Sponsor Links
Courtesy Quick Links
Links that Cove visitors will find useful in your quest for knowledge:
International Standards Organization - ISO Standards and Information
Atul's
Quality Forum Online
Howard's
Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum
Bob Doering
Bob Doering's Blogs and,
Quality Digest
Related LinkedIn Groups
ISO 9001 for Small Businesses
ISO 9001:2015 Revision Discussions
Information Security Community
Medical Devices Group
Quality and Regulatory Network
FDA (Food and Drugs)
AS91XX Series - Tips and Advice
Related Topic Tags
calibration environmental conditions, measurement and monitoring and test devices, calibration (general topics)
Reply
Thread Tools Search this Thread Rating: Thread Rating: 4 votes, 4.75 average. Display Modes
  Post Number #1
Old 9th April 2002, 03:29 PM
paulmorrow
Total Posts: n/a
Question Environmental Conditions for Calibration - Temperature and Humidity
Are there a standard set of environmental conditions, e.g., temperature and humidity, that must be met during the calibration of standard measuring instruments [calipers, micrometers, etc.? A prospective supplier of calibration services is offering to perform such calibrations on site [to save transportation time and $] but I am concerned that the proposed location does not meet the required [?]controls.
Sponsored Links
  Post Number #2
Old 9th April 2002, 04:54 PM
Ryan Wilde's Avatar
Ryan Wilde
Standard conditions for almost all dimensional calibrations is 20C (68F). Now is where it gets tricky. To hold the uncertainties that you are looking for, temperature range varies.
In a nutshell, gauge blocks and instruments grow with temperature increases. They grow (assuming steel gauge blocks and calipers/mic spindles) at a rate of 11.5m/m/C (6.4in/in/F) about 10%.
Therefore, with each C change from 20C, we add to uncertainty of 2.3m/m (or 0.06m per 25mm). Uncertainty on a micrometer at 2.5m (0.0001") over a 25mm (1") span would dictate that you would not want to stray from 20C by more than about 5C (68F 9F). If your micrometers are less accurate, then you could stray more. If they are more accurate, then you need a tighter environment still. This is all reliant on the base uncertainty of the company calibrating your tools, because if they start high, it lessens the room for error allowable for temperature.
The biggest worry is usually rate of change, because the mass of a gauge block versus the mass of a set of calipers or micrometer spindle is a big difference. If your environment can't stay within about 2C/hour, then calibrations start really giving false measurements. If the temperature of the instrument and the temperature of the standard vary by much, errors really start adding up quickly (hence the little chunks of plastic on the micrometer are known as 'heat shields', and are meant to slow the transfer of heat from your hand to the micrometer).
Calipers are more forgiving (because of the 10:1 accuracy of a mic vs. a caliper), and you can generally get away with about 20C a ballpark.
I omitted a lot of calculations here, for brevity and sanity that some of the savvy folks reading this should pick up on, such as distributions, etc., but this is a pretty good rule of thumb for a quick environmental check.
Humidity does not effect caliper or micrometer calibration, although as a supplier, I never liked taking my blocks into an environment with >60% humidity because they rust so quickly.
Hope this helps,
Ryan
Thank You to Ryan Wilde for your informative Post and/or Attachment!
Sponsored Links
  Post Number #3
Old 10th April 2002, 02:29 AM
Atul Khandekar's Avatar
Atul Khandekar
Read This! Some pointers
I can't provide as detailed and elaborate response as Ryan has done.
The two websites I can suggest are:
http://www.a2la.org/
http://www.ukas.com/
Both these have got lots of pdf documentation on various aspects of calibration lab requirements. May be you'll find something there also.
-Atul.
  Post Number #4
Old 10th April 2002, 11:38 AM
Ken K's Avatar
Ken K
The lab which does our gage blocks and weights does so in an environment with humidity < 45%. Temperature range is
68.0 +/-1C.
  Post Number #5
Old 10th April 2002, 02:49 PM
Ryan Wilde's Avatar
Ryan Wilde
68.0 +/-1C.
That is actually a fairly broad temp range for a lab doing gage blocks. Even at 20C 0.25C, temperature accounted for over 50% of our uncertainty on gage blocks, even with thermal compensation. We also had problems with humidity (we spec'd 25-45% RH, and in the winter it always tried to go below 25%, which is not healthy for electronic devices such as gage block comparators).
