The Elsmar Cove Wiki More Free Files The Elsmar Cove Forums Discussion Thread Index Post Attachments Listing Failure Modes Services and Solutions to Problems Elsmar cove Forums Main Page Elsmar Cove Home Page

Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series
Forum Username

Elsmar Cove Forum Visitor Notice(s)


Elsmar Cove Forum Sidebar
Custom Search
Monitor the Elsmar Forum
Monitor New Forum Posts
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Sponsor Links







Donate and $ Contributor Forum Access
Sponsored Links
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's
International Quality Services

Atul's
Symphony Technologies

Marcelo Antunes'
SQR Consulting

Bob Doering's
Correct SPC - Precision Machining


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


Related Topic Tags
iec 60601 - medical electrical equipment, test methods and procedures and plans
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 19th May 2011, 04:09 AM
raymond02302 raymond02302 is offline
Involved in Discussions

 
Registration Date: Mar 2011
Location: China
 
Posts: 56
Thanks Given to Others: 15
Thanked 21 Times in 17 Posts
Karma Power: 14
Karma: 115
raymond02302 is appreciated, and has over 100 Karma points.raymond02302 is appreciated, and has over 100 Karma points.
Please Help! How to understand the "full scale peak-to-valley output" mentioned in IEC 60601-2-27

Dear everybody

Currently, i am conducting tests per IEC 60601-2-27. In clause 50, many subclause mentioned the term "full scale peak-to-valley output" .
ie, about the accuracy of signal reproduction test,the standard says "Adjust the SENSITIVITY of the EQUIPMENT to 10 mm/mV. Adjust the signal generator to produce a peak-to-valley output on a display device that is 10 % of the full scale peak-tovalley output. Increase the output of the signal generator by factors of 2, 5 and 10. The displayed output shall be linear within ± 20 % of the full scale output."

How can i understand the highlighted term, does it mean the channel width or the 10 mm due to the sensitivity.

thaks in advance!

Sponsored Links
  #2  
Old 22nd May 2011, 09:50 PM
Peter Selvey Peter Selvey is offline
Appreciated Information Resource

 
Registration Date: Oct 2009
Location: Japan / Ise
 
Posts: 629
Thanks Given to Others: 35
Thanked 474 Times in 324 Posts
Karma Power: 86
Karma: 4719
Peter Selvey is appreciated, and has over 1700 Karma points.
Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.
Re: How to understand the "full scale peak-to-valley output" mentioned in IEC 60601-2

For a printer strip that has for example 50mm, this would mean a signal that is approximately 5mV peak to peak (+2.5mV / -2.5mV). I say approximate, as most systems will have small sensitivity errors, so the standard usually wants you to adjust the source to give exactly full scale.

On a screen this may be more difficult to define the boundaries of "full scale", as some manufacturers allow the signal to drift well beyond the allocated space. But the designers of the screen layout should be able to indicate what they consider to be "full scale", e.g. point at which two adjacent signals will not overlap. This would be the channel width as you say.

On a related point, the new edition just published (IEC 60601-2-27:2011) has fixed many of the errors in the 2005 edition.

So for example, the 2005 edition says start at 10% of full scale, and increase to 20%, 50% and 100% which makes no sense from a laboratory point of view, because at 10% the resolution is too small.

The 2011 edition says start at 100% full scale, then reduce 50%, 20% and 10% which makes much more sense. Another example is the multichannel cross talk test which is now clearly defined where to apply the signal and where to verify the result.
Sponsored Links

  #3  
Old 22nd May 2011, 11:13 PM
raymond02302 raymond02302 is offline
Involved in Discussions

 
Registration Date: Mar 2011
Location: China
 
Posts: 56
Thanks Given to Others: 15
Thanked 21 Times in 17 Posts
Karma Power: 14
Karma: 115
raymond02302 is appreciated, and has over 100 Karma points.raymond02302 is appreciated, and has over 100 Karma points.
Re: How to understand the "full scale peak-to-valley output" mentioned in IEC 60601-2

Quote:
In Reply to Parent Post by Peter Selvey View Post

For a printer strip that has for example 50mm, this would mean a signal that is approximately 5mV peak to peak (+2.5mV / -2.5mV). I say approximate, as most systems will have small sensitivity errors, so the standard usually wants you to adjust the source to give exactly full scale.

