Question Laser Micrometer Uncertainty - Calibrating Deltronic Pin Gages
Hi guys,
I have a question on a uncertainty calculation that I'm trying to determine.
I want to determine the uncertainty of a measurement using my Zygo Gold laser micrometer when we calibrate Deltronic Pin Gages. Here is the info:
Zygo laser micrometer resolution: .000001"
Zygo laser micrometer accuracy: .00002"
Zygo laser micrometer repeatability: .000005"
Pin Gage Standard XXX (used to cal laser mic): .00001"
By calculating the above, I get an expanded uncertainty of .000026".
My question is: should I be including the accuracy of the Deltronic Pin Gage I'm calibrating? It has a standard accuracy of +.00004 -0".
If I do, then my expanded uncertainty is .000053" which seems higher than I expected. If I do use this, should I subtract this value from the .00004" tolerance of the Deltronic Pin Gage to come up with an uncertainty of .000013"?
Please help a confused soul.
Sponsored Links
Old 29th April 2003, 03:54 PM
Al Dyer's Avatar
Al Dyer Al Dyer is offline
Mod...
Registration Date: Mar 2003
Thanks Given to Others: 4
Karma Power: 46
Al Dyer is appreciated, and has over 300 Karma points.Al Dyer is appreciated, and has over 300 Karma points.Al Dyer is appreciated, and has over 300 Karma points.Al Dyer is appreciated, and has over 300 Karma points.
Send a message via AIM to Al Dyer
Default
My advice would be to drop the pin from the laser mike uncertainty calculation.
The laser is being reviewed, not the pin, and you should be able to use any size pin and get the same uncertainty.
Try a test with increasing sized pins and see if the linearity of the laser is stable
"Para ... Bellum"
Sponsored Links
Old 2nd May 2003, 04:03 PM
Graeme's Avatar
Graeme Graeme is offline
$ Contributor
Registration Date: Sep 2000
Karma Power: 63
Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.Graeme is appreciated, and has over 700 Karma points.
Send a message via AIM to Graeme Send a message via MSN to Graeme Send a message via Yahoo to Graeme
Default Re: Uncertainty question
The measurement uncertainty applies to the measuring system. A measuring system is independent of the items it measures.
The uncertainty is used to help determine if the object being calibrated (the pin gage) is within its own accuracy specification.
The ratio between the measurement uncertainty (of the measuring system) and the accuracy specification (of the object being calibrated) is an indicator of the ability of the system to make that particular measurement.
If the object being calibrated was included as part of the measurement system, then it would be reasonable to be required to perform an MSA for every workload item! (Impractical, outrageous and uneconomic, but weirdly reasonable. )
Graeme C. Payne
ASQ Sr. Member; CQE; CCT
"Does it matter if the measurement result is wrong?
If it does, then calibrate the instrument.
If it doesn't matter, they why are you making the measurement?"
(P. G. Stein, 2000)
Old 12th May 2003, 12:02 PM
Ryan Wilde's Avatar
Ryan Wilde Ryan Wilde is offline
Registration Date: Feb 2001
Karma Power: 41
Ryan Wilde is appreciated, and has over 100 Karma points.Ryan Wilde is appreciated, and has over 100 Karma points.
Hi guys,
Using all rectangular distributions, I got about the same. (0.0000265")
You wouldn't take into account the accuracy of the Deltronic Pin Gage, especially since that is what you are trying to determine. You would, however, need to incorporate the repeatability of measuring a deltronic gauge. Using your system, I would guess this to be around 0.000005" with a single standard deviation.
At this point, we have an uncertainty of about 0.000028".
What we are forgetting is the environmental effects. Temperature is the big wildcard on all dimensional measurements. If your laboratory is 68 ±5 °F (20 ± 2.8 °C), and you don't compensate for the thermal effects, you will have a very different uncertainty.
Thermal expansion of tool steel (such as a Deltronic Pin) is 6.4 µin/in/°F (11.5 µin/in/°C).
Therefore, if you are measuring a 0.5" pin at 72 °F (22.2 °C), and the pin is exactly 0.5" at standard temperature (68 °F/ 20 °C), you would actually measure [0.5" + (0.0000064" *0.5" *4 °F)], or 0.500013". With everything else in the system being perfect (which it isn't), you have an error of 0.000013" due to temperature alone.
I do not know the temperature effect on the Zygo laser, but thermal expansion of the pin, and thermal effects on the Zygo laser must be added to the uncertainty budget.
No, the tolerance of the Deltronic pin has no bearing on your measurement. The tolerance of the pin under test is entirely arbitrary to the uncertainty of measuring the pin.
Ryan
Old 15th May 2003, 12:50 PM
Involved - Posts
Karma Power: 33
Default
Thanks everyone for your feedback.
Old 15th May 2003, 06:57 PM
Marc's Avatar
Marc Marc is offline
Your Elsmar Cove Host
Registration Date: Jan 1996
Thanks Given to Others: 1,895
Karma Power: 605
Lurker Link
Also see this thread for some thoughts on measurement uncertainty. It's a relative oldie (check some of the post dates), but it's a goodie:
Types of Measurement Uncertainty of Different Devices - Is there any Standard?
Reply
Bookmarks
Search this Thread:
Advanced Forum Search
Rate Thread Content:
BB code is On
Smilies are On
[IMG] code is On
Forum Jump
Similar Discussion Threads
Choosing a Laser Micrometer - Your recommendations? mhirand20 ISO 17025 - Calibration, Measurement Gages and Test Laboratories 6 24th February 2009 01:50 PM
Calibrating pin gages - Can we write a procedure "Calibrate before use"? Examples? al40 ISO 17025 - Calibration, Measurement Gages and Test Laboratories 14 19th March 2008 11:19 AM
Micrometer resolution for calibrating thread plug gages RESET ISO 17025 - Calibration, Measurement Gages and Test Laboratories 7 5th May 2006 11:38 PM