Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)


The Elsmar Cove Forum SideBar!
Monitor the Forum
Monitor New Forum Posts
New Threads Feeds
RSS FeedRSS Feed
Sponsor Link










$ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:


Howard's International Quality Services

Atul's Symphony Technologies

Dave Scott's Scott Quality Solutions

Praxiom Research Group


NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest Portal

IEST - Institute of Environmental Sciences and Technology

ASQ - American Society for Quality


All the Important Standards and Related Web Sites in the World
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  #1  
Old 8th July 2002, 07:40 AM
rahayu rahayu is offline
Inactive Registered Visitor

Registration Date: Jul 2002
Location: Malaysia
Age: 32
 
Posts: 1
Thanks Given to Others: 0
Thanked 0 Times in 0 Posts
Karma Power: 30
Karma: 10
rahayu has less than 100 Karma points so far.
Question methods of reviewing confirmation intervals

Hi!
We've been using a OD gauge (75mm) in our lab for our daily OD meter daily verification. We've sent the gauge for calibration once a year.Below are the list of correction factor and measurement uncertainty (MU)written in the calibration certificate

Year Correction factor MU
1998 0 +/- 4um
1999 0.002 +/- 6um
2000 -0.003 +/- 0.007mm
2001 -0.003 +/- 0.002mm

The question is, I would like to analyse whether the interval need to be extend or reduced, waht are the method to be used.

In ISO 10012 : Part 1 : 1994,under A.3 Method of reviewing confirmation intervals, under A.3.1 method 1 is the Automatic or "Staircase" adjustment method. Under this method, "the subsequent intervals is extended if it is found to be within tolerance..". What does it mean by "tolerance" here?

Thank you.
Reply With Quote

Sponsored Links
  #2  
Old 1st August 2002, 11:57 AM
Graeme's Avatar
Graeme Graeme is offline
$ Contributor

Registration Date: Sep 2000
Location: Lilburn, GA, USA
Age: 60
 
Posts: 401
Thanks Given to Others: 0
Thanked 8 Times in 7 Posts
Karma Power: 63
Karma: 894
Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.
Send a message via AIM to Graeme Send a message via MSN to Graeme Send a message via Yahoo to Graeme
Be aware of the differences ...

Hello,

Since it has been a while and nobody else has posted a reply, I will have a go at it. I must say at the beginning, though, that dimensional measurement is not my main area of expertise. (I am an electronics person. )

Calibration interval analysis is a difficult area, especially when you are looking at a very small quanitiy of items - one, in your case. You can get a lot more information from NCSL Recommended Practice RP-1, Establishment and Adjustment of Calibration Intervals. You can purchase a copy from NCSL International, www.ncsli.org .

In your example, the tolerance that is referred to is the performance specification if the item being calibrated - the 75mm OD gage.
  • This is usually the manufacturer's published specification, if any. (I am not sure how the performance of these gages are specified by their manufacturers.)
  • It could also be a usability requirement that you set based on your own measurement needs. For example, you might decide the gage is no longer usable if the correction is more than some specified value.
In the method you describe, the calibration interval would be increased if the reported values are within those specifications.

This is about the simplest method to use, but it is in many ways the least useful. RP-1 notes several problems with this method.
  • This method makes adjustments based on essentially random results. Deming's funnel experiment is used to teach the futility of this.
  • This method cannot account for a target reliability for the equipment. For instance, there is no way to set and achieve a minimum reliability goal of (for example) 95% probability of being in-tolerance at the end of the period.
  • This method does not settle to a stable value for the calibration interval. If it accidentally arrives at a "correct" interval, the result of the next calibration will inevitably change it. Even if you attempt to compute a mean from the interval changes, the time required to reach a stable value often exceeds the useful life of the gage.

Since your gage has only a single measured parameter, I would suggest using Method 2, the next section from the one you cited. Plot the points on a run chart, or on a process decription ("control") chart for individual variables. You will be able to see any long-term trends, and the overall scatter of the points. Once you have "enough" points, a regression analysis will help you predict the future behaviour. If you also plot the calibration uncertainty as error bars, you will see how that relates to the reported value. Note that in all but the last calibration, the uncertainty of the measurement has been larger than the reported correction value, assuming I am interpreting your table correctly.