I now rarely touch dimensional, and I work in a comfy 23 lab. Life is good.
Ryan
  Post Number #6
Old 7th November 2003, 09:37 PM
AlbertPaglinawan
I would like to know what is the ideal environment in calibrating dimensional and electrical equipment. Ours is currently 21C +/- 5C and 50% +/- 10%. Now, I need to justify this figures. Where did we get these figures? What possible justifications can we give?
Thanks in advance!
cheers,
Lord Ituralde
  Post Number #7
Old 10th November 2003, 11:59 AM
Graeme's Avatar
Graeme
In Reply to Parent Post by Lord Ituralde
Here are some figures from a couple of readily available documents --
TEMPERATURE
Dimensional, Optical and Mass
NSCL: 20 C +/- 0.5 C for general calibrations
ISA: 20 C +/- 1 C overall and +/- 0.3 C at the point of measurement
All other disciplines
NCSL: 23 C +/- 2 C for general calibrations
ISA: 23 C +/- 1.5 C
RELATIVE HUMIDITY
Dimensional, Optical, Mass
NCSL: 40% +/- 5% RH at 20 C
ISA: 45% RH maximum at 20 C
NCSL: 40% +/- 5% RH at 23 C
ISA: 20 - 55% RH at 23 C
Each of these recommended practices contain a lot of other information, of course. Both of them are targeted at "standards" laboratories; if you are in a company's production lab, this is the type of lab you would send your transfer standards to. In both cases, the values above are the most relaxed but still might be tighter than what you want to maintain. On the other hand, they also agree with what I remember (10 years ago) of the US Navy's requirements for general calibration labs --
Measurements using gage blocks: 20 C +/- 0.5 C and 10 - 45% RH
Other dimensional/optical/mass: 23 C +/- 3 C and 20 - 60% RH
Electrical/Electronic/RF/Microwave: 23 C +/- 5 C and 20 - 60% RH.
What you have appears to be a compromise. Do you perform all calibrations in the same aera? The temperature range means that you are giving up some accuracy capability in dimensional measurements, and the high humidity is inreasing the risk of corrosion (rust) on your standards. The only way to get better control is to divide the lab into two areas with sepatrate temperature/humidity control systems. (There may be other ways, maybe an air-conditioning engineer would know ...)
You might also want to check the NIST web site. A couple of items that may be useful are the State Weights and Measures Laboratory Handbook (NIST Handbook 143), and the Gage Block Handbook (NIST Monograph 180).
Last edited by Graeme; 10th November 2003 at 12:02 PM. Reason: corrected fumble-fingered spelling
  Post Number #8
Old 10th November 2003, 01:24 PM
howste's Avatar
howste
Reading this thread brought to mind a few questions:
1)If you calibrate all of your gages in a laboratory with a controlled environment, then use the gages in an uncontrolled environment, what kind of results do you get?
2) Does it make sense to let the standards "soak" and calibrate the equipment in the environment it will be used in?
3) Is it better to keep gages in the environment they will be used in, or in a controlled lab?
Reply
Bookmarks
Search this Thread:
Advanced Forum Search
Rate Thread Content:
Forum Posting Settings
BB code is On
Smilies are On
[IMG] code is On
Similar Discussion Threads
Buying a Temperature and Humidity Device, and Calibration DGShane AS9100, Nadcap and related Aerospace Standards and Requirements 8 5th July 2011 08:43 PM
How to use temperature and humidity environmental control data? Dudes Statistical Analysis Tools, Techniques and SPC 18 1st March 2010 05:38 AM
Analytical Lab Environmental Condition (Temperature & Humidity Control Limit) blueicecube Metrology Topics - Measurement Devices, Calibration and Test Laboratories 5 17th February 2010 03:51 PM
Temperature and Humidity Device Calibration - Came back as Out Of Calibration pancho-2007 Metrology Topics - Measurement Devices, Calibration and Test Laboratories 8 4th March 2005 02:45 PM
Laboratory Environmental Controls - Temperature and Humidity Ryan Wilde Metrology Topics - Measurement Devices, Calibration and Test Laboratories 5 13th July 2001 08:58 AM