Determining Working Voltage - Drawing a Medical Device Insulation Diagram


Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed

Monitor the Elsmar Forum
Sponsor Links



Courtesy Quick Links

Links that Cove visitors will find useful in your quest for knowledge:


ASQ - American Society for Quality

International Standards Organization - ISO Standards and Information

Atul's
Quality Forum Online

Howard's
International Quality Services

Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum

Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Related LinkedIn Groups

ISO 9001 for Small Businesses

ISO 9001:2015 Revision Discussions

Information Security Community

Medical Devices Group

Quality and Regulatory Network

FDA (Food and Drugs)

AS91XX Series - Tips and Advice


Related Topic Tags
electrical - voltage resistance amperage etc., medical device electronics and electrical equipment, medical devices (general), voltage (electrical)
Reply
 
Thread Tools Search this Thread Rating: Thread Rating: 1 votes, 5.00 average. Display Modes
  Post Number #1  
Old 16th December 2011, 11:39 AM
Roland chung

 
 
Total Posts: 402
Please Help! Determining Working Voltage - Drawing a Medical Device Insulation Diagram

Hello all,

I am drawing an insulation diagram for class II equipment with BF type applied part which employs a 60950 power supply. Barrier A between Mains (240 V) and secondary circuit (48 V) meets 2 MOOP/ 1 MOPP (288 V). There is a barrier B between secondary (48 V) and applied part (12 V). Since the barrier A is 2 MOOP/ 1 MOPP @ 288 V, the barrier B shall be:
2 MOPP (60 V)
1 MOPP (XX V)
XX = either 240V or 288V.

My question is which working voltage shall be chosen and why.

Thanks,
Roland

Last edited by Roland chung; 16th December 2011 at 12:16 PM. Reason: add words: class II

Sponsored Links
  Post Number #2  
Old 16th December 2011, 11:50 AM
Pads38

 
 
Total Posts: 595
Re: Determining working voltage

250V

Quote:
8.5.2.1 * F-TYPE APPLIED PARTS
The PATIENT CONNECTION(S) of any F-TYPE APPLIED PART shall be separated from all other parts, including the PATIENT CONNECTION(S) of other APPLIED PARTS, by means equivalent to one MEANS OF PATIENT PROTECTION for a WORKING VOLTAGE equal to the MAXIMUM MAINS VOLTAGE
Maximum mains voltage is given in 8.5.3 and is 250V for just about everything.
Sponsored Links

  Post Number #3  
Old 16th December 2011, 12:49 PM
Roland chung

 
 
Total Posts: 402
Re: Determining Working Voltage - Drawing a Medical Device Insulation Diagram

Thank you pads.

I agree with the 250V BUT for the barrier between applied part and accessible part.

What I am concerned about in my case is barrier from Mains to applied part shall be 2 MOPP (barrier A + B). The working voltage for barrier B should seem to be same as barrier A based on the principle of DOUBLE INSULATION. But image barrier A fails, Mains voltage will stress the barrier B. Er, a bit confusing.
  Post Number #4  
Old 18th December 2011, 12:30 AM
raymond02302

 
 
Total Posts: 61
Re: Determining Working Voltage - Drawing a Medical Device Insulation Diagram

The working voltage (for 2 MOOP) for barrier A is not always the sum of mains voltage and secondary voltage. It should be the voltage actually stress barrier A. So you should measure the voltage.
Regarding your question, we should chose the least feavourable between the conditons below
1) We consider mains voltage as the hazard resource. To protect the patient, Barrier A+B should provide 2MOPP with working voltage mains voltage (240 V for your case), therefore, barrier B should provide 1MOPP with working voltage 240.
2) Considering the secondary voltage (48 V) as the hazard resource, So, barrier B should provide 2MOPP with working voltage actually stress barrier B.
  Post Number #5  
Old 18th December 2011, 10:57 PM
Roland chung

 
 
Total Posts: 402
Re: Determining Working Voltage - Drawing a Medical Device Insulation Diagram

Again, I am talking about the case that 2MOPP is divided to 2 single MOPP. However, working voltage of one concerned MOPP (288 V) differ from the working voltage of the whole 2MOPP (240V). In this case, how to consider the working voltage of the other concerned MOPP?

Peter, do you have any comments?
  Post Number #6  
Old 19th December 2011, 12:21 PM
Shine Zhang

 
 
Total Posts: 6
Re: Determining Working Voltage - Drawing a Medical Device Insulation Diagram

The working voltage you mentioned here might be different and it shall be decided case by case basing on the circuit analysis.
I suppose that barrier A is separated by a linear transformer since the working voltage was chose as 288V, so another MOPP on B could be 288V at the worst case in theory.
but in general, I think 250V is enough, especially, A is separated by SMPS. Although the priciple of DI-the working voltage is the whole to which it is subjected, however, B will not subjected 288V when failed in A (SFC) but 250V for AC mains.
The question is when failed in A, what voltage may stress on B considering the actual circuit and possible connection to secondary circuit, also earthed or not earthed of applied part.
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series

Bookmarks



Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Free Quality Tools for drawing Ishikawa Diagram, Pareto Diagram, Histogram, etc. venkat Quality Assurance and Compliance Software Tools / Solutions 5 28th October 2013 01:22 PM
Insulation Diagram for Water Tub netseal IEC 60601 - Medical Electrical Equipment Safety Standards Series 4 14th February 2013 05:22 AM
How to consider the Working Voltage for splitting 2MOPP Roland chung IEC 60601 - Medical Electrical Equipment Safety Standards Series 35 20th July 2012 12:12 AM
Insulation diagram for a medical electrical equipment Tkerrick IEC 60601 - Medical Electrical Equipment Safety Standards Series 4 22nd October 2009 10:53 PM
Medical device with NON medical PC - USB galvanic insulation AmirBaron ISO 13485 and ISO 14969 - Medical Devices - Quality Management Systems 2 18th November 2008 06:01 AM