The Elsmar Cove Forum and Site Map The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > Medical Device Regulations and Standards > Medical Device Related Standards > ISO 13485 and ISO 14969 - Medical Devices - Quality Management Systems


The Elsmar Cove Forum SideBar!
Thread Tags
Tag Cloud
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link











$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World

Tags
calibration, medical devices, calibration frequency or interval
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 8th February 2010, 09:48 AM
Bob Sutton Bob Sutton is offline
Shy Poster (1 to 5 Posts)

Registration Date: Feb 2010
Location: United States, Ohio
 
Posts: 3
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 2
Karma: 10
Bob Sutton has less than 100 Karma points so far.
Please Help! Can I Claim That Periodic Calibration is Not Required?

I've spent the last hour reviewing multiple threads related to calibration and am very impressed with the knowledge that is out there. This is my first posting. I have recently taken on the managing responsibilities for our company’s calibration program. We are a medical company. My question relates to the designation of "Periodic Calibration Not Required". If the device in question cannot be calibrated, would you designate it as PCNR or would it not be in the calibration system at all. I realize that basic tools don't go into the calibration system however, we make fixtures that are used during the process yet no aspect of the fixture can be calibrated. The previous mind set was that all fixtures should either be calibrated (if possible) or have a PCNR label. The justification is written up as: "No aspect of fixture can be calibrated".
In my opinion the first decision to be made is to determine if the device is truly Inspection, monitoring & measuring equipment. If answer is no, then device is not part of calibration program. Your thoughts on this would be greatly appreciated.
Reply With Quote

Sponsored Links
  #2  
Old 8th February 2010, 11:42 AM
Jim Wynne's Avatar
Jim Wynne Jim Wynne is offline
Courtesy Access

Registration Date: Jan 2005
Location: Southeast Wisconsin
Age: 57
 
Posts: 10,049
Thanks Given to Others: 937
Thanked 2,909 Times in 1,915 Posts
Blog Entries: 1
Karma Power: 709
Karma: 24346
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Default Re:Can I Claim That Periodic Calibration is Not Required?

I'm having a hard time trying to imagine any sort of measurement device that can't be calibrated, unless you're saying that the things you're referring to can't be adjusted. "Calibration" is comparison of a device to a controlled standard, and may or may not involve adjustment. If you could give more information about the types of devices you're referring to, it would help.
__________________
To be able to ask a question clearly is two-thirds of the way to getting it answered. ~John Ruskin

Last edited by Al Rosen; 8th February 2010 at 04:26 PM.
Reply With Quote
Thanks to Jim Wynne for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 8th February 2010, 11:54 AM
Bob Sutton Bob Sutton is offline
Shy Poster (1 to 5 Posts)

Registration Date: Feb 2010
Location: United States, Ohio
 
Posts: 3
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 2
Karma: 10
Bob Sutton has less than 100 Karma points so far.
Default Re: Can I Claim That Periodic Calibration is Not Required?

Here is what was given to me.
Equip Description is:
"Delphin display Test Fixture for a Quantum Test Stand."

Justification for designation of Calibration Not Required is:
"The purpose of the Delphin Display test Ficture is to exercise a device under test against a variety of functional requirements. The active element of the fixture (power supply) is routinely cross-checked as part of the functional test process, but it cannot be individually calibrated. A failure of the test fixture would result in a functional test during routing production testing, which would identify the fixture as in need of repair".

Hope this helps. If there is something else you need to know or need clarified just let me know.

Last edited by Al Rosen; 8th February 2010 at 04:26 PM.
Reply With Quote
  #4  
Old 8th February 2010, 12:07 PM
Jim Wynne's Avatar
Jim Wynne Jim Wynne is offline
Courtesy Access

Registration Date: Jan 2005
Location: Southeast Wisconsin
Age: 57
 
Posts: 10,049
Thanks Given to Others: 937
Thanked 2,909 Times in 1,915 Posts
Blog Entries: 1
Karma Power: 709
Karma: 24346
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.Jim Wynne is appreciated, and has over 1700 Karma points.
Default Re: Can I Claim That Periodic Calibration is Not Required?

Quote:
Originally Posted by Bob Sutton View Post

Here is what was given to me.
Equip Description is:
"Delphin display Test Fixture for a Quantum Test Stand."

Justification for designation of Calibration Not Required is:
"The purpose of the Delphin Display test Ficture is to exercise a device under test against a variety of functional requirements. The active element of the fixture (power supply) is routinely cross-checked as part of the functional test process, but it cannot be individually calibrated. A failure of the test fixture would result in a functional test during routing production testing, which would identify the fixture as in need of repair".

