The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


Related Topic Tags
electrical safety, ultrasonic probes, medical devices (general)
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 15th January 2011, 12:50 AM
sankui sankui is offline
Involved in Discussions

 
Registration Date: Jun 2009
 
Posts: 44
Thanks Given to Others: 8
Thanked 7 Times in 7 Posts
Karma Power: 21
Karma: 45
sankui has less than 100 Karma points so far.
Please Help! High Voltage inside Ultrasonic Probes - Do SIP/SOP ports require isolation?

Dear all,

Regarding the ultrasonic equipment, as you may know that the ultrasonic probe driving control circuit will generate high voltage pulse (about 100V) to excite the piezoelectric crystal inside the probe. Because there is no separation between the excitation pulse voltage (100V) and other SELV circuit (e.g. 24V), 100V voltage may appear on the SIP/SOP ports in some situation (e.g. non-safety relevant components failed).

I'm worried about is whether the above mention allowed? Whether SIP/SOP ports need for isolation?

Sponsored Links
  #2  
Old 15th January 2011, 09:12 AM
Peter Selvey Peter Selvey is offline
Appreciated Information Resource

 
Registration Date: Oct 2009
Location: Japan / Ise
 
Posts: 629
Thanks Given to Others: 35
Thanked 474 Times in 324 Posts
Karma Power: 86
Karma: 4719
Peter Selvey is appreciated, and has over 1700 Karma points.
Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.
Re: High Voltage inside Ultrasonic Probes - Do SIP/SOP ports require isolation?

Hi Sankui,

You have many interesting questions - is your company designing many different types of medical devices?

For ultrasound HV supplies:

Typically there are three aspects which make these safe:

First, in "normal condition" although the peak voltage may be high the rms voltage is usually around 5V, and the frequencies in the MHz which are harmless. You can touch this voltage without any fear.

Second, most ultrasounds have a sensor which detects if the probe is connected. If the probe is not connected, the +HV and -HV supplies (dc/dc converters) are turned off, or disconnected by a relay.

Third, the HV supply should be current limited to the minimum necessary for the function. Because both of the above methods of protection are electronic based, the reliability is limited. In practice it should be OK, but to be on the safe side it makes sense to limit the current as far as possible. Literature indicates for dc currents, you need around 150mA for ~0.001 probability of fibrillation. Most ultrasounds work with around 20mA, at which the worst expected effect is a bit of pain.

Hope this helps.

Additional note:
The above is intended for the connector for the ultrasound probe itself. However, your question may be the SIP/SOPs at the rear of the device. This is almost for sure safe but the reason is complicated. Again it relies on a few factors:

1) The HV supply being physically well removed from any components, cabling that directly connects to a SIP/SOP
2) The HV supply having a current limit that is below the sink current of the lower voltage power supply rails (+5V, +/-12V etc)
3) The HV supply having a common ground (0V) to the SIP/SOP circuits

The theory is that if the HV supply were to short to SIP/SOP circuits, in most cases the current limit would operate and pull the voltage down to less than the lower voltage power supply rails. This can be verified by actual tests shorting the HV to the nearby power supply rails.

There are rare cases where the HV shorts to a nearby point that has high impedance or to a lower power supply rail with insufficient sink capacity, for example:

+100V 20mA supply shorting to a point with 4kohm to 0V (results in 80V)
+100V 200mA supply shorting to a -5V rail with only 50mA sink capacity (-5V regulator blows and >60V appears)

These have cause problems in designs where the HV circuit is situated physically close to the SIP/SOP circuits (e.g. same PCB, adjacent circuits, even same connector). But, in most ultrasound designs the HV parts are well removed (e.g. different PCBs). There is no functional reason to have HV circuits adjacent to SIP/SOP circuits except for the ultrasound probe connector which is covered above.

Last edited by Peter Selvey; 15th January 2011 at 07:33 PM. Reason: Additional information
Thank You to Peter Selvey for your informative Post and/or Attachment!
Sponsored Links

  #3  
Old 15th January 2011, 10:46 PM
sankui sankui is offline
Involved in Discussions

 
Registration Date: Jun 2009
 
Posts: 44
Thanks Given to Others: 8
Thanked 7 Times in 7 Posts
Karma Power: 21
Karma: 45
sankui has less than 100 Karma points so far.
Re: High Voltage inside Ultrasonic Probes - Do SIP/SOP ports require isolation?

Hi peter,

Yes, my company manufactures several types of medical equipment, including ventilator, humidifier, patient monitor, ultrasonic equipment (developing).

Your input is very detailed, you must be a medical expert. Do you work for test agency and/or IEC working group?

Regarding the sink current, as far as I know, many low voltage regulators have small sink capability (normally less than 20mA). This can also know from the IC datasheet. If so, the situation will become worse.

Last edited by sankui; 15th January 2011 at 11:35 PM.
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Dates That Require to Stated in Standard Operating Procedure (SOP) & Definitions suka_batang Documentation Control Systems, Procedures, Forms and Templates 1 20th September 2011 04:47 AM
Appearance of an External Voltage on SIP/SOP (Medical Devices) Roland chung IEC 60601 - Medical Electrical Equipment Safety Standards Series 15 4th July 2011 10:23 PM
Measurement and Inspection of SAE J-1926 O-Ring Ports nitemare Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5 24th November 2009 10:46 PM
High Level CAPA (Corrective Action Preventive Action) SOP Example Needed Calico Nonconformance and Corrective Action 6 31st October 2009 09:45 PM
Gage R&R on Oscilloscopes, Fluke voltmeters, Voltage probes, etc. & the like bstone Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0 14th July 2005 07:30 PM



The time now is 10:24 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


   


Marc Timothy Smith - Elsmar.com
8466 LeSourdsville-West Chester Road, West Chester, Ohio 45069-1929
513 341-6272