On a screen this may be more difficult to define the boundaries of "full scale", as some manufacturers allow the signal to drift well beyond the allocated space. But the designers of the screen layout should be able to indicate what they consider to be "full scale", e.g. point at which two adjacent signals will not overlap. This would be the channel width as you say.

On a related point, the new edition just published (IEC 60601-2-27:2011) has fixed many of the errors in the 2005 edition.

So for example, the 2005 edition says start at 10% of full scale, and increase to 20%, 50% and 100% which makes no sense from a laboratory point of view, because at 10% the resolution is too small.

The 2011 edition says start at 100% full scale, then reduce 50%, 20% and 10% which makes much more sense. Another example is the multichannel cross talk test which is now clearly defined where to apply the signal and where to verify the result.
thank you for your detailded interpretation.
but i am still a little intangible.

On a screen, the "full scale" is the maximum scale designed by the manufacturer or the actual scale measured by us.
for example, the maximum scale designed by manufacturer is 30 mm (channel width) but the actual scale measured by us may be larger than that.
so, which is the full scale, the former or the latter?
  #4  
Old 22nd May 2011, 11:50 PM
Peter Selvey Peter Selvey is offline
Appreciated Information Resource

 
Registration Date: Oct 2009
Location: Japan / Ise
 
Posts: 629
Thanks Given to Others: 35
Thanked 474 Times in 324 Posts
Karma Power: 86
Karma: 4719
Peter Selvey is appreciated, and has over 1700 Karma points.
Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.Peter Selvey is appreciated, and has over 1700 Karma points.
Re: How to understand the "full scale peak-to-valley output" mentioned in IEC 60601-2

I would use the scale designated by the manufacturer.

For patient monitors, the manufacturer may allow the ECG signal to go beyond the channel so the user can still see the signal in case of overload or other problems with the baseline (e.g. poor electrode contact).

However, I don't think it is resononable to worry about accuracy and sensitivity in these cases. It is just a kind of user trouble shooting feature.

If the manufacturer says 30mm then test for 30mm.
Thanks to Peter Selvey for your informative Post and/or Attachment!
  #5  
Old 23rd May 2011, 05:42 AM
raymond02302 raymond02302 is offline
Involved in Discussions

 
Registration Date: Mar 2011
Location: China
 
Posts: 56
Thanks Given to Others: 15
Thanked 21 Times in 17 Posts
Karma Power: 14
Karma: 115
raymond02302 is appreciated, and has over 100 Karma points.raymond02302 is appreciated, and has over 100 Karma points.
Re: How to understand the "full scale peak-to-valley output" mentioned in IEC 60601-2

thank you
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
What's the best way to understand IEC 60601-1 standards luloo117117 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4 15th June 2009 09:37 AM
How to implement "assess" mentioned in the clause 7.1.4? zhuoyz2003 ISO 9000, ISO 9001, and ISO 9004 - Questions and Discussions 5 14th November 2008 11:41 PM
How to understand "Monitoring the Process"? "Measurement" of the Process? hwang - 2009 ISO/TS 16949 - International Automotive Quality Systems Standard 14 14th July 2006 07:57 AM
EQA malaysia is at it again - "Guaranteed certification" mentioned all the time darkgelap Registrars (Certification Bodies) and Registration 3 28th November 2004 08:53 AM
Full Scale Tolerance - How to apply full-scale accuracy specification to a gage? e006823 ISO 17025 and related Metrology Topics - Measurement Devices, Calibration and Test Laboratories 4 2nd August 2004 04:34 PM



The time now is 11:17 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


   


Marc Timothy Smith - Elsmar.com
8466 LeSourdsville-West Chester Road, Olde West Chester, Ohio 45069-1929
513 341-6272