times in 800 Posts
Karma Power: 249
Karma: 10982
Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.
Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.Marcelo Antunes is appreciated, and has over 1700 Karma points.
Send a message via ICQ to Marcelo Antunes Send a message via MSN to Marcelo Antunes Send a message via Skype™ to Marcelo Antunes
Re: Insulation Barrier Testing - According IEC 60601-1

It´s weird that your NB aasks only for dielectric strenght, and weirder that they put a limit on the current for the dielectric strenght as the standards does not have that (it does not have a current limit but it defines how a test can fail by defining what is insulation breakdown).

Anyway, there´s a difference between what a standard requires and what a regulation requires. You are not required to comply with IEC 60601, although it´s desirable, and maybe that is the position of your NB. But keep in mind that, the way they´re are asking you, you are NOT in compliance with the standard, and may be at risk of having ptoblems in other places which have a more strict view on standards.
__________________
I'm a moderator on the medical device forums, so if you need help with something, feel free to ask!
  #7  
Old 20th January 2010, 05:11 AM
nikolaos nikolaos is offline
Involved in Discussions

 
Registration Date: Dec 2008
Location: Italy
 
Posts: 64
Thanks Given to Others: 5
Thanked 6 Times in 6 Posts
Karma Power: 25
Karma: 40
nikolaos has less than 100 Karma points so far.
Re: Insulation Barrier Testing - According IEC 60601-1

OK, but I asked to test according 60601-1. In the report is clearly written that they followed the testing procedure according 60601.1(1998). More, I found no isolation diagram in the report.
  #8  
Old 20th January 2010, 05:13 AM
Martin IT Martin IT is offline
Involved in Discussions

 
Registration Date: Oct 2009
Location: Italy
Age: 35
 
Posts: 104
Thanks Given to Others: 32
Thanked 64 Times in 36 Posts
Karma Power: 25
Karma: 384
Martin IT is appreciated, and has over 300 Karma points.Martin IT is appreciated, and has over 300 Karma points.Martin IT is appreciated, and has over 300 Karma points.Martin IT is appreciated, and has over 300 Karma points.
Re: Insulation Barrier Testing - According IEC 60601-1

The compliance with the IEC 60601-1 is not mandatory but highly recommended. In my opinion the suggestion of the NB should be taken into serious consideration.

I've got the same problem ( I'ma a project engineer and I work with an Italian Notify Body ), and at the moment we perform only dielectric test with 8mA of leakage current.

Let me know the way you will forward!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
EN (IEC) 60601-1 – Is in-house testing definitely OK? SteveK IEC 60601 - Medical Electrical Equipment Safety Standards Series 17 11th May 2012 07:48 AM
Do HF accessories require EMC testing according to IEC 60601-1-2? ascherp IEC 60601 - Medical Electrical Equipment Safety Standards Series 9 17th February 2012 07:57 AM
Does a 12V DC device need EMC (IEC 60601-1-2) Testing? SteveK IEC 60601 - Medical Electrical Equipment Safety Standards Series 4 26th October 2010 08:03 AM
IEC/EN 60601 Testing and CE Marking for a Non Medical Device Eamon CE Marking (Conformité Européene) / CB Scheme 9 17th September 2010 01:34 AM
IEC 60601-2-37 - No Dielectric testing specifications? DuttonT IEC 60601 - Medical Electrical Equipment Safety Standards Series 4 22nd May 2007 05:08 PM



The time now is 01:34 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


   


Marc Timothy Smith - Elsmar.com
8466 LeSourdsville-West Chester Road, Olde West Chester, Ohio 45069-1929
513 341-6272