The Cove Business Standards Discussion Forums More Free Files Forum Discussion Thread Post Attachments Listing Cove Discussion Forums Main Page
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series
Forum Username
Wooden Line
IEC 60601 Ground Bond Test Requirements for a Plastic Enclosure
Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed
Monitor the Elsmar Forum
Sponsor Links
Courtesy Quick Links
Links that Cove visitors will find useful in your quest for knowledge:
International Standards Bodies - World Wide Standards Bodies
International Standards Organization - ISO Standards and Information
Howard's
Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum
Bob Doering
Bob Doering's Blogs and,
Quality Digest
Related Topic Tags
class i medical device, iec 60601 - medical electrical equipment, medical equipment (general)
Reply
  Post Number #1
Old 18th May 2012, 10:57 AM
ale_pb
Please Help! IEC 60601 Ground Bond Test Requirements for a Plastic Enclosure
Hello
My question is:
Let's say that I have a CLASS I EQUIPMENT, with a plastic enclosure and no accessible metal parts.
Reading the 60601-1 it seems to me that, if I don't have any accessible metal parts, the ground bond test is not necessary. Am I correct?
Best Regards
Sponsored Links
  Post Number #2
Old 18th May 2012, 11:55 AM
Pads38
Re: Ground bond test - Plastic enclosure
The ground bond test verifies the impedance of the protective earth. If you have a protective earth (defined term in 60601-1) then you need the test.
For a type approval test this could be done 'with the lid off' but it is also one of the tests that a hospital service department would expect to do and you run the risk of problems if you don't make it easy for them. (Things like trying to use 25A test current on a point that goes via a small circuit board trace).
I have found that adding an equipotential point on the rear panel gives a nice easy place to test without being difficult or expensive to fit.
Thanks to Pads38 for your informative Post and/or Attachment!
Sponsored Links
  Post Number #3
Old 18th May 2012, 12:04 PM
ale_pb
Re: IEC 60601 Ground Bond Test Requirements for a Plastic Enclosure
Thanks Pads.
  Post Number #4
Old 18th May 2012, 12:28 PM
Marcelo Antunes's Avatar
Marcelo Antunes
For your equipment to be class I, it has to have accesible or internal parts that are protectively earthed. As yours do not have accessible protectively earthed parts, i'm assuming your equipment has internal parts which are protectively earthed.
The ground bound test verify that protectively earthed parts, accessible or internal, have the right impedance. So, assuming that your have those internal parts, the test has to be done in those parts.
If your equipment do not have protectively earthed parts, it cannot be class I.
Thanks to Marcelo Antunes for your informative Post and/or Attachment!
  Post Number #5
Old 20th May 2012, 07:54 PM
Peter Selvey
Under IEC 60601-1:1988/A2:1995 this is poorly written part of the standard so easy to overlook.
Clause 18g covers only accessible protectively earthed metal parts, with the typical 25A / 0.1ohm test.
Clause 18f covers "other" protectively earthed metal parts and then goes on to describe an alternate test. The way it is written means it is often mistakenly considered not applicable if the alternate test is not applied, and the IECEE CB scheme protocol has promoted this mistake (many test engineers read only the protocol, not the standard). As such, the test of internal protectively earthed parts is frequently overlooked.
But the correct interpretation is that for internal parts, either the test of 18g (i.e. 0.1ohm / 25A) or the alternate test applies. Either way, internal protectively earthed parts must be tested with the result recorded under Clause 18f.
The 3rd edition, IEC 60601-1:2005 has removed the word "accessible" from the applicable clause, so the problem no longer exists.
Thanks to Peter Selvey for your informative Post and/or Attachment!
Reply
Bookmarks
Search this Thread:
Advanced Forum Search
Rate Thread Content:
Forum Posting Settings
BB code is On
Smilies are On
[IMG] code is On
Similar Discussion Threads
Plastic of an infusion pump enclosure and ISO 10993 tzamam ISO 13485:2016 - Medical Device Quality Management Systems 2 26th May 2014 05:11 PM
Enclosure - IEC 60601-1 definition duck5 Definitions, Acronyms, Abbreviations and Interpretations 2 17th June 2013 01:54 AM
IEC 60601-2-49 - Touch Current (Enclosure Leakage Current) requirements Riskcord IEC 60601 - Medical Electrical Equipment Safety Standards Series 2 14th March 2013 06:09 AM