The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Regulations > US Medical Device Regulations > US Medical Devices (21 CFR part 820)
Forum Username


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


Related Topic Tags
21 cfr part 820 - med. device mfger. qsr requirements, clinical laboratories, medical device software, software (general topics), software validation, process verification and validation
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 18th January 2011, 02:27 PM
singber singber is offline
Involved in Discussions

 
Registration Date: Sep 2010
 
Posts: 23
Thanks Given to Others: 6
Thanked 3 Times in 3 Posts
Karma Power: 14
Karma: 25
singber has less than 100 Karma points so far.
Please Help! 21 CFR 820.75 Process Validation for Medical Device Software

Hello,
Can anyone provide examples of how 21 CFR 820.75 Process Validation would apply to a medical device that is strictly software (clinical laboratory information system.) I understand the concept of process validation as it applies to a physical device (i.e. sterility), but cannot determine a scenario where process validation (where the results of a process cannot be fully verified by subsequent inspection and test) would apply to the manufacturing of software.

Any input would be appreciated.

Sponsored Links
  #2  
Old 18th January 2011, 06:44 PM
gholland gholland is offline
Involved in Discussions

 
Registration Date: May 2008
Location: Denver, CO
 
Posts: 95
Thanks Given to Others: 15
Thanked 53 Times in 40 Posts
Karma Power: 31
Karma: 614
gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.gholland is appreciated, and has over 500 Karma points.
Re: 21 CFR 820.75 Process Validation for Medical Device Software

My assumption is that you're asking about validating the installation of software onto a recording device. If you're asking about actually validating the writing of software code... skip this message

One way I've seen it done is that the tool used to burn the CD (assuming this is the process) that you use to distribute the software is Validated. From there the size of the program is compared to the original and assuming you have a successful copy your sizes should be comparable. From there you would demonstrate (Sample size 3 or so) that the software you call 'good' will be what you want it to do.

I'm not a software expert but in my dealings in software land that's what I've seen on the manufacturing floor.


Sponsored Links

  #3  
Old 19th January 2011, 08:03 AM
singber singber is offline
Involved in Discussions

 
Registration Date: Sep 2010
 
Posts: 23
Thanks Given to Others: 6
Thanked 3 Times in 3 Posts
Karma Power: 14
Karma: 25
singber has less than 100 Karma points so far.
Re: 21 CFR 820.75 Process Validation for Medical Device Software

Thank you for your input...here's my problem.

The actual text of the regulation is "Where the results of a process cannot be fully verified by subsequent inspection and test, the process shall be validated with a high degree of assurance and approved according to established procedures."
The process that you described seems like it's one that could be verified by subsequent inspection and test, since as you stated you could verify files sizes and also test out the CD.

I'm interested in any part of the software development life cycle that would apply to this reg, both coding, and transfer to CD or via other means.

Thanks again for your post.
  #4  
Old 19th January 2011, 11:05 AM
yodon yodon is offline
Appreciated Information Resource

 
Registration Date: Jun 2005
Location: Plano, TX
Age: 53
 
Posts: 571
Thanks Given to Others: 374
Thanked 341 Times in 244 Posts
Karma Power: 98
Karma: 4127
yodon is appreciated, and has over 1700 Karma points.
yodon is appreciated, and has over 1700 Karma points.yodon is appreciated, and has over 1700 Karma points.yodon is appreciated, and has over 1700 Karma points.yodon is appreciated, and has over 1700 Karma points.yodon is appreciated, and has over 1700 Karma points.yodon is appreciated, and has over 1700 Karma points.yodon is appreciated, and has over 1700 Karma points.yodon is appreciated, and has over 1700 Karma points.yodon is appreciated, and has over 1700 Karma points.
Re: 21 CFR 820.75 Process Validation for Medical Device Software

gholland is certainly on the right track. You note that the CD replication proces could be verified, but you don't say that you do perform 100% verification of the process output.

Practically speaking,though,there is little chance that the software would install if the CD was corrupted. So you could argue that the output is verified bythe fact that the software installs. You could take that a step further and putting in a checksum test as part of the install process.

Generally speaking, though, there's just not much in the software 'production' that will require such process validation. We typically do the checksum thing and that has been well received. We (or our clients) have not been called to the carpet on this.
Thanks to yodon for your informative Post and/or Attachment!
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Regulations > US Medical Device Regulations > US Medical Devices (21 CFR part 820)

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
21 CFR Part 820 in English - We are an `In vitro´ Medical Device European company invitro_spain US Medical Devices (21 CFR part 820) 6 12th December 2011 03:39 PM
21 CFR 820 - Stability and Validation Requirements Jim-S US Medical Devices (21 CFR part 820) 14 21st May 2010 02:12 PM
Equipment, Software and Processes Validation for Medical Devices - 21CFR 820.75 (a) chasf ISO 13485 and ISO 14969 - Medical Devices - Quality Management Systems 35 22nd December 2008 04:25 PM
Medical Device QMS manual - MDD, AIMD, 21 CFR part 820, CMDCS, and the new ISO 13485 compliance - 2006 ISO 10013 - Quality Management System (QMS) Manuals 3 19th October 2004 12:27 PM
Finished Medical Device - Definition of - FDA 21 CFR Part 820.1 Clarence.L Definitions, Acronyms, Abbreviations and Interpretations 10 20th April 2004 09:45 PM



The time now is 08:52 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


   


Marc Timothy Smith - Elsmar.com
8466 LeSourdsville-West Chester Road, Olde West Chester, Ohio 45069-1929
513 341-6272