Hope this helps. If there is something else you need to know or need clarified just let me know.
I don't know what the "Quantum Test Stand" is measuring or checking, other than "a variety of functional requirements." Perhaps someone else here might have some experience with this sort of thing that I lack, but given the information available I can't say whether there's anything that should be calibrated or not.
__________________
To be able to ask a question clearly is two-thirds of the way to getting it answered. ~John Ruskin

Last edited by Al Rosen; 8th February 2010 at 04:27 PM.
Reply With Quote
  #5  
Old 8th February 2010, 01:38 PM
Bob Sutton Bob Sutton is offline
Shy Poster (1 to 5 Posts)

Registration Date: Feb 2010
Location: United States, Ohio
 
Posts: 3
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 2
Karma: 10
Bob Sutton has less than 100 Karma points so far.
Default Re: Can I Claim That Periodic Calibration is Not Required?

Jim,
I went to our manf floor and watched the fixture in use. First let me explain that a Quantum Test Stand is a functional test stand that we designed and built which is loaded with validated software. This is really a functional test stand used to test different components of the heart lung machine we build.
The Delphin display fixture is a fixture that was designed to hold a PC Band and digital display (PC board and digital display go together as one unti. The fixture is connected to the Quantum (functional) test stand and different test are run to ensure the display functions properly. The fixture itself consists of some electrical connectors, wires and an internal PC board.

Last edited by Al Rosen; 8th February 2010 at 04:27 PM.
Reply With Quote
  #6  
Old 8th February 2010, 03:26 PM
bobdoering's Avatar
bobdoering bobdoering is offline
Stop X-bar/R Madness

Registration Date: May 2002
Location: LaGrange, OH
 
Posts: 2,710
Thanks Given to Others: 830
Thanked 823 Times in 670 Posts
Blog Entries: 16
Karma Power: 203
Karma: 5768
bobdoering is appreciated, and has over 1700 Karma points.
bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.bobdoering is appreciated, and has over 1700 Karma points.
Send a message via Yahoo to bobdoering
Default Re: Can I Claim That Periodic Calibration is Not Required?

Simple question: How do you know that the test stand is accurately passing good parts and failing bad parts?

Is it attribute pass/fail? Can you verify it with certified boundary samples?

Most important: is there any adjustment whatsoever on the device? If so, calibration (assuring the adjustments have not altered the ability of the device to correctly pass or fail product) would be necessary.
__________________
If we didn't care, we wouldn't share. That's why we're there.

Last edited by Al Rosen; 8th February 2010 at 04:28 PM.
Reply With Quote
  #7  
Old 8th February 2010, 04:04 PM
Hershal's Avatar
Hershal Hershal is offline
Metrologist-Auditor

Registration Date: Mar 2004
Location: So Cal, California, U.S.
Age: 54
 
Posts: 1,745
Thanks Given to Others: 1
Thanked 396 Times in 286 Posts
Karma Power: 199
Karma: 6978
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.Hershal is appreciated, and has over 1700 Karma points.
Default Re: Can I Claim That Periodic Calibration is Not Required?

Quote:
Originally Posted by azaplas View Post

Hi everyone,

I have a quick question about calibration labels.

In our surveillance audit, we received a "Sensitive Point" indicating:
"Some test equipments are not calibrated and not identified as 'calibration not required'".

Are there any standard/official formats for the "Calibration Not Required" labels? Must I purchase these labels from a company who makes them or may I use say a standard white mailing label and print out my own "Calibration Not Required" label?

Thanks!
You can make them yourself. Of greater significance is that you have a good reason to give Management and your auditor(s) regarding why it is CNR.

Last edited by Al Rosen; 8th February 2010 at 04:28 PM.
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > Medical Device Regulations and Standards > Medical Device Related Standards > ISO 13485 and ISO 14969 - Medical Devices - Quality Management Systems

Bookmarks

Tags
calibration, medical devices, calibration frequency or interval


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Thermohygrometer Calibration - Number of Required Calibration Points corollax General ISO 17025 Topics - Calibration, Measurement Gages and Test Laboratories 11 21st June 2010 03:51 PM
Claim caused by Gauge Out of Calibration sokolasty General ISO 17025 Topics - Calibration, Measurement Gages and Test Laboratories 2 31st October 2009 02:53 AM
Gage pin calibration - Actual calibration readings required on the calibration cert mfr11 General ISO 17025 Topics - Calibration, Measurement Gages and Test Laboratories 3 9th April 2009 12:58 PM
Design: To claim exemption, or not to claim exemption (exclusion) normzone Design and Development - Process and Product 23 17th September 2008 10:20 AM
Calibration required? We buy part of our machined parts - Supplier gage calibration? svance General ISO 17025 Topics - Calibration, Measurement Gages and Test Laboratories 6 21st February 2007 01:18 PM



The time now is 05:35 AM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.