There is nothing "wrong" with keeping the same calibration interval for a gage like this over its lifetime, especially if you have only a small number of them. Yes, you can save money by calibrating less frequently. However, you also have to evaluate the increased risk of the tool going out of tolerance before the next calibration. Other methods of calculating calibration intervals can account for this risk, but require a large population of identical tools, or a long time period with fewer tools, to accumulate the data for a statisically valid analysis.
__________________
Graeme C. Payne
ASQ Sr. Member; CQE; CCT

"Does it matter if the measurement result is wrong?
If it does, then calibrate the instrument.
If it doesn't matter, they why are you making the measurement?"
(P. G. Stein, 2000)
Reply With Quote
Sponsored Links

  #3  
Old 1st August 2002, 04:46 PM
Al Dyer
Unregistered Guest

 
Posts: n/a
Default

This will sound negative, but is reads like you either have a bad calibration house or someone is sending different gages. There should not be that much variation in a gage over 4 years where it gets better, worse, better, worse.

As a line of thought I would think that any gage would get worse over time and usage.

I assume this is and ID/OD attribute gage?
Reply With Quote
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > ISO (International Organization for Standardization) Standards > ISO 17025 - Calibration, Measurement Gages and Test Laboratories > Calibration Frequency (Interval)

Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors and 1 Unregistered Guests)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump

Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Adjustment of Calibration Intervals (Calibration Frequency) Jerry Eldred Calibration Frequency (Interval) 14 19th February 2009 12:00 PM
Calibration Intervals (Frequency) derived from Variables Data rdragons Calibration Frequency (Interval) 30 12th August 2008 12:45 AM
Calibration Intervals (Frequency) for Stereoscopes, Microscopes, etc. freelovefest Inspection and Test, Sampling and Related Topics 2 12th October 2007 08:55 PM
Factors for Determining Gage Calibration Intervals - Extended Intervals Dave Hiser Calibration Frequency (Interval) 5 14th November 2002 06:01 PM
Determination of Calibration Intervals and Frequency Analysis Marc Calibration Frequency (Interval) 6 21st March 2000 05:50 PM



The time now is 05:35 AM. All times are GMT -4.
The time zone can be changed in your UserCP --> Options.



   

All Y'All Come Back Now, Y' Hear?

Made With A Mac! FreeBSD OS Powered by Apache!
Using php4 Forums provided and maintained by Marc Smith Database by MySQL

FAIR USE and CORRECTNESS NOTICE: This site contains copyrighted material the use of which has not always been specifically authorized by the copyright owner. We are making such material available in our efforts to advance understanding of environmental, political, human rights, economic, democracy, scientific, and social justice issues, etc. We believe herein constitutes a 'fair use' of any such copyrighted material as provided for in section 107 of the US Copyright Law. In accordance with Title 17 U.S.C. Section 107, the material on this site is distributed without profit to those who have expressed a prior interest in receiving the included information for research and educational purposes. For more information go to: http://www.law.cornell.edu/uscode/17/ If you wish to use copyrighted material from this site for purposes of your own that go beyond 'fair use', you must obtain permission from the copyright owner. In addition, I do not guarantee the correctness of the content. The risk of using content from the Elsmar Cove web site and forums remains with the user/visitor.

Responsibility Statement: Each person is responsible for anything they post in the Elsmar Cove forum. Neither I, Marc Timothy Smith, nor any of the forum Moderators, are responsible for the content of posts people make. Liability for post content resides with the poster as does interpretation and/or acceptance and/or use of advice by the reader.

Complaints: If you have a complaint with a post in a forum discussion thread, including Content in general, fighting, flaming, copyright infringement, defamation and/or 'slander', please use the 'Report This Post Report This Post Button button which appears at the top of every post in every thread.

Site courtesy of:
Marc Timothy Smith - Cayman Business Systems, 8466 Lesourdsville-West Chester Road, West Chester, Ohio 45069-1929 - USA
(513) 341-6272

To contact me, click the Google Voice link below, enter Your Name and Your Phone Number and Google will ring your phone and connect you for free!

The Elsmar Cove Web Site is *CopyFree*
no new posts