IEC 60601-1 Single fault conditions of electronic PCB components

Search the Elsmar Cove
Search Elsmar
Monitor the Elsmar Forum
Follow Marc & Elsmar
Elsmar Cove Forum RSS Feed  Marc Smith's Google+ Page  Marc Smith's Linked In Page   Marc Smith's Elsmar Cove YouTube Page  Marc Smith's Facebook Page  Elsmar Cove Twitter Feed
Elsmar Cove Groups
Elsmar Cove Google+ Group  Elsmar Cove LinkedIn Group  Elsmar Cove Facebook Group
Donate and $ Contributor Forum Access
Courtesy Quick Links

Links that Elsmar Cove visitors will find useful in your quest for knowledge:

Howard's
International Quality Services
Marcelo Antunes'
SQR Consulting
Bob Doering's
Correct SPC - Precision Machining

NIST's Engineering Statistics Handbook
IRCA - International Register of Certified Auditors
SAE - Society of Automotive Engineers
Quality Digest Portal
IEST - Institute of Environmental Sciences and Technology
ASQ - American Society for Quality

Related Topic Tags
battery powered devices, iec 60601 - medical electrical equipment, pcba (printed circuit board assembly), risk management and analysis
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 30th April 2013, 02:51 AM
MDDPeter

 
 
Total Posts: 6
Please Help! IEC 60601-1 Single fault conditions of electronic PCB components

Hi All,

We are embarking on a new design and we were wondering to what degree single fault conditions on electronic PCB components must be considered in the design. All PCB mounted components are likely to be surface mounted.
In particular we are referring to the circuits that might be patient contacting where the 10uA current limit applies with a 50uA limit under single fault conditions.

For a Class II medical device, battery powered, non-mains connected, non-earth connected and fully enclosed in a plastic enclosure (with the exception of patient contacts), what do people regard as being adequate single fault risk analysis (ISO 14971 also ties into this).

Two-lead passive components (caps and resistors)
For example, if a patient contacting probe is supplied by a 3V DC battery and current limited through two resistors before returning to 0V. One resistor is 10K, the second 330K… would a single fault short circuit in one of the resistors need to be considered as a potential fault? Note that if the 330K resistor fails short, the current might exceed the 10uA limit. The likelihood of a surface mount resistor going short circuit in electronics is extremely low.

Simple semi-conductors (e.g. MOSFETs)
Similar to the above, if you introduce a MOSFET to switch the current on and off. Would you consider a MOSFET going short circuit as a potential single fault? Again, MOSFETs are very reliable components if they are operated within their bounds.


Integrated circuits (e.g. logic gates and MUXs)
Again, based on the above, if you include some type of multiplexer or other IC, would you consider this device failing as a short circuit to 0V as a failure mode? If it were to go short, there is a risk that the 330K resistor might be shorted and thus current higher than 10uA might flow.

Programmable code executing semi-conductors (e.g. micro-controllers)
If we now tie the input to an analogue input on a micro-controller, there is a risk of failure whereby the micro fails creating a short to ground, again bypassing the 330K resistor. Would that be a single failure mode that would need to be mitigated against?

The IEC 60601-1 standard and ISO 14971 risk analysis standard are a little vague on this and allow for “expert opinion”.


I guess part of the question is, where do you stop with your single fault risk analysis?

Sponsored Links
  Post Number #2  
Old 30th April 2013, 04:08 AM
Peter Selvey

 
 
Total Posts: 772
Re: IEC 60601-1 Single fault conditions of electronic PCB components

Big subject, but useful for everyone.

There is a fundamental problem in the use of the single fault condition (SFC). It served us well for 50 years or so, but it needs to give way to a broader concept of simply providing protection against potential hazards.

The problems with the SFC are:
(1) if applied strictly, it implies an endless FMEA analysis, not only considering thousands of possible fault conditions, but each of these need to be considered under a range of operating conditions, settings etc, possible multiple faults etc etc, reaching millions of combinations;

(2) it often tricks designers into focusing on fault conditions rather than well designed protection;

(3) it ignores other events such as user mistakes, clinical events, the environment which also require protection systems. With the standard expanding into performance and clinical issues, these events are often more important than component faults because they occur at higher frequency, yet fault conditions get more air time;

(4) it fails to capture an assessment of the reliability of the protection system as being appropriate, taking into account the probability of the triggering event (fault, user error, etc) and the severity of the potential harm.

Following ISO 14971, the focus should simply be on designing effective protection against identified hazards. If the protection is reasonably independent, simple, reliable, it should obviously mitigate the risk irrespective of the triggering events. Single fault conditions may form part of the verification tests, but they should not the driver for the original analysis and design of the risk control.

In the particular case in hand, we have an identified risk from patient auxiliary current, with a limit of 10uA in normal condition and 50uAdc in abnormal condition.

Rather than focusing on faults, the designer should ask, what component(s) or feature provide protection? You could intentionally split the 330k resistor into two 160k resistors, such that failure of either one ensure the 50uA cannot be exceeded (kind of double insulation).

Or, you could reasonably argue that a 1608 SMD 330k resistor is being used so far below it's ratings that it will never short, and then write it up as high integrity component (equivalent to reinforced insulation).

Either way, there should be a specific part which is identified as providing the "protection".

Once that is in place, you don't need to worry about MOSFETs shorting, logic circuits or software failures. You might select to short the MOSFET as a worst case verification test, or write some special software that turns it on continuously. Or could just do by inspection 3V / 160k = 18uA --> Pass. All of these are valid methods for verification of a risk control.

So you can see, the single fault condition is really just one of the options you can use in verification, and that's where it should be relegated to.
Thank You to Peter Selvey for your informative Post and/or Attachment!
Sponsored Links

  Post Number #3  
Old 30th April 2013, 11:46 PM
MDDPeter

 
 
Total Posts: 6
Re: IEC 60601-1 Single fault conditions of electronic PCB components

Thanks Peter. I interpret this as "...we must make the device inherently safe by design no matter what the SFC and thus reduce the risk to broadly acceptable".

Thanks,
Peter
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Forum > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series

Do you find this discussion thread helpful and informational?


Bookmarks


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
Requirements for PCB Component Replacements in an IEC 60601-1 Compliant Device nbarabino IEC 60601 - Medical Electrical Equipment Safety Standards Series 9 14th December 2012 07:54 AM
IEC 60601-1 Least Favourable Working Conditions for Medical Equipment indorewalep IEC 60601 - Medical Electrical Equipment Safety Standards Series 3 20th November 2012 07:21 PM
Single Fault Condition for 3-Phase Medical System - IEC 60601 Clause 8.1b gregor_l IEC 60601 - Medical Electrical Equipment Safety Standards Series 3 15th December 2011 08:49 PM
Replacing End of Life (EOL) Components - What triggers retesting to IEC 60601-1 freelui IEC 60601 - Medical Electrical Equipment Safety Standards Series 6 19th July 2011 11:57 AM
Define Fault, Double Fault and Normal Conditions - Preparing a Risk Management File Bunny FMEA and Control Plans 1 17th November 2009 12:42 PM