How to fulfill IEC 60601-1-2 EMC Testing Requirements


Elsmar XML RSS Feed
Elsmar Cove Forum RSS Feed

Monitor the Elsmar Forum
Sponsor Links



Courtesy Quick Links

Links that Cove visitors will find useful in your quest for knowledge:


ASQ - American Society for Quality

International Standards Organization - ISO Standards and Information

Atul's
Quality Forum Online

Howard's
International Quality Services

Marcelo Antunes'
SQR Consulting, and
Medical Devices Expert Forum

Bob Doering
Bob Doering's Blogs and,
Correct SPC - Precision Machining

NIST's Engineering Statistics Handbook

IRCA - International Register of Certified Auditors

SAE - Society of Automotive Engineers

Quality Digest

IEST - Institute of Environmental Sciences and Technology


Related LinkedIn Groups

ISO 9001 for Small Businesses

ISO 9001:2015 Revision Discussions

Information Security Community

Medical Devices Group

Quality and Regulatory Network

FDA (Food and Drugs)

AS91XX Series - Tips and Advice


Related Topic Tags
emc (electromagnetic compatibility), iec 60601 - medical electrical equipment
Reply
 
Thread Tools Search this Thread Rate Thread Content Display Modes
  Post Number #1  
Old 26th September 2013, 04:00 AM
tallu

 
 
Total Posts: 55
Please Help! How to fulfill IEC 60601-1-2 EMC Testing Requirements

Hi Cove Members,

I need some opinions for accomplishing EMC tests.

Here is the brief description of the possible situation:
Your medical device System (consist of main unit and additional units) is EMC tested. You are going to release new unit which is part of the system and it's used with main unit.

Dilemma is that, is it acceptable to perform EMC test only for the new unit? Is there requirement(s) to perform EMC test for the whole system which consists main unit and this new unit?

Sponsored Links
  Post Number #2  
Old 27th September 2013, 03:54 AM
tallu

 
 
Total Posts: 55
Re: How to fulfill IEC 60601-1-2 EMC Testing Requirements

Any thoughts?
Sponsored Links

  Post Number #3  
Old 27th September 2013, 04:16 AM
Pads38

 
 
Total Posts: 585
Re: How to fulfill IEC 60601-1-2 EMC Testing Requirements

I would suggest that you discuss the situation with your EMC test house. If you can get them to agree that the changes are very unlikely to influence the EMC behaviour then that may be enough.

What may be possible is to agree, with the test house, that only certain parts of the test programme need to be repeated. So, if you look through your test results and find that, say, the radiated RF immunity was OK but nearly caused problems at 2GHz, horizontal polarization, then just repeat that part of the tests.

How successful this approach would be is somewhat dependent upon how constructively helpful your test house is. They are, naturally, rather conservative but if the approach throughout is one of constructive problem solving (rather than box ticking) can be very helpful.
Thanks to Pads38 for your informative Post and/or Attachment!
  Post Number #4  
Old 27th September 2013, 05:19 AM
tallu

 
 
Total Posts: 55
Re: How to fulfill IEC 60601-1-2 EMC Testing Requirements

Good point. I had similar point of view.

But it might be easier to test whole system. Feels like it makes device registration and approval processes more clear.

You dont have to combine information between two different EMC Test Reports and explain the relations etc.
  Post Number #5  
Old 27th September 2013, 05:58 AM
katapult

 
 
Total Posts: 7
Re: How to fulfill IEC 60601-1-2 EMC Testing Requirements

Hi Tallu,

I'd suggest to test the whole system in combination with the new part. In general you have to prove the combinability of the new part with the system. So why not check EMC too? If you don't check EMC with the system, you're going to have to explain your testing for every auditor again and again.
  Post Number #6  
Old 2nd October 2013, 12:31 AM
tomshoup

 
 
Total Posts: 11
Re: How to fulfill IEC 60601-1-2 EMC Testing Requirements

The device is the main unit plus whatever "other units" are specified to connect to the main unit. If the new accessory unit is intended to work with the main unit, then the device (main + accessories) has to be tested. Also, if the new accessory unit is expected to work with other generations of main unit or other generations of accessories, then you really should test all combinations. It comes down to what is the device, it's the main unit + accessory units, and that is what has to pass EMC.
Reply

Lower Navigation Bar
Go Back   The Elsmar Cove Business Systems and Standards Discussion Forums > Medical Devices, Medical Information Technology, Medical Software and Health Informatics > Medical Device Related Standards > IEC 60601 - Medical Electrical Equipment Safety Standards Series


Visitors Currently Viewing this Thread: 1 (0 Registered Visitors (Members) and 1 Unregistered Guest Visitors)
 
Thread Tools Search this Thread
Search this Thread:

Advanced Forum Search
Display Modes Rate Thread Content
Rate Thread Content:

Forum Posting Settings
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off


Similar Discussion Threads
Discussion Thread Title Thread Starter Forum Replies Last Post or Poll Vote
What is to be documented to fulfill the requirements of ?820.65 Traceability? Philipp 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1 8th March 2012 02:55 PM
Brazilian Requirements for Medical Device Production Testing to IEC 60601-1 MNewberry IEC 60601 - Medical Electrical Equipment Safety Standards Series 2 23rd September 2011 07:19 AM
Predicate Device - Question on Biocompatibility and IEC 60601-1 Testing Requirements melissa66 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5 2nd January 2010 02:21 AM
Does that Q-Policy fulfill ISO 9001 requirements? Buetze ISO 9000, ISO 9001, and ISO 9004 - Questions and Discussions 4 10th January 2008 02:54 AM
Medical Device EMC requirements - Canada - Testing to FCC part 18, EN 55011, EN 60601 amjadrana IEC 60601 - Medical Electrical Equipment Safety Standards Series 2 1st June 2005 03:40 PM



The time now is 04:03 PM. All times are GMT -4.
Your time zone can be changed in your UserCP --> Options.


 


NOTE: This forum uses "Cookies" -- The Elsmar Cove is *Copyright Free*.
A Peachfarm LLC Internet Property